#59938 - 04/02/12 01:49 AM Micro-Lite PTS-68
Mofe Offline
Mentor

Registered: 01/04/05
Posts: 184
Loc: Portugal
Dear friends,

I have a problem with phototherapy Micro-lite PTS-68, appears the error 05. Sameone can help me to know what means that error?
I need the service manula too. Can you help me?

Regards,

Oliveira Eugénio

Top
#59944 - 05/02/12 05:10 PM Re: Micro-Lite PTS-68 [Re: Mofe]
fmic.biomedical Offline
Dreamer

Registered: 26/07/07
Posts: 21
Loc: Kabul , Haiti , France
hello,
this phototherapy unit is quite simple .
check the 3 lamps , the fan and the lamp timer .


When the unit is switched on, by briefly
depressing the on/standby switch, it goes through a
self-test process. The segments of the display are
checked, the software version is displayed and
then a time is shown. During this start up
sequence, the three halogen lamps are lit in turn.
After the start-up procedure described above, the
timer display shows the length of time since it was
last reset and is designed to be used as a treatment
timer. Pressing and holding the on/standby switch
while the lights are on resets this timer. If the
on/standby switch is held in during the start up
procedure, the Bulb Life Timer is displayed.
Releasing and then holding in the on/standby
switch again resets the Bulb Life Timer; the
treatment timer is unaffected.

Top



Moderator:  DaveC in Oz, KM, RoJo 
Sponsors
Press Releases
Who's Online
14 registered (biomedbill, Pavel, Hasan, webbie, Mark Radbourne, Huw, Kawasaki, Ciaran McGuinness, nerobot, Max, CHJ, Paul Robbins, Timsta, Laiq), 383 Guests and 8 Spiders online.
Key: Admin, Global Mod, Mod
May
Su M Tu W Th F Sa
1 2 3 4 5
6 7 8 9 10 11 12
13 14 15 16 17 18 19
20 21 22 23 24 25 26
27 28 29 30 31
Events
Appropriate Healthcare Technologies for Developing
Welch Allyn - Completing the Picture 2012