#48844 - 07/09/10 10:29 AM CRITICARE 508 Patient Monitor INFALTION PRESSURE
Laiq Offline
Master

Registered: 24/04/05
Posts: 330
Loc: Karachi
Dear all
I have a Criticare 508 Patient Monitor.I have calibrated the nibp pressure with mmHg meter. The problem is that the initial inflation pressure is very low that is 110mmHg. So in its first attempt it fail to get reading, then it inflate again goes to 125mmHg & again no reading same way it gets a reading after 3 or 4 attempts. the patients are not satisfied with it, they need to get reading in first attempt. Can any one share their experience if i can change any setting in setup so that the inflation pressure should start from say 160 or 170 mmHg.
How to get into its setup menu.
Hope to get quick reply.
Thanks in advance & regards
LAIQ

Top
#48845 - 07/09/10 10:36 AM Re: CRITICARE 508 Patient Monitor INFALTION PRESSURE [Re: Laiq]
DaveC in Oz Offline
Philosopher

Registered: 26/06/09
Posts: 606
Loc: Brisbane, Australia
Sounds like the monitor is in neonatal mode (or paediatric) hence the low starting pressure. Don't know how to access the set ups however but perhaps someone else can assist with that,

Top
#48846 - 07/09/10 10:59 AM Re: CRITICARE 508 Patient Monitor INFALTION PRESSURE [Re: Laiq]
Laiq Offline
Master

Registered: 24/04/05
Posts: 330
Loc: Karachi
Thanks Davec
Yes i am thinking on same line. There is a CONFIG button by this u can change some settings, in there Spo2 mode showing ADULT, dont know about NIBP settings as there is none.
Thanks for your help anyway. Hope some one can get my answer.
Regards

Top
#48854 - 07/09/10 02:10 PM Re: CRITICARE 508 Patient Monitor INFALTION PRESSURE [Re: Laiq]
Graham Barnes Offline
Adept

Registered: 03/05/02
Posts: 89
Loc: York District Hospital
I think the Criticare NIBP takes its readings as the cuff inflates and not while deflating like Dinamap etc. It is claimed this is more comfortable for the patient.

Top
#48973 - 14/09/10 02:05 PM Re: CRITICARE 508 Patient Monitor INFALTION PRESSURE [Re: Graham Barnes]
Laiq Offline
Master

Registered: 24/04/05
Posts: 330
Loc: Karachi
Dear Sir
Can u explain a bit why it is getting low inflation pressure during first two attempts. I have replaced the cuff & pipes with new one, but still same behaviour. Is there any operating manual for it so that i can get the picture of what may be the problem.
Waiting for quick reply.
Regards
LAIQ

Top



Moderator:  DaveC in Oz, KM, RoJo 

Sponsors

Press Releases
Who's Online
9 registered (fortronik, Bruno Ferreira, Huw, Ed SWM, aghakhani, Marcelli, Chris Watts, big steve, sylar), 283 Guests and 16 Spiders online.
Key: Admin, Global Mod, Mod
Events
17th National Biomedical & Clinical Engineering Co
Welch Allyn - 12th Annual "Completing the Picture
May
Su M Tu W Th F Sa
1 2 3 4
5 6 7 8 9 10 11
12 13 14 15 16 17 18
19 20 21 22 23 24 25
26 27 28 29 30 31
Newest Members
Bruno Ferreira, EBMEnathan, ferdigia, paqste01, alexmac
7238 Registered Users
Featured Member
Registered: 10/02/03
Posts: 301
Today's Birthdays
Jandre, Lawrie
(Views)Popular Topics
Jokes 2 (son of Jokes!!!) 1133772
Classic Computers 388846
Magic Kingdom Biomed Expats 323129
The VRCT Website 317304
VRCT - Who's Re-Newed? 205932
Trainee EBME Technician Salary? 201721
VRCT renewal 185677
Jokes!! 181750
Sharing the info. 173461
DIRTY equipment 167055