#29492 - 18/04/08 10:48 AM How do you test defibrillators?
DanKel Offline
Scholar

Registered: 26/02/08
Posts: 74
Loc: Manchester, UK
All our equipment is currently tested elsewhere but I would be grateful if someone could tell me how easy is it to test defibrillators & what's involved in doing so?
_________________________
The road to enlightenment is full of confusion, missteps and soggy fries \:\)

Top
Covidien : 01329 224000
Covidien
#29493 - 18/04/08 10:55 AM Re: How do you test defibrillators? [Re: DanKel]
Geoff Hannis Online   content
Hero

Registered: 12/02/04
Posts: 4591
Loc: the path less trodden
It's easy enough when:-

1) You have a decent defib tester.
2) You know what you're doing!

(best leave it to qualified people, Lel) \:\)

PS: lousy spelling again, Mate.

Top
#29494 - 18/04/08 11:03 AM Re: How do you test defibrillators? [Re: Geoff Hannis]
DanKel Offline
Scholar

Registered: 26/02/08
Posts: 74
Loc: Manchester, UK
Spelling looks right to me? \:\)
_________________________
The road to enlightenment is full of confusion, missteps and soggy fries \:\)

Top
#29495 - 18/04/08 11:12 AM Re: How do you test defibrillators? [Re: DanKel]
Geoff Hannis Online   content
Hero

Registered: 12/02/04
Posts: 4591
Loc: the path less trodden
Nicely corrected, then, Lel. But while I'm on, technical (computer) people don't like using ampersands, as they are used to convey other meanings in (for example) computer code! \:\)

(see how much you're learning here?)


Edited by Geoff Hannis (18/04/08 11:13 AM)
Edit Reason: &

Top
#29496 - 18/04/08 11:19 AM Re: How do you test defibrillators? [Re: Geoff Hannis]
DanKel Offline
Scholar

Registered: 26/02/08
Posts: 74
Loc: Manchester, UK
Consider me a sponge, here to absorb information \:\)

Knowledge is your friend \:\)
_________________________
The road to enlightenment is full of confusion, missteps and soggy fries \:\)

Top
#29498 - 18/04/08 11:25 AM Re: How do you test defibrillators? [Re: DanKel]
Geoff Hannis Online   content
Hero

Registered: 12/02/04
Posts: 4591
Loc: the path less trodden
I recall a good old buddy of mine (he was an Egyptian, not that that really matters) saying "I have no friends, my work is my friend". And I could (and still can) relate to that!

Remember too "Arbeit macht frei"!

But, better yet:-

"I keep six honest serving men (they taught me all I knew); their names are What and Why and When and How and Where and Who."
- Rudyard Kipling \:\)


Edited by Geoff Hannis (18/04/08 11:28 AM)
Edit Reason: It's a blue day.

Top
#29533 - 18/04/08 02:46 PM Re: How do you test defibrillators? [Re: Geoff Hannis]
Alan M Offline
Savant

Registered: 07/08/05
Posts: 114
Loc: Chesterfield, Derbyshire
Buy someting to test them with.

We have recently purchased a device from Fluke, although we used to have an older device for testing our Defibs. We used to have S&W's, however recently have changed to Philips Heartstart etc.

What defibrillators do you have?
_________________________
Rock the boat.... Get yer coat!

Top
#29550 - 18/04/08 04:52 PM Re: How do you test defibrillators? [Re: Alan M]
DanKel Offline
Scholar

Registered: 26/02/08
Posts: 74
Loc: Manchester, UK
Lifepak 9's, Lifepak 9P's & Lifepak 12's. wasnt sure if there was a purpose built tester or simply using a meter was an option? i assumed it would not be that simple or wise?
_________________________
The road to enlightenment is full of confusion, missteps and soggy fries \:\)

Top
#29555 - 18/04/08 06:24 PM Re: How do you test defibrillators? [Re: DanKel]
Alan M Offline
Savant

Registered: 07/08/05
Posts: 114
Loc: Chesterfield, Derbyshire
Contact Fluke and ask if there tester is suitable for these (If you have the money!).

Most of your generic defib testers are multi-manufacturer so you can probably pick one up that will test them. It's not as if they are an oddball piece of kit.

Regards

Alan M
_________________________
Rock the boat.... Get yer coat!

Top
#29560 - 18/04/08 06:57 PM Re: How do you test defibrillators? [Re: Alan M]
Geoff Hannis Online   content
Hero

Registered: 12/02/04
Posts: 4591
Loc: the path less trodden
All Lel needs (in my opinion) is a simple go/no-go tester of the old CR type (remember those?), Alan. Less there be no misunderstanding, he is in a warehouse environment (not a hospital, etc.). As I have already said, my advice is that he should leave such things to the "experts" (even though that means he may have to shell out a couple of bob)! \:\)


Edited by Geoff Hannis (18/04/08 06:58 PM)
Edit Reason: A little bit of knowledge ...

Top
Page 1 of 2 1 2 >


Who's Online
11 Registered (Apollo11, Chris Watts, Daddy58, Ed SWM, Eddie, Geoff Hannis, Huw, Kanku, Mark Radbourne, Rob1234, sdah), and 106 guests online.
Newest Members
shihab, Daddy58, Taffy2, Chris Chovaz, redSG
4116 Registered Users
Jobs
EBME Manager - Harlow
-- 6 January 2009

Medical Equipment Library Manager
-- 15 December 2008

Senior Principal Healthcare Scientist
-- 10 December 2008

Senior Medical Engineer Band 5 / 6
-- 8 December 2008

Medical Equipment Repair Technicians
-- 27 November 2008

Biomedical Electronics Team Manager
-- 10 November 2008

Newsletter

Subscribe
Unsubscribe