|
|
#20395 - 03/05/02 01:51 PM
Re: Extension Leads & Patient Connected Equipment
|
Philosopher
Registered: 11/07/00
Posts: 970
Loc: Stockport, Cheshire, England
|
Yes, there was some advice from the MDA about using extension leads. It was quite a while ago and I can't remember the details, but try the MDA Web site - or quicker still, give them a ring !
Our electricians will not supply any extension leads for theatres, ICU, HDU, or anywhere there is a risk of liquid being spilt onto the sockets. We have been working with a local engineering firm who have produced a small stainless steel "pod" on castors with double 13 Amp sockets mounted on the vertical sides and protected from spillages by an overhanging top. They intend to market this in the very near future, so if you would like me to get them to contact you, send me an e-mail giving your telephone number.(and your name etc ofcourse !) It is CE marked, by the way.
_________________________
Today is the day you worried about yesterday - and all is well !
|
|
Top
|
|
|
|
#20396 - 07/05/02 01:40 PM
Re: Extension Leads & Patient Connected Equipment
|
Hero
Registered: 03/07/00
Posts: 1580
Loc: UK
|
Roy How much do they charge?
_________________________
Be Proactive and reactive.
|
|
Top
|
|
|
|
#20397 - 10/05/02 01:10 PM
Re: Extension Leads & Patient Connected Equipment
|
Newbie
Registered: 15/04/02
Posts: 2
Loc: Derby City General Hospital
|
You might like to look at BS EN 60601-1-1:2001 'Safety requirements for medical electrical systems'or you might not
|
|
Top
|
|
|
|
#20398 - 20/05/02 11:57 AM
Re: Extension Leads & Patient Connected Equipment
|
Philosopher
Registered: 11/07/00
Posts: 970
Loc: Stockport, Cheshire, England
|
John - I'll ask the guy to phone you and you can discuss your requirements !
Jon. Surely any requirements would have to be met to achieve the CE mark ? It's not our design - we've just suggested some refinements !
_________________________
Today is the day you worried about yesterday - and all is well !
|
|
Top
|
|
|
|
#20399 - 10/06/02 12:28 PM
Re: Extension Leads & Patient Connected Equipment
|
Newbie
Registered: 06/07/01
Posts: 9
Loc: Dundee
|
We have been supplying 'ELD' extension leads to our users for very many years. They are robust multiway extension leads incorporating an earth leakage monitoring circuit. The circuit alarms if more than 500 microamps flow through the earth wire. The demand for them is very high...
We have been unable to find a commercial partner over the years though several companies have expressed interest.
Mark
|
|
Top
|
|
|
|
#20400 - 11/06/02 07:27 PM
Re: Extension Leads & Patient Connected Equipment
|
Philosopher
Registered: 11/07/00
Posts: 970
Loc: Stockport, Cheshire, England
|
I'll ask our chap to contact you ! There's plenty of room in the stainless steel box to b uild in an alarm circuit. What a good idea ! :p
_________________________
Today is the day you worried about yesterday - and all is well !
|
|
Top
|
|
|
|
#20402 - 07/10/02 03:07 PM
Re: Extension Leads & Patient Connected Equipment
|
Philosopher
Registered: 11/07/00
Posts: 970
Loc: Stockport, Cheshire, England
|
Mark - if you haven't received an e-mail from me (sent 7th Oct) regarding the extension leads & alarm circuit, please can you contact me.
_________________________
Today is the day you worried about yesterday - and all is well !
|
|
Top
|
|
|
|
|
|
32 Registered (abi, Ali-G., Apollo11, Baldrick, bill_mcg, biomedbill, Blinkin Flip, BobA, Dicky, Ed SWM, Eddie, Gav, Geoff Hannis, Gerry, Graham Roberts, Huw, John Stewart, JoLee, Jonathan Wells, Kawasaki, keano, Ken, KM, Laiq, Panander, ray hill, Roger, RoJo, seppy, SUZUKI, Tony Dowman, TonyR),
and 114 guests online.
|
|
|