I have emailed you our infopath test spec we use for annual checks using a sigmapace 1000 on 4170's.

If anything does not conform then its out with the scope and function gen, or send it back to APC.

We used to have a homemade bit of test kit but its now been replaced by the sigmapace - homemade kit was better though IMHO! and also the amount learned by making it :O) Still not that we have time for projects like that these days.

anyway hope this helps

Gary

Top
#51980 - 04/02/11 03:51 PM Re: Testing APC Medical E4170 Bedside Pacemaker [Re: Biomed]
Paul J Offline
Visionary

Registered: 18/04/07
Posts: 37
Loc: Great Western Hospital
Just finished testing one of these today. We bought a new arbitrary wave sig gen (Agilent 33220A) especially to do these (& to replace some older generators we had knocking around). This combined with a Fluke 124 Scopemeter, a calibrated DVM & bench PSU is all you need in terms of test kit. You do have to knock up the 200:1 attenuating test circuit (as shown in the manual) & away you go!
I do find them a bit tricky to test up in our normal workshop as the room seems to be extra noisy from a combination of altitude & loads of fluorescent lights etc., but it can be done.
_________________________
Grey matter migrating to the outside of my head...
http://www.youtube.com/watch?v=iRTnMHW-A_Y

Top
Page 2 of 2 < 1 2



Moderator:  DaveC in Oz, KM, RoJo 
Sponsors
Press Releases
Who's Online
12 registered (murugan18, DaveC in Oz, hieu, Pavel, GeorgeK, Mofe, Paul Tregurtha, mek.sah, DinhVu, ani, Huw, pat), 261 Guests and 10 Spiders online.
Key: Admin, Global Mod, Mod
May
Su M Tu W Th F Sa
1 2 3 4 5
6 7 8 9 10 11 12
13 14 15 16 17 18 19
20 21 22 23 24 25 26
27 28 29 30 31
Events