#44601 - 05/03/10 10:12 AM Graseby 3100
Serroni Offline
Novice

Registered: 21/07/09
Posts: 15
Loc: Barreiro , Portugal
I was reading an older post here that had been placed here about a fault 86 and already did what was indicated to replace the 10v 100uf capacitor C3, but the error continues.
Does anybody know other possible causes for this fault?

Regards

Top
#44604 - 05/03/10 03:44 PM Re: Graseby 3100 [Re: Serroni]
HARI_CHINGAPURAM Offline
Newbie

Registered: 07/04/05
Posts: 4
Loc: London
Fault 86 may be due to a faulty motor. Try replacing the motor.

Top
#44646 - 08/03/10 04:29 PM Re: Graseby 3100 [Re: HARI_CHINGAPURAM]
Serroni Offline
Novice

Registered: 21/07/09
Posts: 15
Loc: Barreiro , Portugal
the error was mine i was changing the wrong capacitor . Problem solved .

Thanks Anyway

Regards

Top
#44650 - 08/03/10 05:27 PM Re: Graseby 3100 [Re: Serroni]
Geoff Hannis Online   content
Super Hero

Registered: 12/02/04
Posts: 10300
Loc: the path less trodden

It takes a Man to own up to mistakes in Open Forum, Serroni.

So ... it was the usual suspect after all then (C3)? smile

Top
#44679 - 09/03/10 09:03 AM Re: Graseby 3100 [Re: Geoff Hannis]
Serroni Offline
Novice

Registered: 21/07/09
Posts: 15
Loc: Barreiro , Portugal

yes it was the c3 but i was changing the c3 from the power board.

Then i see tha it was from the PCB Main that needs to be changed.

Regards smile


Top
#51382 - 05/01/11 12:40 PM Re: Graseby 3100 [Re: Serroni]
Nimish.P Offline
Visionary

Registered: 17/04/10
Posts: 32
Loc: Mumbai-Maharastra-India
Hi guys,

I have one Grasbey 3100, with Fault 91 on screen, anyone can help with the service manual or suggestions most welcome

Regards
Nimish

Top
#51385 - 05/01/11 01:08 PM Re: Graseby 3100 [Re: Nimish.P]
Geoff Hannis Online   content
Super Hero

Registered: 12/02/04
Posts: 10300
Loc: the path less trodden

Code 91 is a "keyboard fault". frown

The manual suggests "replace keyboard"!

But first check for a stuck key "button", fluid "ingress", bad connection(s) ... and all that usual stuff.

Top
#51397 - 06/01/11 03:06 AM Re: Graseby 3100 [Re: Geoff Hannis]
ariel dulay Offline
Adept

Registered: 24/09/10
Posts: 98
Loc: philippines
usually, when it comes to "keyboard fault",its only need general cleaning smile
_________________________
--- PATIENT SAFETY FIRST ---

Top
#51401 - 06/01/11 06:49 AM Re: Graseby 3100 [Re: Serroni]
Nimish.P Offline
Visionary

Registered: 17/04/10
Posts: 32
Loc: Mumbai-Maharastra-India
Geoff and Ariel,

Thanks for msg, will check and let inform you

Top
#51421 - 07/01/11 11:35 AM Re: Graseby 3100 [Re: Serroni]
Nimish.P Offline
Visionary

Registered: 17/04/10
Posts: 32
Loc: Mumbai-Maharastra-India
Yes it does solve, i wonder why couldnt they display " check keypad" instead of Fault 91.... anyways thanks guys need anything just send email....

Top
Page 1 of 2 1 2 >



Moderator:  DaveC in Oz, KM, RoJo 
Sponsors
Press Releases
Who's Online
13 registered (DinhVu, Geoff Hannis, Huw, RoJo, KM, hieu, Lee S, Laiq, lm1253, Jameson, DaveC in Oz, GeorgeK, Hasan), 294 Guests and 7 Spiders online.
Key: Admin, Global Mod, Mod
May
Su M Tu W Th F Sa
1 2 3 4 5
6 7 8 9 10 11 12
13 14 15 16 17 18 19
20 21 22 23 24 25 26
27 28 29 30 31
Events