#60357 - 12/03/12 04:31 PM Multiparameter Monitor patient connectors
Kevin McGinn Offline
Dreamer

Registered: 02/03/01
Posts: 23
Loc: Liverpool,Royal Liverpool Hosp...
I've come across a patient monitor with apparently identical sockets for Spo2 and CO2 patient connections. ie yellow Lemo plastic sockets with a slightly different key. This doesn't seem right to me as there's a cross connection risk. I don't have access to IEC 60601-2-49:2001, could someone have a look or link me to an online resource?

Top
#60358 - 12/03/12 05:00 PM Re: Multiparameter Monitor patient connectors [Re: Kevin McGinn]
Geoff Hannis Offline
Super Hero

Registered: 12/02/04
Posts: 11286
Loc: the path less trodden

Manufacturer? Model? smile

If the key is different ... is it actually possible to fit the wrong plug to either socket?

And then ... can it even be done using "brute force and ignorance"?

Strange that the colours are the same, though. Are the plugs like that as well? think

Top
#60359 - 12/03/12 08:20 PM Re: Multiparameter Monitor patient connectors [Re: Kevin McGinn]
Ed SWM Offline
Mentor

Registered: 19/03/02
Posts: 150
Loc: bristol
Hi All,
sounds like the Contec CMS 7000 range (or similar derivative). They do indeed use yellow lemo style connectors but keyed differently (from memory they use respironics EtCo2 in an external module). From experience they would be quite difficult to cross connect as the internal layout is different. In fact if you come to think of the older style monitors,Datascope Passports for example, they used standard 6 pin style connectors for both ECG and IBP (both black) but with the locating key in a different place so I wouldn't think there would be an issue with standards etc. Doesn't make good sense from a user perspective but sometimes thats nothing new!
Regards
Ed Bennett
Southwest Medical Ltd

Top
#60364 - 13/03/12 08:14 AM Re: Multiparameter Monitor patient connectors [Re: Ed SWM]
Geoff Hannis Offline
Super Hero

Registered: 12/02/04
Posts: 11286
Loc: the path less trodden

Such things are not unknown. I seem to recall H.M.E. and Kontron patient monitors both using green Lemo connectors for ECG. Same to look at, same keying ... interchangeable in fact.

A good idea you might say ... standardisation! Yes. Except that the "wrong" ECG lead(s) would not work in the "wrong" monitor due (if I remember rightly) to them being wired up differently. Not very handy if your hospital held both types of monitor. frown

Top
#60391 - 14/03/12 09:06 AM Re: Multiparameter Monitor patient connectors [Re: Kevin McGinn]
RoJo Offline
Hero

Registered: 08/07/02
Posts: 1653
Loc: Temporarily in "The Smoke" but...
If the key is different then the socket is different. Unfortunately you cannot take in to account common sense (such as making it a different colour) or (Russian shot putter) nurses making it fit.
Quite often you have the gut feeling "this is wrong" but by the letter of the rules it is not. This is where pre-purchase assessment by knowledgeable, technical staff cames to the fore.
Which it sounds like you are doing.
As for the copy of 60601. There are plenty of links to places where you can purchase it such as the BSI Shop , but I suspect you were hoping for a free copy.
Robert
_________________________
Only trying to help and spread the word

Top
#60404 - 14/03/12 04:36 PM Re: Multiparameter Monitor patient connectors [Re: Kevin McGinn]
Kevin McGinn Offline
Dreamer

Registered: 02/03/01
Posts: 23
Loc: Liverpool,Royal Liverpool Hosp...
This is a Daray L505 monitor. Thanks RoJo, I was hoping someone could have a glance at the standard...if it states that the key is all that matters then so be it. However it is possible insert it in to the wrong socket despite the key (without force), although it doesn't fully mate - not ideal.
Yes, Geoff and Ed, common sense is in short supply...except on this site!

Top
#60406 - 14/03/12 05:17 PM Re: Multiparameter Monitor patient connectors [Re: Kevin McGinn]
Geoff Hannis Offline
Super Hero

Registered: 12/02/04
Posts: 11286
Loc: the path less trodden

Yet another Made in China clone then. frown

Top
#60447 - 19/03/12 07:52 AM Re: Multiparameter Monitor patient connectors [Re: Kevin McGinn]
Laiq Offline
Master

Registered: 24/04/05
Posts: 330
Loc: Karachi
Geoff you are right, but good thing from MADE IN CHINA CLONES that if you do complain to them, they will upgrade & remove the faults.
Hope if they will complain to the manufacturer, they will rectify that problem. I am having good experience from chineese manufacturers. They listen & modify fast. Best of luck. :-)

Top
#60448 - 19/03/12 08:02 AM Re: Multiparameter Monitor patient connectors [Re: Laiq]
Geoff Hannis Offline
Super Hero

Registered: 12/02/04
Posts: 11286
Loc: the path less trodden

Fair enough. But I would much prefer that the kit was properly designed in the first place.

Yeah ... great at copying, but not so hot when it comes to innovation!

After all, getting the colour of connectors right (standardised, whatever) is hardly Rocket Science, is it? think

Actually, when you think about it, what you're saying is not so fantastic ... as what you end up with is "similar" kit scattered across the world in various states of "revision".

If they did their research at the onset, then designed and manufactured accordingly, it would spare customers (and patients) from being perpetual "Guinea Pigs"!

But, to be honest, many manufacturers are like that these days (across all sectors) ... in their rush to get to market, they end up dropping the ball! whistle

Top



Moderator:  DaveC in Oz, KM, RoJo 

Sponsors

Press Releases
Who's Online
2 registered (John Stewart, DaveC in Oz), 203 Guests and 8 Spiders online.
Key: Admin, Global Mod, Mod
Events
17th National Biomedical & Clinical Engineering Co
Welch Allyn - 12th Annual "Completing the Picture
May
Su M Tu W Th F Sa
1 2 3 4
5 6 7 8 9 10 11
12 13 14 15 16 17 18
19 20 21 22 23 24 25
26 27 28 29 30 31
Newest Members
Matyas Liptovszk, Luap, Bruno Ferreira, EBMEnathan, ferdigia
7240 Registered Users
Featured Member
Registered: 09/02/13
Posts: 9
Today's Birthdays
paulesc
(Views)Popular Topics
Jokes 2 (son of Jokes!!!) 1134742
Classic Computers 390059
Magic Kingdom Biomed Expats 323505
The VRCT Website 317410
VRCT - Who's Re-Newed? 205971
Trainee EBME Technician Salary? 201809
VRCT renewal 185937
Jokes!! 181877
Sharing the info. 173694
DIRTY equipment 167138