#44896 - 21/03/10 09:21 PM Fresenius Base Intensive Intermittent alarm
madmedic Offline
Newbie

Registered: 19/03/10
Posts: 1
Loc: hull
Hi

Can anyone help with a fresenius base intensive alarm issue, that causes the screen to frezze and has a continous alarm, in order to stop the alarm the battery has to be disconnected.
When all connected back up and switched back on , the unit worked fine, just wondering why it happened and is there a fix for it.

Thanks for your help.
Medic.

Top
#44898 - 22/03/10 08:34 AM Re: Fresenius Base Intensive Intermittent alarm [Re: madmedic]
John Harte Offline
.

Registered: 11/03/05
Posts: 35
Loc: ST Vincents University Hospita...
Medic
We have 32 Base intensives with this same problem Fressenius have being trying to find the reason for nearly 2 yrs with no joy .what version of software are the bases on and is this problem happening in any mode .Fresenius have had a number of units back in their work shop and say they cannot get the fault to happen at last somebody else with the same fault I taught our ICU was haunted as fresenius claim this dose not happen else where .Conntact me if you like and we can compare
00353 12214000 and ask for bleep 464

Top
#45039 - 29/03/10 07:40 AM Re: Fresenius Base Intensive Intermittent alarm [Re: John Harte]
John Harte Offline
.

Registered: 11/03/05
Posts: 35
Loc: ST Vincents University Hospita...
Fresenius e-mailed me on friday saying they have new sofeware update (3.2p ) which repairs this false alarm

Top
#49133 - 20/09/10 12:10 PM Re: Fresenius Base Intensive Intermittent alarm [Re: madmedic]
Paul Kelly Offline
Newbie

Registered: 05/05/10
Posts: 9
Loc: Runcorn, Cheshire
Hi All
If you have any problems with any Infusion Devices please contact me on 01928 533455.

In regards to the Base Intensive alarm, the Base carrys out a self check on switch on. If the Base detects any irregularities it will continuously alarm. To prevent the alarm you must hold down the on/off switch for around 15 seconds which will turn the Base off. Then switch the Base back on and ot will be fine!

Top
#49137 - 20/09/10 02:47 PM Re: Fresenius Base Intensive Intermittent alarm [Re: Paul Kelly]
John Harte Offline
.

Registered: 11/03/05
Posts: 35
Loc: ST Vincents University Hospita...
Paul
what you said will normally fix that alarm proplem but Fresenius have said they have a problem with their software they tried version 3.2p but it was faulty they are due to release a fix soon

Top
#49371 - 30/09/10 02:54 PM Re: Fresenius Base Intensive Intermittent alarm [Re: madmedic]
Paul Kelly Offline
Newbie

Registered: 05/05/10
Posts: 9
Loc: Runcorn, Cheshire
Hi John
The software release is due any week version 3.2Q.

Top