#59161 - 20/11/11 03:24 PM Microscope Olympus CX30
r0ni Offline
Novice

Registered: 02/05/11
Posts: 14
Loc: Kuwait
Hello guys, any idea for microscope olympus cx30. there's no 6V supply from the bulb and the bulb was ok. Any one have a service manual, pls. send it to r0ni.madz@yahoo.com.

Many Thanks.

Top
#59237 - 23/11/11 01:35 PM Re: Microscope Olympus CX30 [Re: r0ni]
Charles Valencia Online   content
Adept

Registered: 05/11/10
Posts: 90
Loc: Philippines
dear friend,

you test the bulb terminal while the unit is power on?
Does your bulb circuit is switching type or simply a transformer?
if it is switching type, try to test the transistors...or the small switching transformer...
how did you determine that bulb is good?
sometimes halogen bulbs have still continuity even if it is busted...
_________________________
There is no work worth dying for.

Top
#59466 - 12/12/11 10:21 AM Re: Microscope Olympus CX30 [Re: r0ni]
r0ni Offline
Novice

Registered: 02/05/11
Posts: 14
Loc: Kuwait
Charles,

It's a switching type, there's no output in the bulb terminal.
I suspected the bipolar IC FA5311B, switching power supply control and a mosfet FS5KM 16A high swtiching. The bulb is still good i exchanged with the other microscope.

Top
#59472 - 13/12/11 01:30 AM Re: Microscope Olympus CX30 [Re: r0ni]
Charles Valencia Online   content
Adept

Registered: 05/11/10
Posts: 90
Loc: Philippines
Fuse is still busted when you turn on the power supply?...if yes, mosfet should be the problem. if fuse is ok upon power on but still no output, switching Ic might be the problem...But still check the connection between your board and halogen lamp, because that in that kind of circuit, usually there is no voltage detected in halogen lamp terminal if there is no load attached to it, meaning your connections must be good and your lamp must be ok before you could detect a voltage to lamp's terminal...
_________________________
There is no work worth dying for.

Top
#59473 - 13/12/11 01:39 AM Re: Microscope Olympus CX30 [Re: r0ni]
Charles Valencia Online   content
Adept

Registered: 05/11/10
Posts: 90
Loc: Philippines
The connection for halogen lamp is directly connected to output of switching transformer right?
_________________________
There is no work worth dying for.

Top
#59736 - 10/01/12 12:28 PM Re: Microscope Olympus CX30 [Re: r0ni]
r0ni Offline
Novice

Registered: 02/05/11
Posts: 14
Loc: Kuwait
I found out that the problem was on IC switching power supply,
no output 12V in pins 4 & 5. The bad thing, it's not available in local market here in kuwait. Need to order in RS electronics.

Top
#59795 - 19/01/12 01:15 AM Re: Microscope Olympus CX30 [Re: r0ni]
Charles Valencia Online   content
Adept

Registered: 05/11/10
Posts: 90
Loc: Philippines
try to find some replacement...maybe there are some available locally...
_________________________
There is no work worth dying for.

Top
#59796 - 19/01/12 01:17 AM Re: Microscope Olympus CX30 [Re: r0ni]
Charles Valencia Online   content
Adept

Registered: 05/11/10
Posts: 90
Loc: Philippines
may i know the part number of that particular IC?...maybe I could Help...
_________________________
There is no work worth dying for.

Top



Moderator:  DaveC in Oz, KM, RoJo 
Sponsors
Press Releases
Who's Online
15 registered (Huw, Kawasaki, Pavel, Ciaran McGuinness, nerobot, Max, CHJ, Paul Robbins, biomedbill, Timsta, Laiq, Charles Valencia, billy11, Hulk, bcarlisle), 378 Guests and 6 Spiders online.
Key: Admin, Global Mod, Mod
May
Su M Tu W Th F Sa
1 2 3 4 5
6 7 8 9 10 11 12
13 14 15 16 17 18 19
20 21 22 23 24 25 26
27 28 29 30 31
Events
Appropriate Healthcare Technologies for Developing
Welch Allyn - Completing the Picture 2012