#48019 - 06/08/10 08:32 AM Stryker 30L highflow insufflator error 04
askrp Offline
Technologist

Registered: 17/06/10
Posts: 45
Loc: tamilnadu,india
hi everybody i have the problem with stryker 30Lhigh flow insufullator, problem is m/c shows E04 at Gasflow any one service manual,pls send me or guide me to rectify the problem,advance thanks

Top
#48039 - 07/08/10 10:37 AM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
NickM Offline
Savant

Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
I've got the service manual for the 40L and it states to check the gas supply. Hope that helps

Top
#48042 - 07/08/10 01:16 PM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
askrp Offline
Technologist

Registered: 17/06/10
Posts: 45
Loc: tamilnadu,india
thanks sir,my email id : rajan_bme@yahoo.co.in pls send to me

Top
#48043 - 07/08/10 03:06 PM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
Pavel Online   content
Mentor

Registered: 11/12/09
Posts: 155
Loc: Puebla, Mexico
Hi NickM. Do yu can send me service manual, too?
hugo_68@yahoo.com
Do you have other Stryker service manuals? Lighth sources per example.
Thanks

Top
#48084 - 09/08/10 10:41 AM Re: Stryker 30L highflow insufflator error 04 [Re: Pavel]
NickM Offline
Savant

Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
Hi Guys, I work for Stryker as a Field Service Engineer, unforunately for you guys all the manuals I have are in electronic form in a program we use, I could probably get extracts if required, but if you post any problems you have I will try to help or atleast point you in the right direction.


Top
#48085 - 09/08/10 11:15 AM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
DaveC in Oz Online   crying
Philosopher

Registered: 26/06/09
Posts: 594
Loc: Brisbane, Australia
Hi Nick, how's the new job going?, PM me if you like (he used to be one of my guys for those who are going....."what the hell?")

Dave


Edited by DaveC in Oz (09/08/10 11:29 AM)

Top
#48674 - 01/09/10 07:30 AM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
askrp Offline
Technologist

Registered: 17/06/10
Posts: 45
Loc: tamilnadu,india
When put on the m/c it shows error 4, kindly guide me,advance thanks

Top
#48720 - 02/09/10 12:37 PM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
NickM Offline
Savant

Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
Like I stated in a previous post you need to check the gas supply, its possibly the internal regulator requires calibration and or repair/replacement.

"4.6.1 Checking Gas Supply Connection/Gas Inlet Filter
(See Fig. 4-6-1-1) The gas inlet filter (2) may be plugged or soiled due to Soiling
unclean gas or gas supply residues. This type of soiling, however, cannot
be detected externally. Heavy soiling is especially noticeable during
higher flow rates and results in the actual gas flow falling short of the
nominal gas flow during active insufflation.
If cleaning the surface of the gas inlet filter does not yield any
improvement, the gas supply connection (1) has to be replaced
completely"

Keep in mind this is for the 40L

Top
#48722 - 02/09/10 12:47 PM Re: Stryker 30L highflow insufflator error 04 [Re: NickM]
askrp Offline
Technologist

Registered: 17/06/10
Posts: 45
Loc: tamilnadu,india
thanks i will check it

Top
#48723 - 02/09/10 12:55 PM Re: Stryker 30L highflow insufflator error 04 [Re: askrp]
NickM Offline
Savant

Registered: 18/12/09
Posts: 103
Loc: Brisbane, Australia
Let me know if you need the part numbers for a replacement regulator or gas inlet filter assy

Top
Page 1 of 5 1 2 3 4 5 >



Moderator:  DaveC in Oz, KM, RoJo 
Sponsors
Press Releases
Who's Online
10 registered (Pavel, GeorgeK, Mofe, Paul Tregurtha, mek.sah, DinhVu, ani, DaveC in Oz, Huw, pat), 284 Guests and 8 Spiders online.
Key: Admin, Global Mod, Mod
May
Su M Tu W Th F Sa
1 2 3 4 5
6 7 8 9 10 11 12
13 14 15 16 17 18 19
20 21 22 23 24 25 26
27 28 29 30 31
Events