#21321 - 06/10/03 01:05 PM IEC leads control
leonius Offline
Savant

Registered: 17/06/03
Posts: 108
Loc: UK
How do all of you deal with the IEC 320 :rolleyes: leads being changed from machine to machine without knowlege of the EBME Dept.
The problem as i see it, as the leads are moved around the correct rated fuse in the lead is swapped. Also some leads are only two pin and therfore could get moved to a class 1 device. being a newbie this issue may aready have been discused in that case a link would be nice.

Top
#21322 - 06/10/03 01:42 PM Re: IEC leads control
Paul Balaam Offline
Expert

Registered: 14/06/01
Posts: 143
Loc: Whipps Cross University hospit...
Big problem!!!!

Class II leads can't be stuck into class I sockets 'cos the earth pins in the way.

But as far as swapping the leads, that’s a different matter. I can't see any way around the problem, but what we do here is to put a laminated label around the lead indicating the lead was test on that particular machine on that date, if the staff swap the lead, then that’s up to them.

I have sent Memo’s, spoken to the General Managers, emailed around the Trust not to swap leads, but still it happens. I’ve ended up sticking 3 amp fuses in all the leads apart from those that need a higher rating and if they are swapped and the fuse is popped, you can ask the question why was the lead changed?

Paul.

Top
#21323 - 06/10/03 02:10 PM Re: IEC leads control
Allan H Offline
Technologist

Registered: 30/09/02
Posts: 49
Loc: United Kingdom
Big problem, no easy solution. I would indicate though that the fuse should not be a problem since the fuse is there to protect the cable and not the appliance, so rate the fuse to suit the cable.

Top
#21324 - 06/10/03 02:36 PM Re: IEC leads control
Anonymous
Unregistered


0.75mm^2 up to 6A fused; 1.0mm^2 up to 10A fused; 1.5mm^2 up to 13A fused (for detachable IEC) isn't it? Depending upon the rating of the plug-top and IEC connector, of course. Any comments welcomed.

Top
#21325 - 06/10/03 07:17 PM Re: IEC leads control
grahamedriver Offline
Mentor

Registered: 18/09/00
Posts: 152
Loc: Rotherham,RDGH,UK
Fuse is there to protect cable and provided it's right, it does.

Aren't detachable mains made detachable so they can be? Safely? What's the issue?

Top
#21326 - 06/10/03 08:17 PM Re: IEC leads control
Anonymous
Unregistered


Don't think there is an issue as long as the fuse is rated correctly, the lead's inspected and safety-tested and identified as such.

Does anyone know of a reliable and effective method of managing the safety of IEC leads, including labelling as "inspected and tested"?

Any companies out there that produce a novel way of marking/labelling/identifying leads?

What inspection intervals should there be on IEC leads intended for use with medical equipment?

Top
#21327 - 06/10/03 08:31 PM Re: IEC leads control
Todd Mallow Offline
Visionary

Registered: 13/02/03
Posts: 35
Loc: UK
Grahame,

I have got to agree, the fuse is there to protect the cable. IEC leads are designed to be detachable so they can be used on different equipment. I have removed ALL 3 amp fuses from our cables as most are rated at 10A or more.
Why fit a 3A fuse to an IEC lead that is rated at 10A then plug it into a piece of equipment that draws 7A ?
3A fuses should not be fitted to IEC cables. I say 7A minimum.

Todd
_________________________
Weaseling out of things is important to learn. It's what separates us from the animals... except the weasel.

Top
#21328 - 07/10/03 10:25 AM Re: IEC leads control
Roy Offline
Philosopher

Registered: 11/07/00
Posts: 970
Loc: Stockport, Cheshire, England
We never use less than a 7 amp fuse. The heavier duty cables we fuse at 10 amps. Equipment which is rated lower than that is protected by its own fuses. In theory there should be no problem with swapping leads (except on high current equipment where it takes out the 7 amp fuse - but better that than melt the cable) but the detatchable leads are a real pain because they generate so many "false" failures - just because they're not plugged into the equipment properly.

There was a notice put out by the MDA a few years ago which suggested that all detatchable mains leads should be tested every six months because they were prone to damage. Ours get tested every 12 months along with all the toasters and kettles. The company who do the work stick a laminated label round the cable with the tested date on - but not a re-test date - just in case they're late getting round to it next year !

They find quite a few with the insulation pulled back or a connection broken - sometimes the earth - so it's definitely worth doing.
_________________________
Today is the day you worried about yesterday - and all is well !

Top
#21329 - 07/10/03 11:01 AM Re: IEC leads control
BSM Offline
Mentor

Registered: 20/07/01
Posts: 187
Loc: Here, of course!
We check the IEC lead of the equipment with every PPM and apply a label indicating the ID no. of the equipment to which the lead is attached (at the time of testing), the date, fuse rating and engineers initials.

If any leads are found to be faulty or damaged, they are replaced.

This doesn't stop the users moving the leads between different pieces of equipment once we've gone, but without fitting retainers to fix the lead to the equipment, what can you do?

Top
#21330 - 07/10/03 05:26 PM Re: IEC leads control
Marcel Eve Offline
Expert

Registered: 09/12/02
Posts: 143
Loc: Epsom & St Helier NHS Trust
Some years ago I got fed up with the spare mains leads hidden in drawers problem. Staff would regularly "find" leads in drawers that hadn't been tested for years and stick them onto equipment.
I decided to serial number & log every IEC lead, following which I planned to forbid staff to use unlabelled "found" leads.

I chose a really robust cable labelling system and started testing & labelling. After about 500 leads (1/4 of the stock)I dropped the project.

It took 2 minutes to test a lead, 6 minutes to label it up!

6 years later I still find my labelled cables around, the numbering beatifully clear but quite impractical.

I used to fuse all syringe pumps at 1A until I started finding the leads swapped onto defibs. I standard fuse leads at 7A these days.

Top
Page 1 of 5 1 2 3 4 5 >


Who's Online
30 Registered (abi, Ali-G., Apollo11, Baldrick, bill_mcg, biomedbill, Blinkin Flip, BobA, Dicky, Ed SWM, Eddie, Gav, Geoff Hannis, Gerry, Graham Roberts, Huw, John Stewart, Jonathan Wells, keano, Ken, KM, Mark Radbourne, Panander, ray hill, Roger, RoJo, seppy, SUZUKI, Tony Dowman, TonyR), and 108 guests online.
Newest Members
abi, jennyhicks, 9999, Wheelybobs, KevByard
4077 Registered Users
Jobs
Medical Equipment Technician
-- 1 December 2008

Medical Equipment Repair Technicians
-- 27 November 2008

Clinical Support Worker :: Technical Support
-- 27 November 2008

Senior Medical Engineer Band 5 / 6
-- 13 November 2008

Snr Asst. Practitioner - Clinical technology
-- 13 November 2008

Biomedical Electronics Team Manager
-- 10 November 2008

Senior Clinical Engineer
-- 3 November 2008

Newsletter

Subscribe
Unsubscribe