#51476 - 11/01/11 10:39 AM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
NickM Offline
Savant

Registered: 18/12/09
Posts: 114
Loc: Brisbane, Australia
Hi Inzunza, unfortunately i've been away from work today due to flooding in Brisbane, Queensland, Australia. Closing down our office.

If I get a chance I will forward through the calibration procedure via email in the next few days.

Basically the things you will need is a full bottle of co2 gas, 2 x 50ml Syringes, and a co2 pressure gauge.

The calibration procedure is fairly straight forward otherwise.


Edited by NickM (11/01/11 10:41 AM)

Top
#51494 - 12/01/11 06:26 AM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
Inzunza Offline
Newbie

Registered: 02/01/11
Posts: 4
Loc: Mexico
Thanks NickM not problem , I'll wait until you have time for it, hope you do not forget it smile.

Top
#52367 - 22/02/11 08:22 AM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
Dps Offline
Newbie

Registered: 17/02/11
Posts: 3
Loc: Punjab
Dear Nick I also would need calibration procedure.please send at my email also dcads_mrkt@yahoo.com

Top
#55239 - 11/03/11 08:14 PM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
remocnab Offline
Newbie

Registered: 10/03/11
Posts: 1
Loc: mexico michoacan
Hi iīm have the same problem with contamination on the display, can you help mi to solve it? i revove a little dusts in the outlet tube, but i donīt know how reset the unit, the unit is a 40 stryker, thanks for your help..

Top
#55559 - 28/03/11 03:44 PM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
TedF Offline
Newbie

Registered: 10/02/11
Posts: 2
Loc: Dallas, Texas USA
Hi, I have two Stryker 40L insufflators that I have brought back from a mission hospital in rural Guatemala. They both have an error message without a number. The error message is "SENSOR DEFECT." Any help would be appreciated. Does anyone have a service manual??

Top
#55566 - 29/03/11 07:58 AM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
NickM Offline
Savant

Registered: 18/12/09
Posts: 114
Loc: Brisbane, Australia
Hi Ted this is due to contaminated gas you will need to replace the fluid sensor and LPU (Low Pressure Unit)

Part Number: 0105199130
Description: COMPLETE PNUEMATICS UNIT LPU

and the fluid sensor

Part Number: 105190933
Description: FLUID SENSOR

As this is for a mission hospital Stryker may be able to sort out some sort of donation repair, contact your local stryker rep if not let me know and I will see what I can do.

All the best

Nick (Field Service Engineer - Stryker South Pacific)

Top
#55627 - 01/04/11 01:41 AM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
TedF Offline
Newbie

Registered: 10/02/11
Posts: 2
Loc: Dallas, Texas USA
thanks Nick,
The Stryker rep here does not seem to know how or if parts can be ordered. Do you know what these parts would cost or how I can get them?
Thanks,
Ted

Top
#55669 - 02/04/11 10:26 PM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
tahir mehmood Offline
Visionary

Registered: 01/06/09
Posts: 34
Loc: pak
dear nick
i have stryker 20 liter and older 40liter with ereror 21 can you send me the copy of service mannula
I had the coppy of stryker newer 40 L high flow and misplaced it can you help
and send me a copy of that
the eror is faulty valve
_________________________
Dr tahir mehmood
138 m d.h.a
lahore pakistan

Top
#56623 - 03/06/11 08:04 PM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
tahir mehmood Offline
Visionary

Registered: 01/06/09
Posts: 34
Loc: pak
dear nick hi
i have two of the 40l insufflator with calibration error on one and regulator error on the other can you help please
_________________________
Dr tahir mehmood
138 m d.h.a
lahore pakistan

Top
#57005 - 28/06/11 01:35 PM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
NotwenCardozo Offline
Newbie

Registered: 28/06/11
Posts: 2
Loc: Rio de Janeiro, Brazil
Does anybody know how to reset a 40L Stryker insufflator?
If you do, can you post here or email me?
My email is notwen_cardozo@hotmail.com
_________________________
Notwen

Top
Page 3 of 6 < 1 2 3 4 5 6 >



Moderator:  DaveC in Oz, KM, RoJo 

Sponsors

Press Releases
Who's Online
4 registered (Neil Porter, dindogurango, Huw, fortronik), 177 Guests and 14 Spiders online.
Key: Admin, Global Mod, Mod
Events
17th National Biomedical & Clinical Engineering Co
Welch Allyn - 12th Annual "Completing the Picture
May
Su M Tu W Th F Sa
1 2 3 4
5 6 7 8 9 10 11
12 13 14 15 16 17 18
19 20 21 22 23 24 25
26 27 28 29 30 31
Newest Members
Socio, Nigel Cross, apeters, Jakeyfox, davidescafa
7228 Registered Users
Featured Member
Registered: 17/05/09
Posts: 117
Today's Birthdays
No Birthdays
(Views)Popular Topics
Jokes 2 (son of Jokes!!!) 1124605
Classic Computers 384980
Magic Kingdom Biomed Expats 318865
The VRCT Website 316232
VRCT - Who's Re-Newed? 205493
Trainee EBME Technician Salary? 201243
VRCT renewal 185184
Jokes!! 180580
Sharing the info. 172698
DIRTY equipment 166601