#61281 - 08/06/12 12:29 AM Dry Side Calibration Of Omnifuse Syringe Driver
Simpo Offline
Dreamer

Registered: 10/12/02
Posts: 20
Loc: Cyprus
Gents

Can anyone throw some light on the method for carrying out the dry side calibration test on the Omnifuse syringe driver using the Omnifuse software and a Fluke IDA 4 Channel tester. I'm being asked to input an ambiguous measured value in KG?? and only have the standard IDA options of mmHg, cmH2O etc.

Any help appreciated

Paul

Top
#61284 - 08/06/12 09:49 AM Re: Dry Side Calibration Of Omnifuse Syringe Driver [Re: Simpo]
Paul Robbins Offline
Technologist

Registered: 16/10/03
Posts: 48
Loc: cambridgeshire england
Hate to be a party pooper but the short answer is you can not.

To complete the dry side occulsion pressure measurement you need to use the the approved pressurement tool, - a load cell (mounted to a dummy syringe) (dry side occulsion pressure is a direct measurement of the force being applied by the pump drive system to the syringe plunger) with the relevent control box that allows you to zero the load cell etc. This system then gives you the applied force values to enter into the omnifuse calibration program to allow the right values to be written back to the pump software.

The tool is available from Smiths but it is made by Force Measurement Systems in Glasgow, who may be able to help you.

Top
#61286 - 08/06/12 12:27 PM Re: Dry Side Calibration Of Omnifuse Syringe Driver [Re: Simpo]
bcarlisle Offline
Master

Registered: 16/08/07
Posts: 301
Loc: carlisle uk
The part No. is 0151-0287 load assembly kit.
For the size gauge set if you are calibrating 0151-0259 (sizes 8-35mm)

You will also need the software depends on your software issue in your pumps, contact the techs at smiths for the one you will need.

Top
#61287 - 08/06/12 01:01 PM Re: Dry Side Calibration Of Omnifuse Syringe Driver [Re: Simpo]
Geoff Hannis Offline
Super Hero

Registered: 12/02/04
Posts: 11286
Loc: the path less trodden

Top
#61295 - 08/06/12 08:57 PM Re: Dry Side Calibration Of Omnifuse Syringe Driver [Re: Simpo]
Simpo Offline
Dreamer

Registered: 10/12/02
Posts: 20
Loc: Cyprus
Many thanks chaps, lets see if Smiths will sell the test equipment now!

Top



Moderator:  DaveC in Oz, KM, RoJo 

Sponsors

Press Releases
Who's Online
5 registered (Huw, Matyas Liptovszk, Ed SWM, big steve, mimi), 233 Guests and 19 Spiders online.
Key: Admin, Global Mod, Mod
Events
17th National Biomedical & Clinical Engineering Co
Welch Allyn - 12th Annual "Completing the Picture
May
Su M Tu W Th F Sa
1 2 3 4
5 6 7 8 9 10 11
12 13 14 15 16 17 18
19 20 21 22 23 24 25
26 27 28 29 30 31
Newest Members
Matyas Liptovszk, Luap, Bruno Ferreira, EBMEnathan, ferdigia
7240 Registered Users
Featured Member
Registered: 30/05/08
Posts: 37
Today's Birthdays
paulesc
(Views)Popular Topics
Jokes 2 (son of Jokes!!!) 1134742
Classic Computers 390059
Magic Kingdom Biomed Expats 323505
The VRCT Website 317410
VRCT - Who's Re-Newed? 205971
Trainee EBME Technician Salary? 201809
VRCT renewal 185937
Jokes!! 181877
Sharing the info. 173694
DIRTY equipment 167138