#13810 - 17/12/03 02:55 PM
Re: Eschmann RX 500 operating table
|
Philosopher
Registered: 11/07/00
Posts: 969
Loc: Stockport, Cheshire, England
|
We've got the RX500 and 600 tables and both have had problems with fluid ingress into the base - where all the electronics, hydraulics, batteries and motors are located ! The problem is usually with the rubber sealing strips between the two halves of the cowling or between the cowling and the pedestal. Usually it just blows the fuse, but it has caused some odd effects on occasions.
The handsets are usually just smashed by being dropped on the floor or swinging about at the end of their cable. If you need replacement handsets be careful you get the right ones because the newer one has an extra function button - but the same plug, so it plugs into the older table and the buttons do different things than they are intended to do !
_________________________
Today is the day you worried about yesterday - and all is well !
|
|
Top
|
|
|
|
#13812 - 18/12/03 12:25 PM
Re: Eschmann RX 500 operating table
|
Philosopher
Registered: 11/07/00
Posts: 969
Loc: Stockport, Cheshire, England
|
Because you need to retain access to the components inside the base, you can't seal the covers "permanently". Our theatres have tried some waterproof drapes which prevent fluid getting onto the top of the base and hence through the joins. Even in the parked position, the base is high enough off the floor to be clear of any spills - unless you operate under several inches of water ! I'll try and find out where they got them from and iff they were a success. Our base covers got bent out of shape when they wheeled a "C" arm X-ray machine into it ! 
_________________________
Today is the day you worried about yesterday - and all is well !
|
|
Top
|
|
|
|
|
|
10 registered (Nozza, DaveC in Oz, Chris Glass, Mofe, Huw, kmadhavesh, Ed SWM, Pete B, READ, ipetev),
297
Guests and
12
Spiders online. |
|
Key:
Admin,
Global Mod,
Mod
|
|
|
|
|
1
|
2
|
3
|
4
|
5
|
|
6
|
7
|
8
|
9
|
10
|
11
|
12
|
|
13
|
14
|
15
|
16
|
17
|
18
|
19
|
|
20
|
21
|
22
|
23
|
24
|
25
|
26
|
|
27
|
28
|
29
|
30
|
31
|
|
|
|
|