#44601 - 05/03/10 10:12 AM
Graseby 3100
|
Novice
Registered: 21/07/09
Posts: 15
Loc: Barreiro , Portugal
|
I was reading an older post here that had been placed here about a fault 86 and already did what was indicated to replace the 10v 100uf capacitor C3, but the error continues. Does anybody know other possible causes for this fault?
Regards
|
|
Top
|
|
|
|
#44646 - 08/03/10 04:29 PM
Re: Graseby 3100
[Re: HARI_CHINGAPURAM]
|
Novice
Registered: 21/07/09
Posts: 15
Loc: Barreiro , Portugal
|
the error was mine i was changing the wrong capacitor . Problem solved .
Thanks Anyway
Regards
|
|
Top
|
|
|
|
#44679 - 09/03/10 09:03 AM
Re: Graseby 3100
[Re: Geoff Hannis]
|
Novice
Registered: 21/07/09
Posts: 15
Loc: Barreiro , Portugal
|
yes it was the c3 but i was changing the c3 from the power board. Then i see tha it was from the PCB Main that needs to be changed. Regards 
|
|
Top
|
|
|
|
#51382 - 05/01/11 12:40 PM
Re: Graseby 3100
[Re: Serroni]
|
Visionary
Registered: 17/04/10
Posts: 32
Loc: Mumbai-Maharastra-India
|
Hi guys,
I have one Grasbey 3100, with Fault 91 on screen, anyone can help with the service manual or suggestions most welcome
Regards Nimish
|
|
Top
|
|
|
|
#51385 - 05/01/11 01:08 PM
Re: Graseby 3100
[Re: Nimish.P]
|
Super Hero
Registered: 12/02/04
Posts: 10021
Loc: the path less trodden
|
Code 91 is a "keyboard fault".  The manual suggests "replace keyboard"! But first check for a stuck key "button", fluid "ingress", bad connection(s) ... and all that usual stuff.
|
|
Top
|
|
|
|
#51401 - 06/01/11 06:49 AM
Re: Graseby 3100
[Re: Serroni]
|
Visionary
Registered: 17/04/10
Posts: 32
Loc: Mumbai-Maharastra-India
|
Geoff and Ariel,
Thanks for msg, will check and let inform you
|
|
Top
|
|
|
|
#51421 - 07/01/11 11:35 AM
Re: Graseby 3100
[Re: Serroni]
|
Visionary
Registered: 17/04/10
Posts: 32
Loc: Mumbai-Maharastra-India
|
Yes it does solve, i wonder why couldnt they display " check keypad" instead of Fault 91.... anyways thanks guys need anything just send email....
|
|
Top
|
|
|
|
|
|
20 registered (SUZUKI, Chris Jones, Darth Welder, jackwu, Stephen Don, BobBanks, obiwan, lesh, Huw, RS, Hulk, walaa, Gordovan, Bala Vignesh, John Stewart, Armenag, ipetev, Rob1234, RoJo, 1 invisible),
368
Guests and
10
Spiders online. |
|
Key:
Admin,
Global Mod,
Mod
|
|
|
|
|
|
1
|
2
|
3
|
4
|
|
5
|
6
|
7
|
8
|
9
|
10
|
11
|
|
12
|
13
|
14
|
15
|
16
|
17
|
18
|
|
19
|
20
|
21
|
22
|
23
|
24
|
25
|
|
26
|
27
|
28
|
29
|
|
|
|
|
|