#60881 - 20/04/12 11:38 AM Maquet Alphamaxx Operating Table - problem list
Paul Owens Online   content
Scholar

Registered: 17/03/03
Posts: 51
Loc: Bristol
Looking for feedback from those who look after this bariatric table.
We've had them in use here in Bristol since 2005 & we've had a succession of recurring problems with them. Of 4 tables in use we had to replace 5 main PCBs before Maquet admitted there was a problem with a component & we eventually had all the boards changed to the same version but only after we had forked out the money for the first 5.
The next major recurring problem relates to the locking mechanism on the leg extensions - bent shafts & bent bearing pins. As the shafts incorporate RFID encoders both to tell the table that it is a solid section & which is left/right it's vital that proper secure locking is achieved else you can end up with a split leg section. After that it's the telescopic section, handswitch & handswitch cable breaking & we have our hands full. The other serious & major probem is it's possible to dislodge a position sensor cable as it's not guarded & then the table doesn't know where it is & the result can be specatular. Any other users have these problems

Top
#60887 - 21/04/12 11:40 AM Re: Maquet Alphamaxx Operating Table - problem list [Re: Paul Owens]
Pablo orra Offline
Newbie

Registered: 04/02/12
Posts: 7
Loc: Italy
I have a ALPHAMAXX built in 2008
I hade problems on power board (hydraulic pump always ON)

Top
#60906 - 24/04/12 11:04 AM Re: Maquet Alphamaxx Operating Table - problem list [Re: Paul Owens]
Paul Owens Online   content
Scholar

Registered: 17/03/03
Posts: 51
Loc: Bristol
Hello Pablo.
Hydraulic pump constantly running is one of the symptoms of a failed power board - worry when this happens in the middle of an operation is that the operators continue working, the pump overheats & with all that hydraulic oil we wonder what the danger is of a fire. You can't cut the power to the pump without lifting the plinth covers & that's not possible with a patient on the table.

Top
#60912 - 25/04/12 10:45 AM Re: Maquet Alphamaxx Operating Table - problem list [Re: Paul Owens]
Geoff Hannis Online   content
Super Hero

Registered: 12/02/04
Posts: 11278
Loc: the path less trodden

Yet another example of brilliant design, then, by the sound of it. frown

Sometimes (often, usually in fact) the old - proven - ways are the best.

Top



Moderator:  DaveC in Oz, KM, RoJo 

Sponsors

Press Releases
Who's Online
17 registered (Paul Allum, Huw, Paul Owens, tomiteus, Chris Watts, RoJo, Barney, robertenglp, Geoff Hannis, Aymen, Paul.Pichur, Gee Bee, titan, Lee S, SLamosa, PauloSousa, BK), 382 Guests and 18 Spiders online.
Key: Admin, Global Mod, Mod
Events
17th National Biomedical & Clinical Engineering Co
Welch Allyn - 12th Annual "Completing the Picture
May
Su M Tu W Th F Sa
1 2 3 4
5 6 7 8 9 10 11
12 13 14 15 16 17 18
19 20 21 22 23 24 25
26 27 28 29 30 31
Newest Members
Bruno Ferreira, EBMEnathan, ferdigia, paqste01, alexmac
7238 Registered Users
Featured Member
BSM
BSM
Registered: 20/07/01
Posts: 223
Today's Birthdays
Jandre, Lawrie
(Views)Popular Topics
Jokes 2 (son of Jokes!!!) 1133524
Classic Computers 388623
Magic Kingdom Biomed Expats 323075
The VRCT Website 317279
VRCT - Who's Re-Newed? 205925
Trainee EBME Technician Salary? 201701
VRCT renewal 185656
Jokes!! 181733
Sharing the info. 173436
DIRTY equipment 167035