Home | Articles | Products | Services | Jobs | Forum | Downloads | Press | Contact Us

Topic Options
Rate This Topic
#5762 - 03/11/06 08:46 PM Fluke Ventilator Analizer
Eliane_Brazil Offline
Newbie

Registered: 19/09/06
Posts: 1
Dear Friends,

We are looking for a ventilator analyser.
Do you know the Fluke ESA601?
We can realize all IEC 601-1 and 601-2-12 with this analyser?
Thanks in advance,

Eliane Apolinario
Biomedical Engineer
Brazil

Top
#5763 - 04/11/06 02:06 PM Re: Fluke Ventilator Analizer
Geoff Hannis Offline
Super Hero

Registered: 12/02/04
Posts: 10300
Loc: the path less trodden
Looks like the ESA601 handles EN 60601-1.

Regarding a ventilator analyzer, are you looking at the VT Mobile or VT Plus, I wonder?

But the hand-held Rigel 355 also looks interesting.

Better check-out this lot too! smile

Top
#5764 - 05/11/06 11:18 PM Re: Fluke Ventilator Analizer
mantunes Offline
Visionary

Registered: 28/07/05
Posts: 37
Loc: São Paulo
Ventilators analyzers cannot, strictly speaking, perform all the tests of IEC 60601-1-2 correctly - the volume test accoording to the standard has to be made with lung simulators of the rigid conteiner type, and all analyzers that i know, when the have lung simulators, use the pneumatic type (springs) which are not fit for the tests. Anyway, they are made to be used in hospitals, which do not need the metrological accuracy needed in test houses - labs, for which the standard is made. So i think you can use safely any analyzer, if you pay attention to theses aspects. Also, take care with the tests of IEC 60601-2-12. A lot of them have to be adapted to work on a particular ventilator, this meaning that the analyzer alone won´t perform the tests, but it´ll need your knowledge of the ventilator and the test mechanics to perform it correctly.
_________________________
Medical Electrical Equipment Information Blog - Recently updatet! :-)
http://www.electromedicalinfo.com

Top



Moderator:  DaveC in Oz, KM, RoJo 
Sponsors
Press Releases
Who's Online
16 registered (Ciaran McGuinness, Pavel, nerobot, Max, CHJ, Paul Robbins, biomedbill, Timsta, Laiq, Charles Valencia, billy11, Hulk, bcarlisle, Huw, Paul Owens, Lee S), 370 Guests and 7 Spiders online.
Key: Admin, Global Mod, Mod
May
Su M Tu W Th F Sa
1 2 3 4 5
6 7 8 9 10 11 12
13 14 15 16 17 18 19
20 21 22 23 24 25 26
27 28 29 30 31
Events
Appropriate Healthcare Technologies for Developing
Welch Allyn - Completing the Picture 2012