#48039 - 07/08/10 10:37 AM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Savant
Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
|
I've got the service manual for the 40L and it states to check the gas supply. Hope that helps
|
|
Top
|
|
|
|
#48084 - 09/08/10 10:41 AM
Re: Stryker 30L highflow insufflator error 04
[Re: Pavel]
|
Savant
Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
|
Hi Guys, I work for Stryker as a Field Service Engineer, unforunately for you guys all the manuals I have are in electronic form in a program we use, I could probably get extracts if required, but if you post any problems you have I will try to help or atleast point you in the right direction.
|
|
Top
|
|
|
|
#48720 - 02/09/10 12:37 PM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Savant
Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
|
Like I stated in a previous post you need to check the gas supply, its possibly the internal regulator requires calibration and or repair/replacement.
"4.6.1 Checking Gas Supply Connection/Gas Inlet Filter (See Fig. 4-6-1-1) The gas inlet filter (2) may be plugged or soiled due to Soiling unclean gas or gas supply residues. This type of soiling, however, cannot be detected externally. Heavy soiling is especially noticeable during higher flow rates and results in the actual gas flow falling short of the nominal gas flow during active insufflation. If cleaning the surface of the gas inlet filter does not yield any improvement, the gas supply connection (1) has to be replaced completely"
Keep in mind this is for the 40L
|
|
Top
|
|
|
|
#48722 - 02/09/10 12:47 PM
Re: Stryker 30L highflow insufflator error 04
[Re: NickM]
|
Technologist
Registered: 17/06/10
Posts: 45
Loc: tamilnadu,india
|
|
|
Top
|
|
|
|
#48723 - 02/09/10 12:55 PM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Savant
Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
|
Let me know if you need the part numbers for a replacement regulator or gas inlet filter assy
|
|
Top
|
|
|
|
#50400 - 19/11/10 02:16 AM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Technologist
Registered: 22/09/08
Posts: 46
Loc: philippines
|
Hi NickM,
thanks for your post it helps me a lot, by the way can i have a copy of your manual? please email it to me, thanks in advance
Regards,
June
|
|
Top
|
|
|
|
#50751 - 02/12/10 12:12 PM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Master
Registered: 24/04/05
Posts: 322
Loc: Karachi
|
Hi NICK I am having ERROR 21 in stryker 30L insufflator. And there is another one ERROR 03 in Stryker 20L insufflator. Please do help me out.. Or if u can send me service manual. What ever easy for you. Waiting for quick reply. Regards
|
|
Top
|
|
|
|
#50774 - 04/12/10 12:10 PM
Re: Stryker 30L highflow insufflator error 04
[Re: Laiq]
|
Savant
Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
|
Error 21 is Fluid Sensor Error - Check and clean fluid sensor and all tube, replace if necessory.
Error 3 - Initialization error - Replace the CPU PCB.
Please beaware this unit is obsolete and stryker do not support it or spare parts anymore.
|
|
Top
|
|
|
|
#50787 - 06/12/10 01:25 PM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Master
Registered: 24/04/05
Posts: 322
Loc: Karachi
|
Dear NickM Thanks brother. About the location of FLUID SENSOR. Any location diagram or where to look. Thanks in advance. Regards
|
|
Top
|
|
|
|
#50801 - 07/12/10 11:44 AM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Savant
Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
|
It's an "L" shaped connector at the front of the unit.
|
|
Top
|
|
|
|
#50806 - 07/12/10 06:43 PM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Master
Registered: 24/04/05
Posts: 322
Loc: Karachi
|
Thanks Nick. I hsve found out one sensor in one of two horizontal tubes just behind the display board. It is HONEYWELL MICRO SWITCH three pin (+IN, GND, OUT) black body with some kind of optical sensor in the path of that pipe... Is this the same..??
|
|
Top
|
|
|
|
#51345 - 02/01/11 06:11 PM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Newbie
Registered: 02/01/11
Posts: 4
Loc: Mexico
|
Hi can you help me, I have a insunfflator 40 L Stryker with a error on the display say CONTAMINATION And nothing more I can not do any other thing ,and also ask for secure code. I'll appreciate it any of your help. Thanks.
|
|
Top
|
|
|
|
#51346 - 03/01/11 04:21 AM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Savant
Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
|
Requires a calibration, do you know when this was last done?
|
|
Top
|
|
|
|
#51439 - 08/01/11 06:10 PM
Re: Stryker 30L highflow insufflator error 04
[Re: NickM]
|
Newbie
Registered: 02/01/11
Posts: 4
Loc: Mexico
|
|
|
Top
|
|
|
|
#51445 - 09/01/11 02:12 AM
Re: Stryker 30L highflow insufflator error 04
[Re: NickM]
|
Newbie
Registered: 02/01/11
Posts: 4
Loc: Mexico
|
How can I do it? Or what should do? Thanks.
|
|
Top
|
|
|
|
#51476 - 11/01/11 10:39 AM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Savant
Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
|
Hi Inzunza, unfortunately i've been away from work today due to flooding in Brisbane, Queensland, Australia. Closing down our office.
If I get a chance I will forward through the calibration procedure via email in the next few days.
Basically the things you will need is a full bottle of co2 gas, 2 x 50ml Syringes, and a co2 pressure gauge.
The calibration procedure is fairly straight forward otherwise.
Edited by NickM (11/01/11 10:41 AM)
|
|
Top
|
|
|
|
#51494 - 12/01/11 06:26 AM
Re: Stryker 30L highflow insufflator error 04
[Re: NickM]
|
Newbie
Registered: 02/01/11
Posts: 4
Loc: Mexico
|
Thanks NickM not problem , I'll wait until you have time for it, hope you do not forget it  .
|
|
Top
|
|
|
|
#52367 - 22/02/11 08:22 AM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Newbie
Registered: 17/02/11
Posts: 3
Loc: Punjab
|
Dear Nick I also would need calibration procedure.please send at my email also dcads_mrkt@yahoo.com
|
|
Top
|
|
|
|
#55239 - 11/03/11 08:14 PM
Re: Stryker 30L highflow insufflator error 04
[Re: NickM]
|
Newbie
Registered: 10/03/11
Posts: 1
Loc: mexico michoacan
|
Hi i´m have the same problem with contamination on the display, can you help mi to solve it? i revove a little dusts in the outlet tube, but i don´t know how reset the unit, the unit is a 40 stryker, thanks for your help..
|
|
Top
|
|
|
|
#55559 - 28/03/11 03:44 PM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Newbie
Registered: 10/02/11
Posts: 2
Loc: Dallas, Texas USA
|
Hi, I have two Stryker 40L insufflators that I have brought back from a mission hospital in rural Guatemala. They both have an error message without a number. The error message is "SENSOR DEFECT." Any help would be appreciated. Does anyone have a service manual??
|
|
Top
|
|
|
|
#55566 - 29/03/11 07:58 AM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Savant
Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
|
Hi Ted this is due to contaminated gas you will need to replace the fluid sensor and LPU (Low Pressure Unit)
Part Number: 0105199130 Description: COMPLETE PNUEMATICS UNIT LPU
and the fluid sensor
Part Number: 105190933 Description: FLUID SENSOR
As this is for a mission hospital Stryker may be able to sort out some sort of donation repair, contact your local stryker rep if not let me know and I will see what I can do.
All the best
Nick (Field Service Engineer - Stryker South Pacific)
|
|
Top
|
|
|
|
#55627 - 01/04/11 01:41 AM
Re: Stryker 30L highflow insufflator error 04
[Re: NickM]
|
Newbie
Registered: 10/02/11
Posts: 2
Loc: Dallas, Texas USA
|
thanks Nick, The Stryker rep here does not seem to know how or if parts can be ordered. Do you know what these parts would cost or how I can get them? Thanks, Ted
|
|
Top
|
|
|
|
#55669 - 02/04/11 10:26 PM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Visionary
Registered: 01/06/09
Posts: 34
Loc: pak
|
dear nick i have stryker 20 liter and older 40liter with ereror 21 can you send me the copy of service mannula I had the coppy of stryker newer 40 L high flow and misplaced it can you help and send me a copy of that the eror is faulty valve
_________________________
Dr tahir mehmood 138 m d.h.a lahore pakistan
|
|
Top
|
|
|
|
#56623 - 03/06/11 08:04 PM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Visionary
Registered: 01/06/09
Posts: 34
Loc: pak
|
dear nick hi i have two of the 40l insufflator with calibration error on one and regulator error on the other can you help please
_________________________
Dr tahir mehmood 138 m d.h.a lahore pakistan
|
|
Top
|
|
|
|
#57005 - 28/06/11 01:35 PM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Newbie
Registered: 28/06/11
Posts: 2
Loc: Rio de Janeiro, Brazil
|
Does anybody know how to reset a 40L Stryker insufflator? If you do, can you post here or email me? My email is notwen_cardozo@hotmail.com
_________________________
Notwen
|
|
Top
|
|
|
|
#57223 - 14/07/11 06:10 PM
Re: Stryker 30L highflow insufflator error 04
[Re: NickM]
|
Newbie
Registered: 14/07/11
Posts: 1
Loc: Brazil - Mato Grosso
|
I'm a brazilian doctor. I have a styker insufflator 40L with calibration error. Do you know how I solve this? Can you help me? Please, do you have the stryker insufflator service manual? Can you send me?
mjosefr@uol.com.br
Thanks.
|
|
Top
|
|
|
|
#58128 - 09/09/11 01:24 AM
Re: Stryker 30L highflow insufflator error 04
[Re: NickM]
|
Newbie
Registered: 05/04/11
Posts: 6
Loc: MERIDA, YUCATAN, MEXICO
|
Dear Sir NickM could send the process of calibration or if possible the service manual to my email omar.glz.mendez @ gmail.com, as well as how to acquire each of the parts for the Stryker 40L High Flow Insufflator:
LPU (Low Pressure Unit) Part Number: 0105199130 Description: COMPLETE UNIT PNUEMATICS LPU
and the fluid sensor Part Number: 105190933 Description: FLUID SENSOR
Thanks in advance. Best regards
|
|
Top
|
|
|
|
#59398 - 05/12/11 01:58 PM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Newbie
Registered: 30/11/11
Posts: 2
Loc: Illinois
|
Hi NickM, I need the service manual for Stryker 40L insufflator. Please email to: mike.crowley@passavanthospital.com Thanks so much for your help.
|
|
Top
|
|
|
|
#59743 - 10/01/12 11:47 PM
Re: Stryker 30L highflow insufflator error 04
[Re: NickM]
|
Newbie
Registered: 16/12/11
Posts: 2
Loc: California, U.S.A
|
Hi Nick, Please give me a instruction of how to reset error 21 and service alarm after replaced fluid sensor on 30L that will be appreciated. chuleema@yahoo.com Thanks. Best Joel
|
|
Top
|
|
|
|
#59761 - 13/01/12 06:25 PM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Newbie
Registered: 13/01/12
Posts: 1
Loc: guadalajara, jalisco,mexico
|
hello I have a styker insufflator 40L with contamination. Do you know how I solve this? Can you help me? Can you send me the service code? or the way to reset
Thanks.
|
|
Top
|
|
|
|
#59799 - 19/01/12 08:23 AM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Newbie
Registered: 17/01/12
Posts: 2
Loc: tbilisi
|
hi everybody, I need service manual for insufflator 30L, please send me, thanks.
|
|
Top
|
|
|
|
#59800 - 19/01/12 08:28 AM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Newbie
Registered: 17/01/12
Posts: 2
Loc: tbilisi
|
My mail is bmalkhaz@gmail.com
|
|
Top
|
|
|
|
#59818 - 20/01/12 08:48 PM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Newbie
Registered: 16/12/11
Posts: 2
Loc: California, U.S.A
|
Hello, I need a reset code for error 21 (fluid sensor) on stryker 30L if anybody can help me I really appreciate it. pls post here or email me chuleema@yahoo.com
|
|
Top
|
|
|
|
#59833 - 24/01/12 09:14 PM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Newbie
Registered: 24/01/12
Posts: 1
Loc: México D.F.
|
Hello, I need service manuals stryker? Does any body helpme, please post here or email me segzack@hotmail.com
|
|
Top
|
|
|
|
#59928 - 02/02/12 09:49 PM
Re: Stryker 30L highflow insufflator error 04
[Re: NickM]
|
Newbie
Registered: 14/07/11
Posts: 2
Loc: mexico
|
hello nick We would appreciate it if I could, provide a copy of the manual stryker 30L insufflator
require assistance with error 21
my email is: mgarcia@magbiomedica.com
thank you very much for your help be most grateful to you
|
|
Top
|
|
|
|
#60118 - 21/02/12 08:23 AM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Newbie
Registered: 14/02/12
Posts: 1
Loc: Budapest, Hungary
|
Hi Nick, Would you be so kind and help me out with an error code E04 in a Stryker 40L insufflator? How could we resolve this problem? What could have caused it? The unit has been running perfectly for over 10 years and suddenly it stopped with this message. We also have a similar type but 30L flow unit. Can that have the same problem? Any helps or materials would be very welcome. My email address is at.kalman@gmail.com. Thanks a lot for your kind assistance in advance! Attila
|
|
Top
|
|
|
|
#60427 - 16/03/12 01:01 AM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Newbie
Registered: 15/03/12
Posts: 1
Loc: Brazil, Rio de Janeiro, RJ
|
Hi.
Error code 4 is for General gas supply error.
Lutar
|
|
Top
|
|
|
|
#61024 - 08/05/12 02:04 AM
Re: Stryker 30L highflow insufflator error 04
[Re: askrp]
|
Savant
Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
|
Sorry for the delay havn't been on here in a while, if it's been running perfectly for 10 years is it time to upgrade?
If not how often is the unit serviced and calibrated?
|
|
Top
|
|
|
|
|
|
10 registered (Pavel, GeorgeK, Mofe, Paul Tregurtha, mek.sah, DinhVu, ani, DaveC in Oz, Huw, pat),
284
Guests and
8
Spiders online. |
|
Key:
Admin,
Global Mod,
Mod
|
|
|
|
|
1
|
2
|
3
|
4
|
5
|
|
6
|
7
|
8
|
9
|
10
|
11
|
12
|
|
13
|
14
|
15
|
16
|
17
|
18
|
19
|
|
20
|
21
|
22
|
23
|
24
|
25
|
26
|
|
27
|
28
|
29
|
30
|
31
|
|
|
|
|