OK ... here's some more "historical" (pedantic) trivia ...
"Officially", you were supposed to always carry out the Protective Earth Continuity test first, and were not to carry on to the more (potentially) "hazardous" tests (insulation resistance, and - if you insisted - the dreaded "Mains on Applied Parts",
etc.) if the first test failed.
Bearing in mind that I'm thinking of Class I equipment here, and at a time (back in an era) when most electromedical equipment had a fixed mains cable ... and a lot of it wasn't even fused at the input (apart from at the mains plug - in the UK, that is) like the kit is today.
In other words, you only moved on to insulation tests if the EUT was at least basically safe (protected by a known good protective conductor - or, if you must, "earth"). This advice was, of course, for your own protection.
This was the reason I always reckoned that the "Rigel tests" (that is, the Good Old Rigel
233) started off in the wrong order - in my opinion 1,2, then 5, then 3,4 ... 6
etc. may have been a better sequence ... especially for those biomeds who liked to just crank the knobs (and throw the switches) without seeming to bother too much about what they were actually trying to do. Yes; we had them back then, too (in fact a fair number, if I pause to recall).