#58257 - 22/09/11 02:04 PM
Centrifuge lid interlock
|
Adept
Registered: 03/05/02
Posts: 88
Loc: York District Hospital
|
Does anyone know anything about requirements for lid interlocks on centrifuges? We’ve had a small centrifuge delivered for commissioning. The lid can be opened when it’s still running – though the power is turned off by the action of opening the lid. The supplier says a label warning users not to open the lid while it’s running is sufficient. I may be being over cautious but it does not seem right to me. Thanks Graham
|
|
Top
|
|
|
|
#58267 - 22/09/11 06:01 PM
Re: Centrifuge lid interlock
[Re: Graham Barnes]
|
Super Hero
Registered: 12/02/04
Posts: 10300
Loc: the path less trodden
|
I wouldn't be happy accepting that either. See ECRI (1992). There will be an IEC "Particular Requirements" Standard covering this (and no doubt someone will let us know the reference) which may talk about speed and diameter etc., but in my opinion, any modern (new) centrifuge should have a true "zero RPM" lid-locking system. That is, the interlock should keep the lid locked at all times during rotation (even during power failure) and require that the rotor be at rest before the lid can be unlocked. Is this thing from a well-known manufacturer ... or some "back-door" *import?  See also this thread from the depths of the archives. * Does it carry a CE mark?
|
|
Top
|
|
|
|
#58284 - 23/09/11 12:59 PM
Re: Centrifuge lid interlock
[Re: Graham Barnes]
|
Savant
Registered: 11/04/02
Posts: 122
Loc: Up North
|
Hi there. We have had this problem too. We had a rep come in to the hospital direct to the Breast Surgery team selling her centrifuge. It has no interlock, and the same excuse was given - the label is sufficient. She also tried to say that no other centrifuge was acceptable because the syringes being spun in the thing would be compromised (the sleeves they sit in could get contaminated). Anyway, I asked our senior scientist in the Pathology team to help us, and his advice was that the centrifuge was unsuitable. I can do a bit of digging and see what machine we got in the end, but he was most supportive. He said they had had the same problem with the same company before, and the result had been the same; they bought a different machine. The rep wasn't happy, but in my book that's just tough. She should have come through the right channels. Also, having been asked to bring the machine in through Medical Physics and agreeing to do so, she didn't. I caught her out by asking the senior ODA if he had seen her, and he said she was in his theatre as we spoke, with him using the machine... not his fault, and he was most upset that she was lying to him and using him to get her own way. If I had my way, she would now be banned from selling to our Trust. Among all this, there were no decontamination instructions or documentation provided, which is essential. The procedure this machine was being used to support was quite interesting. Apparently after cutting out the cancerous breast tissue and rebuilding the breast, the wounds heal best if you can inject some of the fat tissue under the stitched up skin. There must be no blood in the fat (hence using the centrifuge) but it gives a really good finish to what would otherwise be pretty grim scars. They heal flatter and cleaner. It is nice to think that such a small idea has such big benefits.
|
|
Top
|
|
|
|
#58290 - 23/09/11 03:33 PM
Re: Centrifuge lid interlock
[Re: Clare Walsh]
|
Super Hero
Registered: 12/02/04
Posts: 10300
Loc: the path less trodden
|
If I had my way, she would now be banned from selling to our Trust. Bravo! Clare. But why stop there? She should be banned from doing business across the whole NHS!  I've seen it a thousand times (both here and overseas) ... pretty faces (UK) or shiny suits ("out there") sneaking into the back corridors of the theatres etc. endeavouring to flog their [censored]. If the stuff was any good they would be proud to pass it by the biomed shop. I usually send them packing, with a flea in their ear, or worse (... better?).
|
|
Top
|
|
|
|
#58301 - 24/09/11 09:02 AM
Re: Centrifuge lid interlock
[Re: Graham Barnes]
|
Hero
Registered: 08/07/02
Posts: 1395
Loc: Temporarily in "The Smoke" but...
|
And they are often breaking Trust policies... If you look carefully at most procurement policies reps are only allowed in after having contacted the purchasing department and sought permission.* In a previous job I asked our head of precurement to write to the company MD saying we would consider banning a rep if he continued sneaking in - no rep=no potential for any sales. It stopped him. RoJo
* Though this is often subcontracted to Clinical Engineering etc by mutual agreement and to save the purchasing department all the hassle.
_________________________
Only trying to help and spread the word
|
|
Top
|
|
|
|
#58319 - 26/09/11 09:38 AM
Re: Centrifuge lid interlock
[Re: Graham Barnes]
|
Savant
Registered: 11/04/02
Posts: 122
Loc: Up North
|
We have just such a policy. Unfortunately, Supplies is listed as their first point of contact, and they might not realise the full implications of the reps patter. This directive is so that reps cannot sell to the Trust while there is a contract in place, so Supplies is the best place on that score, but as to suitability and so forth - no. We have used this clause to ban a rep before. It was only a temporary ban, but it made the point. Reps attempting to get advantageous information out of Trust staff by lying and claiming to have permission for that information when they expressly didn't - it does not go down well! On the centrifuge point - we ended up with the Heraeus Labofuge 200 because it has a lid interlock. The rep didn't like it, so she was promoting a different one instead although the company sold both. Graham has PM'ed me because we may have to share info that is best not public, but if you have similar troubles feel free to PM me.
|
|
Top
|
|
|
|
#58320 - 26/09/11 09:46 AM
Re: Centrifuge lid interlock
[Re: Clare Walsh]
|
Super Hero
Registered: 12/02/04
Posts: 10300
Loc: the path less trodden
|
Not going to "name and shame" then, Clare!  But (meanwhile) can you disclose if the Heraeus was purchased from the same company (that the rep worked for)? And, what was it the rep didn't like:- the *Heraeus, or that her commission was being threatened? * What other wonderful features could the other machine have had, I wonder? 
|
|
Top
|
|
|
|
|
|
15 registered (Pavel, Huw, Ciaran McGuinness, nerobot, Max, CHJ, Paul Robbins, biomedbill, Timsta, Laiq, Charles Valencia, billy11, Hulk, bcarlisle, Paul Owens),
377
Guests and
7
Spiders online. |
|
Key:
Admin,
Global Mod,
Mod
|
|
|
|
|
1
|
2
|
3
|
4
|
5
|
|
6
|
7
|
8
|
9
|
10
|
11
|
12
|
|
13
|
14
|
15
|
16
|
17
|
18
|
19
|
|
20
|
21
|
22
|
23
|
24
|
25
|
26
|
|
27
|
28
|
29
|
30
|
31
|
|
|
|
|