#39315 - 02/07/09 06:10 PM
Re: Renal Dialysis water treatment plant
[Re: Geoff Hannis]
|
Hero
Registered: 23/02/09
Posts: 1499
Loc: Jeddah, Saudi Arabia
|
Thats how I lost the tender for the water treatment, it went to gambro, as did Ben soon after, good tech though
Edited by Neil Porter (02/07/09 06:11 PM)
_________________________
Stress is for other people
|
|
Top
|
|
|
|
#39320 - 02/07/09 08:06 PM
Re: Renal Dialysis water treatment plant
[Re: Geoff Hannis]
|
Hero
Registered: 23/02/09
Posts: 1499
Loc: Jeddah, Saudi Arabia
|
Open units, easy maintenance access
_________________________
Stress is for other people
|
|
Top
|
|
|
|
#39322 - 03/07/09 06:36 AM
Re: Renal Dialysis water treatment plant
[Re: Neil Porter]
|
Dreamer
Registered: 21/04/05
Posts: 24
Loc: Eldoret,Kenya
|
hello....no one is telling me if the two are compatible or not Thanks
|
|
Top
|
|
|
|
#39325 - 03/07/09 09:01 AM
Re: Renal Dialysis water treatment plant
[Re: mtalam]
|
Super Hero
Registered: 12/02/04
Posts: 10297
Loc: the path less trodden
|
Without standing there alongside you it would be foolhardy of anyone here to make a declaration quite that firm, would it not? However ... unless there is some sort of monitoring feedback loop from the (dialysis) machines, and/or some sort of controlling (of the water treatment) from there ... what could be incompatible? The connections at the (dialysis) machines' end? 
|
|
Top
|
|
|
|
#39328 - 03/07/09 09:13 AM
Re: Renal Dialysis water treatment plant
[Re: Geoff Hannis]
|
Hero
Registered: 23/02/09
Posts: 1499
Loc: Jeddah, Saudi Arabia
|
two things you have to worry about 1) the type of connectors between the supply line and the machines. 2) that the supply line can handle an increase in volume. The water quality, pressure and quantity from from DWA, Gambro, Fresenius and any other company should be acceptable for any Dialysis machine. If you have a small loop ensure that it can handle the inreased volume, the pressure will remain the same, I had to split a single loop into 2 parallel loops in order to cope with upgrading the unit from 15 machined to 35
_________________________
Stress is for other people
|
|
Top
|
|
|
|
#39331 - 03/07/09 09:35 AM
Re: Renal Dialysis water treatment plant
[Re: Neil Porter]
|
Super Hero
Registered: 12/02/04
Posts: 10297
Loc: the path less trodden
|
The over-riding consideration is water quality, naturally. All the rest, capacity, pressure, flow, volumes, "ring main", loops or branches ... is after all, just straight forward plumbing (er, engineering)!  What about relative locations, mtalam? You haven't mentioned that. Next door (yes, good), or (as I have known) a kilometer away, across a road (bad). Or upstairs (OK ... gravity), or down in the basement (not so good ... more pumping required).
|
|
Top
|
|
|
|
#39335 - 03/07/09 10:08 AM
Re: Renal Dialysis water treatment plant
[Re: Geoff Hannis]
|
Hero
Registered: 23/02/09
Posts: 1499
Loc: Jeddah, Saudi Arabia
|
NO BRANCHES. the distance between the loop and the connector must not exceed 2 1/2 inches, gravity does not come into play as you will be using impeller pumps. There should also be no storage of product water, There is another consideration though and that is the chiller, can the chiller handle the extra capacity as you need to keep the water below 30 degrees.
_________________________
Stress is for other people
|
|
Top
|
|
|
|
#39337 - 03/07/09 10:19 AM
Re: Renal Dialysis water treatment plant
[Re: Neil Porter]
|
Super Hero
Registered: 12/02/04
Posts: 10297
Loc: the path less trodden
|
Yes Mate. It all has to be correctly engineered. Don't forget, though, that we may not be talking about "ideal conditions", here. But (yet again), you've afforded me the opportunity to remind everyone that:- "An engineer is someone who can do for a Riyal (insert your favourite currency here, as appropriate) what any darned fool can do for ten"! 
|
|
Top
|
|
|
|
#39341 - 03/07/09 10:44 AM
Re: Renal Dialysis water treatment plant
[Re: Geoff Hannis]
|
Dreamer
Registered: 21/04/05
Posts: 24
Loc: Eldoret,Kenya
|
The location is just next door though we are avoiding putting up a palnt house since that will mean extra money.We also intend to have a permeate ring to serve the 10 units and lastly we dont intend to store the permeate-this is usually the cause of infection due to micro organism growth
|
|
Top
|
|
|
|
|
|
12 registered (Laiq, infinix, GeorgeK, Lee S, Huw, ferhat__aydin, bosch, Geoff Hannis, Angus, NickM, DaveC in Oz, Neil Porter),
281
Guests and
8
Spiders online. |
|
Key:
Admin,
Global Mod,
Mod
|
|
|
|
|
1
|
2
|
3
|
4
|
5
|
|
6
|
7
|
8
|
9
|
10
|
11
|
12
|
|
13
|
14
|
15
|
16
|
17
|
18
|
19
|
|
20
|
21
|
22
|
23
|
24
|
25
|
26
|
|
27
|
28
|
29
|
30
|
31
|
|
|
|
|