#48019 - 06/08/10 08:32 AM Stryker 30L highflow insufflator error 04
askrp Offline
Technologist

Registered: 17/06/10
Posts: 45
Loc: tamilnadu,india
hi everybody i have the problem with stryker 30Lhigh flow insufullator, problem is m/c shows E04 at Gasflow any one service manual,pls send me or guide me to rectify the problem,advance thanks

Top
#48039 - 07/08/10 10:37 AM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
NickM Offline
Savant

Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
I've got the service manual for the 40L and it states to check the gas supply. Hope that helps

Top
#48042 - 07/08/10 01:16 PM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
askrp Offline
Technologist

Registered: 17/06/10
Posts: 45
Loc: tamilnadu,india
thanks sir,my email id : rajan_bme@yahoo.co.in pls send to me

Top
#48043 - 07/08/10 03:06 PM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
Pavel Online   content
Mentor

Registered: 11/12/09
Posts: 155
Loc: Puebla, Mexico
Hi NickM. Do yu can send me service manual, too?
hugo_68@yahoo.com
Do you have other Stryker service manuals? Lighth sources per example.
Thanks

Top
#48084 - 09/08/10 10:41 AM Re: Stryker 30L highflow insufflator error 04 [Re: Pavel]
NickM Offline
Savant

Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
Hi Guys, I work for Stryker as a Field Service Engineer, unforunately for you guys all the manuals I have are in electronic form in a program we use, I could probably get extracts if required, but if you post any problems you have I will try to help or atleast point you in the right direction.


Top
#48085 - 09/08/10 11:15 AM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
DaveC in Oz Online   crying
Philosopher

Registered: 26/06/09
Posts: 594
Loc: Brisbane, Australia
Hi Nick, how's the new job going?, PM me if you like (he used to be one of my guys for those who are going....."what the hell?")

Dave


Edited by DaveC in Oz (09/08/10 11:29 AM)

Top
#48674 - 01/09/10 07:30 AM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
askrp Offline
Technologist

Registered: 17/06/10
Posts: 45
Loc: tamilnadu,india
When put on the m/c it shows error 4, kindly guide me,advance thanks

Top
#48720 - 02/09/10 12:37 PM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
NickM Offline
Savant

Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
Like I stated in a previous post you need to check the gas supply, its possibly the internal regulator requires calibration and or repair/replacement.

"4.6.1 Checking Gas Supply Connection/Gas Inlet Filter
(See Fig. 4-6-1-1) The gas inlet filter (2) may be plugged or soiled due to Soiling
unclean gas or gas supply residues. This type of soiling, however, cannot
be detected externally. Heavy soiling is especially noticeable during
higher flow rates and results in the actual gas flow falling short of the
nominal gas flow during active insufflation.
If cleaning the surface of the gas inlet filter does not yield any
improvement, the gas supply connection (1) has to be replaced
completely"

Keep in mind this is for the 40L

Top
#48722 - 02/09/10 12:47 PM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
askrp Offline
Technologist

Registered: 17/06/10
Posts: 45
Loc: tamilnadu,india
thanks i will check it

Top
#48723 - 02/09/10 12:55 PM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
NickM Offline
Savant

Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
Let me know if you need the part numbers for a replacement regulator or gas inlet filter assy

Top
#50400 - 19/11/10 02:16 AM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
june Offline
Technologist

Registered: 22/09/08
Posts: 46
Loc: philippines
Hi NickM,

thanks for your post it helps me a lot, by the way can i have a copy of your manual? please email it to me, thanks in advance

Regards,

June

Top
#50751 - 02/12/10 12:12 PM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
Laiq Offline
Master

Registered: 24/04/05
Posts: 322
Loc: Karachi
Hi NICK
I am having ERROR 21 in stryker 30L insufflator. And there is another one ERROR 03 in Stryker 20L insufflator.
Please do help me out..
Or if u can send me service manual.
What ever easy for you.
Waiting for quick reply.
Regards

Top
#50774 - 04/12/10 12:10 PM Re: Stryker 30L highflow insufflator error 04 [Re: Laiq]
NickM Offline
Savant

Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
Error 21 is Fluid Sensor Error - Check and clean fluid sensor and all tube, replace if necessory.

Error 3 - Initialization error - Replace the CPU PCB.

Please beaware this unit is obsolete and stryker do not support it or spare parts anymore.

Top
#50787 - 06/12/10 01:25 PM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
Laiq Offline
Master

Registered: 24/04/05
Posts: 322
Loc: Karachi
Dear NickM
Thanks brother.
About the location of FLUID SENSOR. Any location diagram or where to look.
Thanks in advance.
Regards

Top
#50801 - 07/12/10 11:44 AM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
NickM Offline
Savant

Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
It's an "L" shaped connector at the front of the unit.

Top
#50806 - 07/12/10 06:43 PM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
Laiq Offline
Master

Registered: 24/04/05
Posts: 322
Loc: Karachi
Thanks Nick. I hsve found out one sensor in one of two horizontal tubes just behind the display board. It is HONEYWELL MICRO SWITCH three pin (+IN, GND, OUT) black body with some kind of optical sensor in the path of that pipe... Is this the same..??

Top
#51345 - 02/01/11 06:11 PM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
Inzunza Offline
Newbie

Registered: 02/01/11
Posts: 4
Loc: Mexico
Hi can you help me, I have a insunfflator 40 L Stryker with a error on the display say CONTAMINATION And nothing more I can not do any other thing ,and also ask for secure code.
I'll appreciate it any of your help. Thanks.

Top
#51346 - 03/01/11 04:21 AM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
NickM Offline
Savant

Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
Requires a calibration, do you know when this was last done?

Top
#51439 - 08/01/11 06:10 PM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
Inzunza Offline
Newbie

Registered: 02/01/11
Posts: 4
Loc: Mexico
Not I do not.

Top
#51445 - 09/01/11 02:12 AM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
Inzunza Offline
Newbie

Registered: 02/01/11
Posts: 4
Loc: Mexico
How can I do it? Or what should do?
Thanks.

Top
#51476 - 11/01/11 10:39 AM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
NickM Offline
Savant

Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
Hi Inzunza, unfortunately i've been away from work today due to flooding in Brisbane, Queensland, Australia. Closing down our office.

If I get a chance I will forward through the calibration procedure via email in the next few days.

Basically the things you will need is a full bottle of co2 gas, 2 x 50ml Syringes, and a co2 pressure gauge.

The calibration procedure is fairly straight forward otherwise.


Edited by NickM (11/01/11 10:41 AM)

Top
#51494 - 12/01/11 06:26 AM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
Inzunza Offline
Newbie

Registered: 02/01/11
Posts: 4
Loc: Mexico
Thanks NickM not problem , I'll wait until you have time for it, hope you do not forget it smile.

Top
#52367 - 22/02/11 08:22 AM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
Dps Offline
Newbie

Registered: 17/02/11
Posts: 3
Loc: Punjab
Dear Nick I also would need calibration procedure.please send at my email also dcads_mrkt@yahoo.com

Top
#55239 - 11/03/11 08:14 PM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
remocnab Offline
Newbie

Registered: 10/03/11
Posts: 1
Loc: mexico michoacan
Hi i´m have the same problem with contamination on the display, can you help mi to solve it? i revove a little dusts in the outlet tube, but i don´t know how reset the unit, the unit is a 40 stryker, thanks for your help..

Top
#55559 - 28/03/11 03:44 PM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
TedF Offline
Newbie

Registered: 10/02/11
Posts: 2
Loc: Dallas, Texas USA
Hi, I have two Stryker 40L insufflators that I have brought back from a mission hospital in rural Guatemala. They both have an error message without a number. The error message is "SENSOR DEFECT." Any help would be appreciated. Does anyone have a service manual??

Top
#55566 - 29/03/11 07:58 AM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
NickM Offline
Savant

Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
Hi Ted this is due to contaminated gas you will need to replace the fluid sensor and LPU (Low Pressure Unit)

Part Number: 0105199130
Description: COMPLETE PNUEMATICS UNIT LPU

and the fluid sensor

Part Number: 105190933
Description: FLUID SENSOR

As this is for a mission hospital Stryker may be able to sort out some sort of donation repair, contact your local stryker rep if not let me know and I will see what I can do.

All the best

Nick (Field Service Engineer - Stryker South Pacific)

Top
#55627 - 01/04/11 01:41 AM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
TedF Offline
Newbie

Registered: 10/02/11
Posts: 2
Loc: Dallas, Texas USA
thanks Nick,
The Stryker rep here does not seem to know how or if parts can be ordered. Do you know what these parts would cost or how I can get them?
Thanks,
Ted

Top
#55669 - 02/04/11 10:26 PM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
tahir mehmood Offline
Visionary

Registered: 01/06/09
Posts: 34
Loc: pak
dear nick
i have stryker 20 liter and older 40liter with ereror 21 can you send me the copy of service mannula
I had the coppy of stryker newer 40 L high flow and misplaced it can you help
and send me a copy of that
the eror is faulty valve
_________________________
Dr tahir mehmood
138 m d.h.a
lahore pakistan

Top
#56623 - 03/06/11 08:04 PM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
tahir mehmood Offline
Visionary

Registered: 01/06/09
Posts: 34
Loc: pak
dear nick hi
i have two of the 40l insufflator with calibration error on one and regulator error on the other can you help please
_________________________
Dr tahir mehmood
138 m d.h.a
lahore pakistan

Top
#57005 - 28/06/11 01:35 PM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
NotwenCardozo Offline
Newbie

Registered: 28/06/11
Posts: 2
Loc: Rio de Janeiro, Brazil
Does anybody know how to reset a 40L Stryker insufflator?
If you do, can you post here or email me?
My email is notwen_cardozo@hotmail.com
_________________________
Notwen

Top
#57223 - 14/07/11 06:10 PM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
Marcello Offline
Newbie

Registered: 14/07/11
Posts: 1
Loc: Brazil - Mato Grosso
I'm a brazilian doctor.
I have a styker insufflator 40L with calibration error.
Do you know how I solve this? Can you help me?
Please, do you have the stryker insufflator service manual?
Can you send me?

mjosefr@uol.com.br

Thanks.

Top
#58128 - 09/09/11 01:24 AM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
Omar Gonzalez Offline
Newbie

Registered: 05/04/11
Posts: 6
Loc: MERIDA, YUCATAN, MEXICO
Dear Sir NickM
could send the process of calibration or if possible the service manual to my email omar.glz.mendez @ gmail.com, as well as how to acquire each of the parts for the Stryker 40L High Flow Insufflator:

LPU (Low Pressure Unit)
Part Number: 0105199130
Description: COMPLETE UNIT PNUEMATICS LPU

and the fluid sensor
Part Number: 105190933
Description: FLUID SENSOR

Thanks in advance.
Best regards

Top
#59398 - 05/12/11 01:58 PM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
troutn222 Offline
Newbie

Registered: 30/11/11
Posts: 2
Loc: Illinois
Hi NickM, I need the service manual for Stryker 40L insufflator. Please email to: mike.crowley@passavanthospital.com
Thanks so much for your help.

Top
#59743 - 10/01/12 11:47 PM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
JSP Offline
Newbie

Registered: 16/12/11
Posts: 2
Loc: California, U.S.A
Hi Nick, Please give me a instruction of how to reset error 21 and service alarm after replaced fluid sensor on 30L
that will be appreciated. chuleema@yahoo.com
Thanks.
Best Joel

Top
#59761 - 13/01/12 06:25 PM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
cristobal Offline
Newbie

Registered: 13/01/12
Posts: 1
Loc: guadalajara, jalisco,mexico
hello
I have a styker insufflator 40L with contamination.
Do you know how I solve this? Can you help me?
Can you send me the service code? or the way to reset

Thanks.

Top
#59799 - 19/01/12 08:23 AM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
abcd Offline
Newbie

Registered: 17/01/12
Posts: 2
Loc: tbilisi
hi everybody, I need service manual for insufflator 30L, please send me, thanks.

Top
#59800 - 19/01/12 08:28 AM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
abcd Offline
Newbie

Registered: 17/01/12
Posts: 2
Loc: tbilisi
My mail is bmalkhaz@gmail.com

Top
#59818 - 20/01/12 08:48 PM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
JSP Offline
Newbie

Registered: 16/12/11
Posts: 2
Loc: California, U.S.A
Hello, I need a reset code for error 21 (fluid sensor) on stryker 30L if anybody can help me I really appreciate it.
pls post here or email me chuleema@yahoo.com

Top
#59833 - 24/01/12 09:14 PM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
Begaman Offline
Newbie

Registered: 24/01/12
Posts: 1
Loc: México D.F.
Hello, I need service manuals stryker?
Does any body helpme, please post here or email me segzack@hotmail.com

Top
#59928 - 02/02/12 09:49 PM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
mgarcia Offline
Newbie

Registered: 14/07/11
Posts: 2
Loc: mexico
hello nick
We would appreciate it if I could, provide a copy of the manual stryker 30L insufflator

require assistance with error 21

my email is: mgarcia@magbiomedica.com

thank you very much for your help be most grateful to you

Top
#60118 - 21/02/12 08:23 AM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
Attila Offline
Newbie

Registered: 14/02/12
Posts: 1
Loc: Budapest, Hungary
Hi Nick,
Would you be so kind and help me out with an error code E04 in a Stryker 40L insufflator? How could we resolve this problem? What could have caused it?
The unit has been running perfectly for over 10 years and suddenly it stopped with this message. We also have a similar type but 30L flow unit. Can that have the same problem?
Any helps or materials would be very welcome. My email address is at.kalman@gmail.com.
Thanks a lot for your kind assistance in advance!
Attila

Top
#60427 - 16/03/12 01:01 AM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
lutar Offline
Newbie

Registered: 15/03/12
Posts: 1
Loc: Brazil, Rio de Janeiro, RJ
Hi.

Error code 4 is for General gas supply error.

Lutar

Top
#61024 - 08/05/12 02:04 AM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
NickM Offline
Savant

Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
Sorry for the delay havn't been on here in a while, if it's been running perfectly for 10 years is it time to upgrade?

If not how often is the unit serviced and calibrated?

Top
Page 1 of 5 1 2 3 4 5 >



Moderator:  DaveC in Oz, KM, RoJo 
Sponsors
Press Releases
Who's Online
10 registered (Pavel, GeorgeK, Mofe, Paul Tregurtha, mek.sah, DinhVu, ani, DaveC in Oz, Huw, pat), 284 Guests and 8 Spiders online.
Key: Admin, Global Mod, Mod
May
Su M Tu W Th F Sa
1 2 3 4 5
6 7 8 9 10 11 12
13 14 15 16 17 18 19
20 21 22 23 24 25 26
27 28 29 30 31
Events