Joined: Mar 2014
Posts: 11
Novice
|
OP
Novice
Joined: Mar 2014
Posts: 11 |
We have an old T2000 treadmill which I realise is well out of support but because it really hasn’t done that much work we would like to keep going. The attached CASE stress system is reporting “no communication with treadmill” which would appear to indicate a treadmill control board failure as the CASE including comms cable, works fine with another treadmill. Has anyone out there experienced these issues and found a solution?
|
|
|
|
Joined: Feb 2004
Posts: 14,654 Likes: 60
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,654 Likes: 60 |
I felt sure we had seen this one before ... but all I could find was this. Meanwhile, here is the manual. I remember reading somewhere that the "cal" procedure has been known to cure many ills with this unit; but not this particular problem, perhaps. But I don't suppose it's the emergency stop button ... ?
|
|
|
|
Joined: Mar 2014
Posts: 11
Novice
|
OP
Novice
Joined: Mar 2014
Posts: 11 |
So far we have deduced that; a. it is not the stop button b. performing the cal does not help although it seems to run that satisfactorily. c. the comms signals appear Ok
What we have found however is that attempting to manually control the treadmill via the 9pin D connector debug port we can control the elevation but not the speed. Applying a 0-5v signal to MANSPD does nothing.
|
|
|
|
Joined: Jun 2009
Posts: 796 Likes: 13
Philosopher
|
Philosopher
Joined: Jun 2009
Posts: 796 Likes: 13 |
Have you checked that you have the correct treadmill type and comms port selected in the system config/devices tab of the setup file (section 11 page 53 if you have the same manual as me)?
Dave (yippee, nearly beer o'clock... down under at least)
Last edited by DaveC in Oz; 11/03/16 4:53 AM.
Thoughts and information provided on this forum are mine and mine alone and do not necessarily reflect the policy of NSW Health. They may also be complete bollocks!!
|
|
|
|
Joined: Feb 2004
Posts: 14,654 Likes: 60
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,654 Likes: 60 |
Sounds logical. Also, which model (era) of CASE are we talking about here? ...works fine with another treadmill.
Can we assume a treadmill of the same type (T-2000)?
|
|
|
|
Joined: Mar 2014
Posts: 11
Novice
|
OP
Novice
Joined: Mar 2014
Posts: 11 |
The CASE is a P2 series ,2006. and no, the other treadmill that does work with this is a T2100.
Has anybody used the Debug port on the treadmill to manually control it. Just wondering if I'm not doing it right. Could control the elevation but not the speed.
Trying to avoid having to purchase another treadmill.
|
|
|
|
Joined: Mar 2014
Posts: 11
Novice
|
OP
Novice
Joined: Mar 2014
Posts: 11 |
We deduced that the Control Board must have been faulty so we purchased a replacement PCB and the treadmill now performs as it should. I believe it wasn't performing the speed calibration correctly which then caused a comms error as it couldn't respond to the commands from the exercise system.
All good now. thanks all for your input.
Ray
|
|
|
|
Joined: Feb 2004
Posts: 14,654 Likes: 60
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,654 Likes: 60 |
Nice chance to do a bit of component-level fault finding on the old board, then.
If you don't inspect ... don't expect.
|
|
|
|
Joined: Jun 2001
Posts: 246
Master
|
Master
Joined: Jun 2001
Posts: 246 |
We had this and a ge engineer said this was common if the speed is not completed every 6 months
ERRATIC MEANS STATIC SO BE ERRATIC AND NOT STATIC WE ARE ALL IN THE NHS AND THIS IS "ERRATIC AND STATIC"
|
|
|
|
Joined: Sep 2010
Posts: 407 Likes: 2
Sage
|
Sage
Joined: Sep 2010
Posts: 407 Likes: 2 |
What do you mean "speed is not completed every 6 months"?
..Calibration? :-)