Home | Articles | Products | Services | Jobs | Forum | Downloads | Press | Contact Us

Topic Options
Rate This Topic
#59805 - 19/01/12 01:41 PM SLE2000 PIP HIGH ALARM
Laiq Online   content
Master

Registered: 24/04/05
Posts: 322
Loc: Karachi
Dear all
I am having problem in one of SLE2000 infant ventilator.
I have calibrated it electronically & phneumatically.
Problem is that it is showing PIP HIGH alarm continously.
Also by changing the alarm setting of PIP control the ventilator is making unnecessary PEEP of approx 10cmH2O..
Can some one help me out here what i should do & look for.
Waiting for quick reply.
regards

Top
#59948 - 07/02/12 09:07 AM Re: SLE2000 PIP HIGH ALARM [Re: Laiq]
Charles Valencia Online   content
Adept

Registered: 05/11/10
Posts: 90
Loc: Philippines
in what mode are you working pal? as far as i know, PIP is used in Support ventilation mode and PEEP is in Assist /Control Mode...
_________________________
There is no work worth dying for.

Top
#59960 - 09/02/12 10:33 AM Re: SLE2000 PIP HIGH ALARM [Re: Laiq]
obiwan Offline
Scholar

Registered: 08/10/07
Posts: 71
Loc: uk
Not sure what you are asking, what mode are you in?
you have to remember with this vent you set up the CPAP/PEEP first with the pressure display set to MIN, i would suggest in cpap mode, then you set up PIP with pressure display set to max in CMV is best mode. Remember after setup, if PEEP pressure is adjusted this the will be reflecte on you PIP so if you have a Peep of 4 and a pip of 20 and you then adjust your peep pressure to 6 you will automaticaly have a pip of 22.

when you set the pressures up make sure you min alarm is all the way down and your max is all the way up and the cycle fail is above your peep but not your pip so you dont get the vent dumping thinking is an over pressure.

I think you are getting an high alarm what is causing the vent to dump the pip and thats why your getting that peep, not sure need more info

Top
#60041 - 14/02/12 01:04 PM Re: SLE2000 PIP HIGH ALARM [Re: Laiq]
Laiq Online   content
Master

Registered: 24/04/05
Posts: 322
Loc: Karachi
Dear charles & obiwan
Thanks a lot. Got it working. The PIP control pneumatic was faulty.
Met with SLE engineer in Arab health 2012, he helped us a lot.
Now only blender problem remaining, that is always whistle alarm even if both gases at same input pressure.
Thanks again for your replies.

Top
#60051 - 15/02/12 01:36 PM Re: SLE2000 PIP HIGH ALARM [Re: Laiq]
obiwan Offline
Scholar

Registered: 08/10/07
Posts: 71
Loc: uk
sounds like a sticking shuttle but that may have caused your fluctuating pressures as it will vent cass through the alarm

Top
#60052 - 15/02/12 01:37 PM Re: SLE2000 PIP HIGH ALARM [Re: Laiq]
obiwan Offline
Scholar

Registered: 08/10/07
Posts: 71
Loc: uk
cass = gass

Top



Moderator:  DaveC in Oz, KM, RoJo 
Sponsors
Press Releases
Who's Online
14 registered (biomedbill, Pavel, Hasan, webbie, Mark Radbourne, Huw, Kawasaki, Ciaran McGuinness, nerobot, Max, CHJ, Paul Robbins, Timsta, Laiq), 383 Guests and 8 Spiders online.
Key: Admin, Global Mod, Mod
May
Su M Tu W Th F Sa
1 2 3 4 5
6 7 8 9 10 11 12
13 14 15 16 17 18 19
20 21 22 23 24 25 26
27 28 29 30 31
Events
Appropriate Healthcare Technologies for Developing
Welch Allyn - Completing the Picture 2012