#48039 - 07/08/10 10:37 AM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Savant
Registered: 18/12/09
Posts: 114
Loc: Brisbane, Australia
|
I've got the service manual for the 40L and it states to check the gas supply. Hope that helps
|
|
Top
|
|
|
|
#48084 - 09/08/10 10:41 AM
Re: Stryker 30L highflow insufflator error 04
[Re: Pavel]
|
Savant
Registered: 18/12/09
Posts: 114
Loc: Brisbane, Australia
|
Hi Guys, I work for Stryker as a Field Service Engineer, unforunately for you guys all the manuals I have are in electronic form in a program we use, I could probably get extracts if required, but if you post any problems you have I will try to help or atleast point you in the right direction.
|
|
Top
|
|
|
|
#48720 - 02/09/10 12:37 PM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Savant
Registered: 18/12/09
Posts: 114
Loc: Brisbane, Australia
|
Like I stated in a previous post you need to check the gas supply, its possibly the internal regulator requires calibration and or repair/replacement.
"4.6.1 Checking Gas Supply Connection/Gas Inlet Filter (See Fig. 4-6-1-1) The gas inlet filter (2) may be plugged or soiled due to Soiling unclean gas or gas supply residues. This type of soiling, however, cannot be detected externally. Heavy soiling is especially noticeable during higher flow rates and results in the actual gas flow falling short of the nominal gas flow during active insufflation. If cleaning the surface of the gas inlet filter does not yield any improvement, the gas supply connection (1) has to be replaced completely"
Keep in mind this is for the 40L
|
|
Top
|
|
|
|
#48722 - 02/09/10 12:47 PM
Re: Stryker 30L highflow insufflator error 04
[Re: NickM]
|
Technologist
Registered: 17/06/10
Posts: 47
Loc: tamilnadu,india
|
|
|
Top
|
|
|
|
#48723 - 02/09/10 12:55 PM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Savant
Registered: 18/12/09
Posts: 114
Loc: Brisbane, Australia
|
Let me know if you need the part numbers for a replacement regulator or gas inlet filter assy
|
|
Top
|
|
|
|
|
|
9 registered (San Garwe, Rob.Rsch, Huw, Geoff Hannis, Marcelli, Neil Porter, tomiteus, JackFlash, BK),
354
Guests and
14
Spiders online. |
|
Key:
Admin,
Global Mod,
Mod
|
|
|
|
|
|
1
|
2
|
3
|
4
|
|
5
|
6
|
7
|
8
|
9
|
10
|
11
|
|
12
|
13
|
14
|
15
|
16
|
17
|
18
|
|
19
|
20
|
21
|
22
|
23
|
24
|
25
|
|
26
|
27
|
28
|
29
|
30
|
31
|
|
|
Registered: 22/01/12
Posts: 3
|
|
|