We have had three occurrences of air being spotted down steam of the pump (Graseby 500) towards the end of blood infusions recently. The pumps are being used by experienced, trained staff who have had competency based training on the product.
The staff ensured that the air was not injected however their confidence has been knocked.
The bubbles have been between 3 and 12mm long in a standard blood set (not I believe enough to be dangerous but still worrying) with one infusion having 4 separate bubbles.
The giving sets on close inspection show many small bubbles above the pump typically smaller than 0.5mm diameter and the same below the pump but up to 2mm diameter; these bubbles are not apparent when the infusion starts. Watching the small bubbles during an infusion showed them slowly getting larger but not moving down the set. (sizes of bubbles at end of infusion)
The infusions are typically of around 250ml infused over two hours twenty minutes, so infusion rates around 100ml an hour.
The incidents have occurred with different pumps, bloods and batches of giving set.
I have tested these pumps to Graseby’s Checklist and found no problems; they have also been tested by a third party; I could not get air to go past the first blister and its air detection system, Graseby are also investigating this.
The blood is coming down immediately after matching so will still be relatively cold.
Does anyone know what is causing this problem?
What is a safe level of air to be injected into a vein? (Yes I have read the Infusion Pump Technical Article).