#61053 - 11/05/12 06:58 AM Calibration Laboratory
Neil Porter Online   content
Hero

Registered: 23/02/09
Posts: 1635
Loc: Jeddah, Saudi Arabia
I would like to get details of which equipment I need, the pricing and the supplier of calibration equipment for medical devices. I have 'googled' and all I get is "send your equipment to us" type messages. I am trying to put a proposal together so any help would be welcome, thanking you in advance.
_________________________
Stress is for other people

Top
#61054 - 11/05/12 09:57 AM Re: Calibration Laboratory [Re: Neil Porter]
Geoff Hannis Online   content
Super Hero

Registered: 12/02/04
Posts: 11267
Loc: the path less trodden

Thinking of having a go yourself, then, Neil? think

Expensive!

My first thoughts are:- whatever instruments you go for, they will all themselves need to be calibrated against a "higher level" - eg, ultimately traceable to the National Physical Laboratory or whatever.

Do you have such a body out there? And who runs (owns) it? MODA, I wonder?

And what (if anything) are the other Big Healthcare Providers out there doing in this regard?

Off the top of my head, I would have thought that one medical equipment calibration lab should be able to serve the whole Kingdom ... with, perhaps, just "cal checks" (in spec/out of spec) checks being carried out locally.

Top
#61058 - 12/05/12 05:38 AM Re: Calibration Laboratory [Re: Neil Porter]
Neil Porter Online   content
Hero

Registered: 23/02/09
Posts: 1635
Loc: Jeddah, Saudi Arabia
All hospitals, companies are now chasing accreditation, there are standards organizations including weights and measures. The companies are sending their instruments abroad for calibration and this is expensive, there are people here willing to invest in the creation of a calibration laboratory and the business potential is huge!

Three labs would be perfect, Damman, Jeddah and Riyadh.

Geoff do you have any clues as to where I can find the equipment?
_________________________
Stress is for other people

Top
#61060 - 12/05/12 08:37 AM Re: Calibration Laboratory [Re: Neil Porter]
Geoff Hannis Online   content
Super Hero

Registered: 12/02/04
Posts: 11267
Loc: the path less trodden

What exactly are you hoping to do, Neil ... set up a lab to carry out annual calibrations on biomed test kit? think

If you were to approach one the companies already doing that in the UK (or USA, wherever) with the "promise" of a Joint Venture, or some other incentive, I'm sure you would be able to obtain all the information you require.

For instance:- you could try Young's. You could also speak to Rigel ... and Ultramedic.

Otherwise, you're going to have to sit down and work out your own calibration methods. And from that, the equipment you will need. Much more fun doing it that way, I would have thought! smile

Top
#61063 - 12/05/12 11:06 AM Re: Calibration Laboratory [Re: Neil Porter]
Neil Porter Online   content
Hero

Registered: 23/02/09
Posts: 1635
Loc: Jeddah, Saudi Arabia
Not only biomedical test equipment, but all measuring equipment. The idea is to set up an independent laboratory complete with certification, instead of sending the equipment abroad DO IT HERE.
_________________________
Stress is for other people

Top
#61065 - 12/05/12 04:57 PM Re: Calibration Laboratory [Re: Neil Porter]
biomedical87 Offline
Scholar

Registered: 14/11/10
Posts: 68
Loc: samson
Mr porter its really a good idea for buisness as well as for hospitals and companies who needs an urgent calibration to set standards...
there is a variety of these equipments.but it will be expansive.

Top
#61067 - 12/05/12 07:13 PM Re: Calibration Laboratory [Re: Neil Porter]
Neil Porter Online   content
Hero

Registered: 23/02/09
Posts: 1635
Loc: Jeddah, Saudi Arabia
So Mr. Biomedical87 do you have any companies that I can approach?
_________________________
Stress is for other people

Top
#61068 - 12/05/12 07:17 PM Re: Calibration Laboratory [Re: Neil Porter]
Geoff Hannis Online   content
Super Hero

Registered: 12/02/04
Posts: 11267
Loc: the path less trodden

My bet is that someone is already doing it. As long ago as 1987 I remember some French company in Jubail doing all our biomed test kit.

If I were you I would check with the oil companies, Neil. Jubail, Yanbu for starters. ARAMCO must have some sort of instrument calibration laboratory, for sure.

Other than that, do you really want to upset the status quo (as in the various parties with a vested interest in keeping matters as they currently stand)? If my experiences "out there" are anything to go by, it's probably far better (if you want to survive) to keep your head down.

Why not just start off in a small way, and take it from there? If I were you I would begin by calibrating electrical safety testers, and maybe multimeters (plus voltmeters, and so forth) and see how you get on with that. smile

Top
#61069 - 12/05/12 08:30 PM Re: Calibration Laboratory [Re: Neil Porter]
Neil Porter Online   content
Hero

Registered: 23/02/09
Posts: 1635
Loc: Jeddah, Saudi Arabia
Geoff, this is not for me, I am getting the information for a company I am acting as a consultant. The more companies, hospitals we speak to the more we find out how much calibration is being done abroad!
_________________________
Stress is for other people

Top
#61070 - 12/05/12 08:34 PM Re: Calibration Laboratory [Re: Neil Porter]
Neil Porter Online   content
Hero

Registered: 23/02/09
Posts: 1635
Loc: Jeddah, Saudi Arabia
Geoff in the time you have been away a lot has changed.standards are improving and test there is accountability!
_________________________
Stress is for other people

Top
#61072 - 13/05/12 06:53 AM Re: Calibration Laboratory [Re: Neil Porter]
biomedical87 Offline
Scholar

Registered: 14/11/10
Posts: 68
Loc: samson
Mr.Geoff you are the most active person in the forum....

Mr.porter try this link...
http://www.netech.org/products.asp

Top
#61073 - 13/05/12 07:04 AM Re: Calibration Laboratory [Re: biomedical87]
Geoff Hannis Online   content
Super Hero

Registered: 12/02/04
Posts: 11267
Loc: the path less trodden

Off topic for a while (sorry Neil):- yes, the Netech range has really grown, and it now covers "almost everything" (in biomed terms).

But it looks like a lot of that stuff is bought in from elsewhere ... unless Netech is in fact the OEM (and others re-brand their stuff).

There again, the way things are these days, I often wonder who is the OEM of what (and for whom) - for most equipment! think

Top
#61074 - 13/05/12 07:48 AM Re: Calibration Laboratory [Re: Neil Porter]
Alan Ong Offline
Savant

Registered: 04/07/07
Posts: 103
Loc: Al Ain, UAE
Hi Neil,

Am based in Abu Dhabi UAE and we send our tools for calibration to Arabian Calibration and Dev. Rep. Co. LLC . Website: http://www.arabcal.com/

I've talked with our contact person there and she said that they also sell calibration equipment. If you are interested, her name is Maria and her phone number is +971554132136 .

Good luck on your project. smile

Top
#62408 - 25/10/12 01:35 AM Re: Calibration Laboratory [Re: Neil Porter]
Maczx636 Offline
Visionary

Registered: 29/05/12
Posts: 32
Loc: Cornwall, Devon, UK
Look any large engineering company ie BAE, BABCOCK, THALES, have a dedicated calibration centre, so as an independent small engineering company wishing to supply other firms with calibration requirements, all you have to is ensure that you can cal your kit with these firms, then calibrate your clients with this equipment. Don't think that you are restricted to purely medical engineering companies. Radio graphics, well seeing BAE do it on a daily basis on submarines, well you see where we are going, this is engineering, you are an engineer, sometimes look for a solution in the engineering world and not in the medical one.


Edited by Maczx636 (25/10/12 01:36 AM)
Edit Reason: Typo

Top
#62410 - 25/10/12 09:08 AM Re: Calibration Laboratory [Re: Maczx636]
Geoff Hannis Online   content
Super Hero

Registered: 12/02/04
Posts: 11267
Loc: the path less trodden

Neil is in Jeddah. I know Neil from "the Old Days", and can confirm his long experience, so no doubt he understands the principles of calibration (traceability, and all the rest). Kit out there still gets calibrated (by and large), but often the test kit (and sometimes the medical equipment itself) needs to be sent overseas to be re-certified.

Yes, there are industrial cal labs around "out there" (but not too many submarines). But I expect that Neil was looking for someone to offer him a list of test (calibration lab) kit, complete with suppliers' details, so he can have a go at setting up a cal lab in-house to serve the hospitals and clinics within his bailiwick.

So here we have a case where Neil is hoping for something that he has often complained about when replying to the posts of others ... that is, presenting him with a quick and easy answer!

As I may have observed before:- "if it was that easy, everyone would be doing it"! whistle

Top
Page 1 of 2 1 2 >



Moderator:  DaveC in Oz, KM, RoJo 

Sponsors

Press Releases
Who's Online
12 registered (Geoff Hannis, Marcelli, Neil Porter, tomiteus, San Garwe, Huw, JackFlash, Rob.Rsch, BK, Ed SWM, John Stewart, spectrumsteve), 337 Guests and 14 Spiders online.
Key: Admin, Global Mod, Mod
Events
17th National Biomedical & Clinical Engineering Co
Welch Allyn - 12th Annual "Completing the Picture
May
Su M Tu W Th F Sa
1 2 3 4
5 6 7 8 9 10 11
12 13 14 15 16 17 18
19 20 21 22 23 24 25
26 27 28 29 30 31
Newest Members
EBMEnathan, ferdigia, paqste01, alexmac, BrunoAraujo
7236 Registered Users
Featured Member
Registered: 25/04/13
Posts: 1
Today's Birthdays
bandaid, stefbois
(Views)Popular Topics
Jokes 2 (son of Jokes!!!) 1132072
Classic Computers 387859
Magic Kingdom Biomed Expats 322644
The VRCT Website 316952
VRCT - Who's Re-Newed? 205806
Trainee EBME Technician Salary? 201560
VRCT renewal 185473
Jokes!! 181303
Sharing the info. 173311
DIRTY equipment 166952