ebme medical engineering website
Page 1 of 2 1 2 >
Topic Options
Rate This Topic
#27403 - 14/01/08 02:40 PM Evita 2 & 2 Dura
micknand Offline
Scholar

Registered: 23/01/05
Posts: 60
Loc: Calderdale
Hi

We have the Drager Evita 2 and 2 Dura and are experiencing regular "Flow Inop" messages. This across is all the ventilators. We have serviced the inspiratory valve.

Any one else got this problem? Or a remedy for it?

Mick
_________________________
I was looking back to see if she was looking back to see if I was looking back at her!

Top
#27408 - 14/01/08 03:57 PM Re: Evita 2 & 2 Dura [Re: micknand]
JAKE Offline
Newbie

Registered: 19/05/04
Posts: 3
Loc: Darlington Memorial Hospital
Hi
If you are using disposable exp valves I would suggest that you have some that have a harder rubber seal for the flow sensor connection. This can prevent the flow sensor from being fully engaged to the right (ie not activating the microswitch).
Draeger are aware of the issue and will exchange.

Paul

Top
#27513 - 20/01/08 06:58 AM Re: Evita 2 & 2 Dura [Re: JAKE]
gringo Offline
Scholar

Registered: 12/09/05
Posts: 71
Loc: Novi Sad
the new generation of flow sensors are slightly different in shape and size than the old ones (yellowish colour). we had similar problems with new ones but most of the time it was the switch engaging problem. Just pay a little more attention when replacing
best regards
Milan

Top
#27519 - 21/01/08 08:19 AM Re: Evita 2 & 2 Dura [Re: gringo]
KM Offline
Philosopher

Registered: 30/08/01
Posts: 754
Loc: LHCH
Weve experienced this with flow sensors problems on all Drager Evita vents.
Drager have been informed of the fact.
Weve found that the contacts at the machine end need cleaning with a little contact cleaner. After this the m/c works first time with the sensor that originally failed.
Buzz the flow sensors out before you bin them, dont assume they are faulty. Or if you have a magnifier check the flow sensor wires in out to see theres no breakage.

Top
#29308 - 15/04/08 12:17 AM Re: Evita 2 & 2 Dura [Re: KM]
dennisbacani Offline
Newbie

Registered: 08/04/08
Posts: 7
Loc: Philippines
hi to everybody


I have a prolem on evita 2 dura, FIO2 high i check on the O2 analyzer and it's really high on 21% setting i measure 33% is it the O2 mixer problem please help and if someone have a service manual it will be a big help

thank a lot
dennis

Top
#29358 - 16/04/08 06:32 AM Re: Evita 2 & 2 Dura [Re: dennisbacani]
gringo Offline
Scholar

Registered: 12/09/05
Posts: 71
Loc: Novi Sad
try recalibrating the O2 sensor or replacing it. It usually solves the problem.
best regards
Milan

Top
#29361 - 16/04/08 08:32 AM Re: Evita 2 & 2 Dura [Re: gringo]
micknand Offline
Scholar

Registered: 23/01/05
Posts: 60
Loc: Calderdale
And to add to the "knowledge base"

Following from the original post - we replaced the socket and lead, ot appears to have solved the problem and quite cheaply too.

Mick
_________________________
I was looking back to see if she was looking back to see if I was looking back at her!

Top
#29378 - 16/04/08 12:04 PM Re: Evita 2 & 2 Dura [Re: dennisbacani]
dennisbacani Offline
Newbie

Registered: 08/04/08
Posts: 7
Loc: Philippines
Hi everybody

Is anyone encounter problem with EVITA 2 DURA, FIO2 high alarms, setting at 21% and it monitors 33% even you check on oxygen analyzer it really reads 33%, when you remove the Oxygen gas it will read 21%, is it the problem on O2/AIR mixer pls advice

thanks
dennis

Top
#29379 - 16/04/08 12:15 PM Re: Evita 2 & 2 Dura [Re: dennisbacani]
Geoff Hannis Online   content
Super Hero

Registered: 12/02/04
Posts: 11368
Loc: the path less trodden
Misaligned control knob? smile


Edited by Geoff Hannis (16/04/08 12:17 PM)
Edit Reason: Hit or miss

Top
#29390 - 16/04/08 01:20 PM Re: Evita 2 & 2 Dura [Re: Geoff Hannis]
BBottomley Offline
Mentor

Registered: 07/03/01
Posts: 165
Loc: Huddersfield Royal Infirmary, ...
Misaliged control knob - eh??? sick

Dennis - you may have a problem with the basic calibration mechanism in the Evita. I have had this on Evita 4's and suspect that the methodology is the same for the Dura.

When the O2 cell is being calibrated there is an automatic sequence which provides air and oxygen separately, to provide a known 21% and 100% O2 calibraion point.

This is achieved by a small shuttle valve being driven across the inlet to the O2 cell and providing the requiste gas at the appropriate time. The valve seal is provided by a rubber/neoprene? disc which should eliminate any dilution of the intended calibration gas. If the disc wears then some dilution can occur and the calibration settings are assumed by the vent to have taken place as expected, however there will have been some electronic bias placed on the measuring circuit to compensate for this drift.

In your case, if oxygen is leaking round the seal when the vent assumes a 21% cal point to have been accepted, it will allow enough oxygen into the patient circuit to achieve the same O2 cell voltage, thus giving you an unexpectedly high O2 reading both on the vent and your monitor.

You can fairly easily check to see if you have this problem - assuming the Dura to be the same as the Evita - by removing the O2 cell and looking into the opening thus revealed in the calibration block. The black sealing disc in question should be visible seated neatly on a brass shaft in the centre of the opening. The wear on the seal tends to occur at the shaft when the mounting hole becomes eccentric thus allowing the seal to slip slightly to one side during calibration.

Replacement of the sealing disc is simple if a little fiddly and the results are remarkable.

Failing that I think you might be looking at replacing one of the main mixing valves - expensive!

Hope this may be of use.

Brian
_________________________
**** Still grumbling in the Pennines ****

Top
Page 1 of 2 1 2 >



Moderator:  DaveC in Oz, KM, RoJo 

Sponsors

Press Releases
Who's Online
22 registered (vins, Roger, Gary Smith, Rob.Rsch, Inter07, Paul Robbins, RoJo, John Stewart, Huw, noah, Electric Blobby, Angus, Neil Porter, David Mulvey, FrankD, Lee S, DaveC in Oz, Paul Owens, Geoff Hannis, hieu, Laiq, 1 invisible), 250 Guests and 8 Spiders online.
Key: Admin, Global Mod, Mod
Events
17th National Biomedical & Clinical Engineering Co
Welch Allyn - 12th Annual "Completing the Picture
June
Su M Tu W Th F Sa
1
2 3 4 5 6 7 8
9 10 11 12 13 14 15
16 17 18 19 20 21 22
23 24 25 26 27 28 29
30
Newest Members
vinu, MacD33, bindu, leonardkumbi, Richard Medical
7272 Registered Users
Featured Member
mkm
mkm
Registered: 08/01/08
Posts: 30
Today's Birthdays
anu
(Views)Popular Topics
Jokes 2 (son of Jokes!!!) 1160867
Classic Computers 398669
Magic Kingdom Biomed Expats 348794
The VRCT Website 321328
VRCT - Who's Re-Newed? 209914
Trainee EBME Technician Salary? 207979
VRCT renewal 190294
Jokes!! 185923
Sharing the info. 176813
DIRTY equipment 169550