#5186 - 27/12/02 06:53 PM Water in Medical Air
johnmac Offline
Newbie

Registered: 02/07/02
Posts: 2
Loc: Great Western Hospital Swindon
Two weeks ago, due to a breakdown in the dryers on the medical air compressor, water condensate passed through our piped medical gas system into a ventilator. Just as we were moving hospitals too. Exciting times. Have any other hospitals experienced this and what action did you take to protect your patients and make sure the vents were safe to use in the long term?

Top
#5187 - 28/12/02 07:18 PM Re: Water in Medical Air
John Sandham Offline

Hero

Registered: 03/07/00
Posts: 1711
Loc: UK
We also had a problem with our driers a couple of weeks ago. We replaced all the inlet filters on our Siemens 300 vents due to moisture ingress. We also checked all other areas on that pipeline.

It coincided with a cold snap, so condensation was suspected. Medaes were very quick coming in (a couple of hours). (They replaced the drying crystals.)
_________________________
Be Proactive and reactive.

Top
#5188 - 29/12/02 05:05 AM Re: Water in Medical Air
EH Offline
Expert

Registered: 23/10/02
Posts: 130
Hi Guy's, probably sounds a silly question but did you guys involve a pharmacist for a QC check ? confused
_________________________
A problem is a gift to be unwrapped :rolleyes:

Top
#5189 - 30/12/02 12:28 PM Re: Water in Medical Air
KM Online   content
Philosopher

Registered: 30/08/01
Posts: 729
Loc: LHCH
Ffloyd is right amongst the many jargonesc documents that control the work we all deal with are HTM 2022 and a pharmacy document "whos name i cannot rememember at this point" which stipulate what medical pipeline supplies are given to users and what quality they have to pass. rolleyes
Any authorised person worth their salt would have had a QC test done on the gas supply after any repair, I assume this is so. Otherwise you could end up in a situation that we had many moons ago when driers failed on MA4 supply. Several months later mixers in vents started to fail at an alarming rate. Id suggest you keep a close eye on your gas analysers and if the vents are using the air as a drive gas eg Erica / Elvira via an EDV get in and check theres no crud got past the line filters. wink

Top
#5190 - 30/12/02 04:35 PM Re: Water in Medical Air
John Sandham Offline

Hero

Registered: 03/07/00
Posts: 1711
Loc: UK
We didn't involve the pharmacist, Where does it say we should?
The equipment is on a comprehensive agreement with our service supplier. Surely they carry out all relevent tests??
confused
_________________________
Be Proactive and reactive.

Top
#5191 - 31/12/02 08:38 AM Re: Water in Medical Air
Tone Offline
Novice

Registered: 11/12/02
Posts: 12
Loc: London
I doubt whether a contractor is able to perform quality tests that would normally be carried out by a Quality Pharmacists. One of the things that they check is gas purity, I very much doubt if a visiting engineer would even have the instruments to perform such a test. Making sure the air is dry is one thing, ensuring that it contains nothing else is another (which is what the Pharmacist primarily checks for). John I suggest you speak to your Pharmacy department. The document HTM 2022 referred earlier should be available in your Esatates dept, a quick search on HTM 2022 mentions QC checks by Quality Pharmacists.

Top
#5192 - 31/12/02 08:51 AM Re: Water in Medical Air
Tone Offline
Novice

Registered: 11/12/02
Posts: 12
Loc: London
Sorry forgot to mention, when I worked for another Trust a few years ago we also had problems with water in our medical air piped supply. Estates, Pharmacy and infection control all worked together to ensure the the problem was rectified. Even though the compressor got fixed relatively quickly it was weeks before the supply was restored. Infection Control and Pharmacy insisted that all the pipes were throughly dried as the moisture represented an infection risk, once they were convinced it was dry purity checks were performed. Some areas I believe failed which meant that some sections of piping had to be flushed through and retested, this was repeated until pharmacy and infection control was happy. The whole saga lasted about a month with some very unhappy ward staff and porters (the poor chaps had to lug G sized cylinders all over the place), although I believe BOC and mobile compressor suppliers benefited quite well out of it. Therefore completing repairs to the piped gas system is only the tip of the iceberg, QC checks should not ne ignored as it can uncover infection issues.

Top
#5193 - 31/12/02 10:13 AM Re: Water in Medical Air
EH Offline
Expert

Registered: 23/10/02
Posts: 130
John, I am somewhat confused here, confused who is your hospital “Authorized Person” [AP], I attended the “Competent Persons” course in Eastwood Park, this is a internationally recognized standard, cool even out here in the gulf, eek and my manual stipulates that any “OPEN WORK” done on “Any” medical gas pipeline system must have a QC on completion of the work and before use. Obviously under AP supervision. John, you can call eastwood park for more details, cool

Regards

Ffloyd
_________________________
A problem is a gift to be unwrapped :rolleyes:

Top
#5194 - 31/12/02 12:07 PM Re: Water in Medical Air
KM Online   content
Philosopher

Registered: 30/08/01
Posts: 729
Loc: LHCH
Sorry chaps, you can go on as many courses as you like.
The facts are that the type of work mentioned should be under the supervision of an AP that is working with the interesrs of the trust in mind.
Once again I would say any AP worth their salt would be aware of exactly what they need to do and would cover their backs if nothing else, certainly the ones ive come into contact with would hit the roof if they new people where messing about with MGPS without permits etc.
rolleyes

Top
#5195 - 31/12/02 12:40 PM Re: Water in Medical Air
Tone Offline
Novice

Registered: 11/12/02
Posts: 12
Loc: London
Hi Karl

I agree that any work to the medical gas piped system should be only be undertaken under a permit to work. The point being debated is the QC after the necessary repair work has been performed. An authorised person generally hasn't got the skill or the knowledge to perform purity checks hence why they are performed by quality pharmacists in conjunction with people like AP's.

Top
Page 1 of 2 1 2 >



Moderator:  DaveC in Oz, KM, RoJo 
Sponsors
Press Releases
Who's Online
15 registered (Paul Robbins, Gordovan, DinhVu, Geoff Hannis, Huw, RoJo, KM, hieu, Lee S, Laiq, lm1253, Jameson, DaveC in Oz, GeorgeK, Hasan), 312 Guests and 8 Spiders online.
Key: Admin, Global Mod, Mod
May
Su M Tu W Th F Sa
1 2 3 4 5
6 7 8 9 10 11 12
13 14 15 16 17 18 19
20 21 22 23 24 25 26
27 28 29 30 31
Events