#5186 - 27/12/02 06:53 PM
Water in Medical Air
|
Newbie
Registered: 02/07/02
Posts: 2
Loc: Great Western Hospital Swindon
|
Two weeks ago, due to a breakdown in the dryers on the medical air compressor, water condensate passed through our piped medical gas system into a ventilator. Just as we were moving hospitals too. Exciting times. Have any other hospitals experienced this and what action did you take to protect your patients and make sure the vents were safe to use in the long term?
|
|
Top
|
|
|
|
#5188 - 29/12/02 05:05 AM
Re: Water in Medical Air
|
Expert
Registered: 23/10/02
Posts: 130
|
Hi Guy's, probably sounds a silly question but did you guys involve a pharmacist for a QC check ? 
_________________________
A problem is a gift to be unwrapped :rolleyes:
|
|
Top
|
|
|
|
#5193 - 31/12/02 10:13 AM
Re: Water in Medical Air
|
Expert
Registered: 23/10/02
Posts: 130
|
John, I am somewhat confused here,  who is your hospital “Authorized Person” [AP], I attended the “Competent Persons” course in Eastwood Park, this is a internationally recognized standard,  even out here in the gulf,  and my manual stipulates that any “OPEN WORK” done on “Any” medical gas pipeline system must have a QC on completion of the work and before use. Obviously under AP supervision. John, you can call eastwood park for more details, Regards Ffloyd
_________________________
A problem is a gift to be unwrapped :rolleyes:
|
|
Top
|
|
|
|
#5194 - 31/12/02 12:07 PM
Re: Water in Medical Air
|
Philosopher
Registered: 30/08/01
Posts: 729
Loc: LHCH
|
Sorry chaps, you can go on as many courses as you like. The facts are that the type of work mentioned should be under the supervision of an AP that is working with the interesrs of the trust in mind. Once again I would say any AP worth their salt would be aware of exactly what they need to do and would cover their backs if nothing else, certainly the ones ive come into contact with would hit the roof if they new people where messing about with MGPS without permits etc. 
|
|
Top
|
|
|
|
#5197 - 02/01/03 12:00 PM
Re: Water in Medical Air
|
Philosopher
Registered: 11/07/00
Posts: 969
Loc: Stockport, Cheshire, England
|
Do you want to put money on that ?
I've spoken to Estates staff who think that air is piped round the hospital just to drive tools ! They seem to think that if it isn't Oxygen, then it isn't given to a patient !
_________________________
Today is the day you worried about yesterday - and all is well !
|
|
Top
|
|
|
|
#5198 - 02/01/03 07:12 PM
Re: Water in Medical Air
|
Newbie
Registered: 02/07/02
Posts: 2
Loc: Great Western Hospital Swindon
|
Thanks you all. Our Pharmacist is principally responsible for medical gases and has been involved and is very helpful. We immediately started to run our vents off bottled Air and there was much lugging of G size cylinders. We added an extra HME filter to the inspiratory limb of the patient circuit and Microbiology took samples of the fluid coming through the air lines. (They didn't grow anything). Our AP-authorised person- was around commissioning the new site and was able to offer plenty of advice about the gas system. After the compressor was repaired, they flushed the system for some time to get the humidity down. However, the clinicians are still nervous about what may be left inside the vents and the risk to already compromised ICU patients. The Public Health labs at Colindale have recommended a strip down and we await the manufacturers price with interest. In the meantime we are still running those vents with extra HME filters (a fairly cheap precaution). Of course the Medical Gas system is provided and run by our PFI partner and there may be some contractual argy bargy. I will keep you up to date with the outcome.
|
|
Top
|
|
|
|
|
|
15 registered (Paul Robbins, Gordovan, DinhVu, Geoff Hannis, Huw, RoJo, KM, hieu, Lee S, Laiq, lm1253, Jameson, DaveC in Oz, GeorgeK, Hasan),
312
Guests and
8
Spiders online. |
|
Key:
Admin,
Global Mod,
Mod
|
|
|
|
|
1
|
2
|
3
|
4
|
5
|
|
6
|
7
|
8
|
9
|
10
|
11
|
12
|
|
13
|
14
|
15
|
16
|
17
|
18
|
19
|
|
20
|
21
|
22
|
23
|
24
|
25
|
26
|
|
27
|
28
|
29
|
30
|
31
|
|
|
|
|