Hi
Im a newcomer. My Alaris Ivac P2000 has a error: Erf confused. I dont have service manual. Can everyone help me repair.
Thank.

Top
#46797 - 09/06/10 07:22 AM Re: Alaris IVAC P2000 [Re: nhockiller]
Geoff Hannis Offline
Super Hero

Registered: 12/02/04
Posts: 10298
Loc: the path less trodden

Welcome to the forum. smile

The "Er" error codes all indicate what the manual calls "an internal malfunction":-

Er1 : Motor speed too fast.
Er2 : Motor speed too slow.
Er3 : Too many encoder pulses for rate set.
Er4 : Zero encoder pulses.
Er5 : RAM data invalid.
Er6 : Watchdog failure (too slow).
Er7 : Watchdog failure (too fast).
Er8 : Invalid syringe calibration data.
Er9 : Motor malfunction.

See what you get at the Self-Test routine:-

1) Switch off
2) Hold down Start button and switch On/Off switch to On
3) (display will show - - -)
4) Enter 1 2 3 using the rate buttons
5) Press Stop to commence Self-Test sequence

The pump should now proceed through the tests. Press Stop to move on the next test, or Start at any time to exit the Self-Tests routines. There should be twenty or so tests.

I am guessing that you are reading Er8 as ErF ... so pay particular attention to Test No.3 (Syringe Calibration Check). This will show a "Sy" code (Sy0 to Sy8) according to the type (manufacturer etc.) of syringe you are using (or plan to use).

Check out Test No.4 as well. This is the Default Syringe Size. "d00" should be a safe bet.

Er8 may be due to RAM not retaining data when the pump is switched off. Take a look at battery B1 on the Control PCB.

Let us know how you get on. smile

Top
#46810 - 09/06/10 09:48 AM Re: Alaris IVAC P2000 [Re: Geoff Hannis]
Eagle Offline
Adept

Registered: 04/03/03
Posts: 78
Loc: South West
As in the training notes:

ErF - "Internal microcontroller RAM or external RAM error" caused by excessive RF interference or problems with IC1 or IC18 on control board.

Hope this helps.

Top
#46814 - 09/06/10 11:13 AM Re: Alaris IVAC P2000 [Re: Eagle]
Geoff Hannis Offline
Super Hero

Registered: 12/02/04
Posts: 10298
Loc: the path less trodden

Interesting! I wonder if the B1 battery is (still) the culprit! smile

I don't suppose you have all the other "ErA" (alphas) codes, and what they mean?

Top
#46822 - 09/06/10 03:31 PM Re: Alaris IVAC P2000 [Re: Geoff Hannis]
Eagle Offline
Adept

Registered: 04/03/03
Posts: 78
Loc: South West
ErA: Voltage reference fault - IC0 on power board, or power board.

ErB: Strain gauge current fault - Bonded beam or control board.

ErC: Motor drive transistor fault (off when should be on) - Check motor wires not catching, TR1 & 2 on control board, TR4 on power board.

ErD: Motor drive transistor fault (on when should be off) - TR1 & 2 on control board, TR4 on power board.

ErE: Plunger detect error - Flex cct or plunger stuck.

Err: Pumping pressure offset error - Bonded beam or IC2 on control board.

ErY: Motor data error - Control board.

ErL: CRC error - Control board.

ErO: Microprocessor error - Control board.

ErN: Watchdog window cct error - Control board.

Finally

Eric - my brother's name!

Top
#46826 - 09/06/10 04:23 PM Re: Alaris IVAC P2000 [Re: Eagle]
Geoff Hannis Offline
Super Hero

Registered: 12/02/04
Posts: 10298
Loc: the path less trodden

Thanks for that (new information, for me at any rate)! smile

Top
#46868 - 10/06/10 02:29 PM Re: Alaris IVAC P2000 [Re: Geoff Hannis]
nhockiller Offline
Newbie

Registered: 08/06/10
Posts: 3
Loc: Ho Chi Minh City
Thank you very much. Let me check IC1 and IC18 laugh .


Edited by nhockiller (10/06/10 02:53 PM)

Top
#47006 - 15/06/10 10:06 AM Re: Alaris IVAC P2000 [Re: nhockiller]
nhockiller Offline
Newbie

Registered: 08/06/10
Posts: 3
Loc: Ho Chi Minh City
Vcc of IC1 and IC18 equal 5V, but V battery of IC18 does not equal 3V. Do anybody have technical service. If you have, please send me: nhoc1met84@gmail.com

Top
#47007 - 15/06/10 10:13 AM Re: Alaris IVAC P2000 [Re: nhockiller]
Geoff Hannis Offline
Super Hero

Registered: 12/02/04
Posts: 10298
Loc: the path less trodden

We only have the manual in hard copy, unfortunately. frown

Meanwhile, how is the RAM back-up battery B1?

Top



Moderator:  DaveC in Oz, KM, RoJo 
Sponsors
Press Releases
Who's Online
9 registered (Mofe, GeorgeK, Paul Tregurtha, mek.sah, DinhVu, ani, DaveC in Oz, Huw, pat), 299 Guests and 10 Spiders online.
Key: Admin, Global Mod, Mod
May
Su M Tu W Th F Sa
1 2 3 4 5
6 7 8 9 10 11 12
13 14 15 16 17 18 19
20 21 22 23 24 25 26
27 28 29 30 31
Events