#39322 - 03/07/09 06:36 AM
Re: Renal Dialysis water treatment plant
[Re: Neil Porter]
|
Dreamer
Registered: 21/04/05
Posts: 24
Loc: Eldoret,Kenya
|
hello....no one is telling me if the two are compatible or not Thanks
|
|
Top
|
|
|
|
#39325 - 03/07/09 09:01 AM
Re: Renal Dialysis water treatment plant
[Re: mtalam]
|
Super Hero
Registered: 12/02/04
Posts: 10021
Loc: the path less trodden
|
Without standing there alongside you it would be foolhardy of anyone here to make a declaration quite that firm, would it not? However ... unless there is some sort of monitoring feedback loop from the (dialysis) machines, and/or some sort of controlling (of the water treatment) from there ... what could be incompatible? The connections at the (dialysis) machines' end? 
|
|
Top
|
|
|
|
#39331 - 03/07/09 09:35 AM
Re: Renal Dialysis water treatment plant
[Re: Neil Porter]
|
Super Hero
Registered: 12/02/04
Posts: 10021
Loc: the path less trodden
|
The over-riding consideration is water quality, naturally. All the rest, capacity, pressure, flow, volumes, "ring main", loops or branches ... is after all, just straight forward plumbing (er, engineering)!  What about relative locations, mtalam? You haven't mentioned that. Next door (yes, good), or (as I have known) a kilometer away, across a road (bad). Or upstairs (OK ... gravity), or down in the basement (not so good ... more pumping required).
|
|
Top
|
|
|
|
#39337 - 03/07/09 10:19 AM
Re: Renal Dialysis water treatment plant
[Re: Neil Porter]
|
Super Hero
Registered: 12/02/04
Posts: 10021
Loc: the path less trodden
|
Yes Mate. It all has to be correctly engineered. Don't forget, though, that we may not be talking about "ideal conditions", here. But (yet again), you've afforded me the opportunity to remind everyone that:- "An engineer is someone who can do for a Riyal (insert your favourite currency here, as appropriate) what any darned fool can do for ten"! 
|
|
Top
|
|
|
|
#39341 - 03/07/09 10:44 AM
Re: Renal Dialysis water treatment plant
[Re: Geoff Hannis]
|
Dreamer
Registered: 21/04/05
Posts: 24
Loc: Eldoret,Kenya
|
The location is just next door though we are avoiding putting up a palnt house since that will mean extra money.We also intend to have a permeate ring to serve the 10 units and lastly we dont intend to store the permeate-this is usually the cause of infection due to micro organism growth
|
|
Top
|
|
|
|
|
|
24 registered (Sandy King, mugtree, Huw, zilog, zedrok, SUZUKI, Chris Jones, Darth Welder, jackwu, Stephen Don, BobBanks, obiwan, lesh, RS, Hulk, walaa, Gordovan, Bala Vignesh, John Stewart, Armenag, ipetev, Rob1234, RoJo, 1 invisible),
372
Guests and
12
Spiders online. |
|
Key:
Admin,
Global Mod,
Mod
|
|
|
|
|
|
1
|
2
|
3
|
4
|
|
5
|
6
|
7
|
8
|
9
|
10
|
11
|
|
12
|
13
|
14
|
15
|
16
|
17
|
18
|
|
19
|
20
|
21
|
22
|
23
|
24
|
25
|
|
26
|
27
|
28
|
29
|
|
|
|
|
|