#40853 - 16/09/09 02:40 AM Drager GA primus
mammad Offline
Scholar

Registered: 05/11/07
Posts: 74
Loc: PCMC, Malaysia
Hye friend, I have one unit of Drager Primus. Initiatly when run the self test, the valves/ventilator test failed. I check the ventilator and change the whole ventilator systems with the effective one. Then I run the self test, it shows leak (man.spot) failed and the reading is too high, 800 plus. And also the breathing system also failed.

I checked all the breathing circuit, still cannot find the problem... can anybody help me or anybody experienced the same problem.. thanks ya...

Top
#41000 - 28/09/09 07:06 AM Re: Drager GA primus [Re: mammad]
Diego de Jesus Gonzalez Beltran Offline
Scholar

Registered: 03/05/02
Posts: 58
Loc: Guadalajara Jalisco Mexico
Do you still need help?

let me check I have a Friend that knows about drager machines
_________________________
Diego Gonzalez
Service Engineer
GAS Latam Mexico SA de CV

Top
#41015 - 28/09/09 02:13 PM Re: Drager GA primus [Re: Diego de Jesus Gonzalez Beltran]
Diego de Jesus Gonzalez Beltran Offline
Scholar

Registered: 03/05/02
Posts: 58
Loc: Guadalajara Jalisco Mexico
Hi:

my friend toll me that the leak must not be in the Breathing circuit
it could be in the Ataching system betwen the patient module and the machine (check the seals betwen the systems)

and may be the other problem is a leak with the APL valve


I hoppe this help

Have a great day
_________________________
Diego Gonzalez
Service Engineer
GAS Latam Mexico SA de CV

Top
#41071 - 02/10/09 04:59 AM Re: Drager GA primus [Re: Diego de Jesus Gonzalez Beltran]
mammad Offline
Scholar

Registered: 05/11/07
Posts: 74
Loc: PCMC, Malaysia
Thanks Diego.. So how to check the leaking at APL valve?

Top


Who's Online
9 registered (shijith, Neil Porter, umish, mhmtech, nguyen phong, BioAwaken, Kiwi Phil, Dominic14, june), 89 Guests and 35 Spiders online.
Key: Admin, Global Mod, Mod
Newest Members
Teflon, Anjum, lochardil, yazar71, shihab.s
4816 Registered Users
Jobs
Medical Technical Officer
-- 8 March 2010

Service Engineer Required
-- 2 March 2010

Sales Representatives
-- 2 March 2010

Sister / Charge Nurse (Equipment Library)
-- 24 February 2010

Senior Technical Officer (Electronics)
-- 6 February 2010

Clinical Technical Assistant Manager
-- 2 February 2010

Medical Technical Officer
-- 29 January 2010

Dynamic PFI Biomedical Manager required.
-- 21 January 2010

Radiotherapy Equipment Technician
-- 15 January 2010

Newsletter

Subscribe
Unsubscribe