#44559 - 03/03/10 03:18 AM Modify a RF302 ESU Analyzer for Hyfrecator checks
SCANNER Offline
Newbie

Registered: 13/12/08
Posts: 2
Loc: Alabama , USA

Greetings,

Has anyone ever modified a Bio-Tek RF302 Electrosurgery Analyzer to test Hyfrecators ? If so is there a schematic diagram to use to do the modification. When I use my RF302 to test Hyfrecators I use the higher resistance load setting for the Hyfrecator but then I lose the my scale sensitivity (500 Ohms / 500W) and the max output of the Hyfrecator is only 35W. I know I've seen photos of what looks like a RF302 ESU analyzers with added input for Hyfrecators. Can anyone help. Please drop me an email at spectrum_dynamics@yahoo.com. or post a reply.

Thanks

Randy Reynolds
Biomedical Tech. USA

Top
#44565 - 03/03/10 11:56 AM Re: Modify a RF302 ESU Analyzer for Hyfrecator checks [Re: SCANNER]
Sean Fearon Offline
Mentor

Registered: 10/02/03
Posts: 166
Loc: CMFT Manchester
Hello Randy
You can get more accurate Hyfrecator output readings with a Pearson current monitor and a True RMS meter. You already know the RF302 impedance (nominal 500 ohms), square the current measured through the "Doughnut" and this will provide the power.
I believe Pearson are located in Palo Alto CA, and the True RMS meter needs to have a suitable bandwidth.
I used to calibrate the RF302's and because of the hot wire thermocouple response and the quoted accuracy +/- 5% of scale, (worst case scenario at that range +/- 25 Watts error) you are not going to get accurate low power readings of 30 to 40 W on a 500 watts/ohms selected scale, no matter what modification's you do using that technology.

Regards
Sean

Top



Moderator:  DaveC in Oz, KM, RoJo 
Sponsors
Press Releases
EBME Seminar 2012
19 January 2012
Who's Online
20 registered (SUZUKI, Chris Jones, Darth Welder, jackwu, Stephen Don, BobBanks, obiwan, lesh, Huw, RS, Hulk, walaa, Gordovan, Bala Vignesh, John Stewart, Armenag, ipetev, Rob1234, RoJo, 1 invisible), 368 Guests and 10 Spiders online.
Key: Admin, Global Mod, Mod
February
Su M Tu W Th F Sa
1 2 3 4
5 6 7 8 9 10 11
12 13 14 15 16 17 18
19 20 21 22 23 24 25
26 27 28 29