#61093 - 16/05/12 12:46 AM PHILIPS MP 70
Mofe Offline
Master

Registered: 01/04/05
Posts: 201
Loc: Portugal
Hi,

I have an monitor PHILPS MP 70 and sometimes he turns off alone. someone can help me?

Best regards

Top
#61105 - 16/05/12 04:33 PM Re: PHILIPS MP 70 [Re: Mofe]
Chandana S Offline
Scholar

Registered: 16/07/09
Posts: 70
Loc: Boeblingen, Germany
Hi Eugenio,

Do you have an IntelliVue Support Tool installed on your laptop? Are you able to extract the device error logs?
_________________________
Chandana Samaranayake MD

Chandana Samaranayake is an employee of Philips Healthcare.
All comments made on this forum are made on a personal capacity, and do not necessarily represent the views of my employer.


As per FDA guidelines on on-line communication, please note that the Intended Use of a Philips IntelliVue product is as stated in the Instructions for Use document.

Top
#61106 - 16/05/12 06:22 PM Re: PHILIPS MP 70 [Re: Chandana S]
Mofe Offline
Master

Registered: 01/04/05
Posts: 201
Loc: Portugal
Hi Chandana,

Please see the PM.

Best regards

Top
#61108 - 16/05/12 11:29 PM Re: PHILIPS MP 70 [Re: Mofe]
Mofe Offline
Master

Registered: 01/04/05
Posts: 201
Loc: Portugal
Hi Chandana,

In the error log the monitor I have:

Monitor: H

Ident: 17205

Cód: 20071

This appears nine times. Can you help me?

Best regards,

Top
#61116 - 17/05/12 01:37 PM Re: PHILIPS MP 70 [Re: Mofe]
Gavinb Offline
Newbie

Registered: 12/10/04
Posts: 3
Loc: Wrexham
Hi

We have had a similar problem ourselves first thing to do is replaced the module with another from a different MP70 and then place the unit in demo mode to confirm that it is the MP70. If the problem still occurs it looks like you will need to change the main board on MP70. Part no. M8050-68423

Hope this helps
Best regards

Top
#61147 - 22/05/12 08:00 AM Re: PHILIPS MP 70 [Re: Mofe]
Chandana S Offline
Scholar

Registered: 16/07/09
Posts: 70
Loc: Boeblingen, Germany
Hi Eugenio,

I checked with tech marketing, and they said that this error code indicates a failure of the Ethernet background check. The “H” in the monitor column indicates that the monitor performed a hot-start due to the failed background check. There are 3 possibilities:
A. If this monitor is connected to a network, it could be a network misconfiguration
B. If this monitor is connected to a network, it could be a problem with the local Ethernet port on the wall, or patching
C. It could be a failure of the monitor mainboard

To troubleshoot, the entire error log file (extracted as XML via the IntelliVue Support Tool) is required. However, if you don’t have access to a support tool, you can do the following troubleshooting:
To exclude A: if there are other monitors on the network showing the same behavior, it could be A. If this is the only monitor with this issue, A could be excluded.
To exclude B: Unmount this monitor and put another one in place (same model, with the same SW revision and configuration), and connect it to the network. If that monitor exhibits the same behavior, the problem is B.
If you exclude both A and B, then the problem is C, and the mainboard needs to be replaced.
_________________________
Chandana Samaranayake MD

Chandana Samaranayake is an employee of Philips Healthcare.
All comments made on this forum are made on a personal capacity, and do not necessarily represent the views of my employer.


As per FDA guidelines on on-line communication, please note that the Intended Use of a Philips IntelliVue product is as stated in the Instructions for Use document.

Top
#61826 - 04/08/12 01:07 AM Re: PHILIPS MP 70 [Re: Mofe]
Mofe Offline
Master

Registered: 01/04/05
Posts: 201
Loc: Portugal
Hi Chandana,

The problem is on Main Board.

Thank´s a lot.

Best Regards

Top



Moderator:  DaveC in Oz, KM, RoJo 

Sponsors

Press Releases
Who's Online
14 registered (Huw, Chris Watts, RoJo, Barney, robertenglp, Geoff Hannis, Aymen, Paul.Pichur, Gee Bee, titan, Lee S, SLamosa, PauloSousa, BK), 365 Guests and 14 Spiders online.
Key: Admin, Global Mod, Mod
Events
17th National Biomedical & Clinical Engineering Co
Welch Allyn - 12th Annual "Completing the Picture
May
Su M Tu W Th F Sa
1 2 3 4
5 6 7 8 9 10 11
12 13 14 15 16 17 18
19 20 21 22 23 24 25
26 27 28 29 30 31
Newest Members
Bruno Ferreira, EBMEnathan, ferdigia, paqste01, alexmac
7238 Registered Users
Featured Member
Registered: 27/02/07
Posts: 85
Today's Birthdays
Jandre, Lawrie
(Views)Popular Topics
Jokes 2 (son of Jokes!!!) 1133505
Classic Computers 388618
Magic Kingdom Biomed Expats 323071
The VRCT Website 317272
VRCT - Who's Re-Newed? 205924
Trainee EBME Technician Salary? 201698
VRCT renewal 185656
Jokes!! 181720
Sharing the info. 173436
DIRTY equipment 167033