audiometer-room

Hearing tests

Before the development of electro acoustic equipment for generating and measuring sound, the available tests of hearing gave approximate answers at best. A person's hearing could be specified in terms of the ability to distinguish the ticking of a watch or the clicking of coins or the distance at which conversational speech or a whispered voice could be understood. The examiner also might note the length of time the person could hear the gradually diminishing note of a tuning fork, comparing the performance with his own.

Tuning-fork tests

A qualitative assessment of hearing loss can be carried out with a tuning fork. These tests exploit the ability of sound to be conducted through the bones of the skull. In the Rinne test the sounding tuning fork is placed on the mastoid process, and the person being tested is asked to report when it is no longer heard. The examiner then removes the fork immediately and holds the prongs close to the open ear canal. The normal ear continues to hear it for about 45 seconds, and this "positive" result occurs also with incomplete sensor neural impairment of hearing. When the result is "negative" and the fork is heard longer by bone conduction than by air conduction, a conductive type of deafness is present. In the Schwabach test the presence of a sensor neural impairment is indicated when the individual being tested cannot hear the bone-conducted sound as long as the examiner with normal hearing can. The individual with a conductive hearing loss, however, can hear the fork for a longer period of time than the examiner because the conductive lesion excludes the extraneous airborne masking noise of the surroundings. A bone-conduction audiometer would give a similar result.

For the Weber test, the fork is simply placed on the person's forehead, and the examiner asks in which ear the person hears it. If a sensor neural lesion is present in one ear, the person will localize the sound in the opposite, or "better," ear. If a conductive defect is present, the person will localize it in the "worse" ear--i.e., the one that is protected from interference by extraneous sounds. This simple test has been a valuable aid in the diagnosis of otosclerosis for many years.


History

With the introduction of the electric audiometer in the 1930s, it became possible to measure an individual's hearing threshold for a series of pure tones ranging from a lower frequency of 125 hertz to an upper frequency of 8,000 or 10,000 hertz. This span includes the three octaves between 500 and 4,000 hertz that are most important for speech. One of the standard frequency correction curves (or weighting) applied to sounds in a measurement device to simulate the hearing capability of the human hearing mechanism. The A weighting is the most common frequency weighting used in sound level measurements and is found in virtually every sound level meter. Other frequency weightings include the 'B', 'C' and 'Z' curves.


The Audiometer

The audiometer consists of an oscillator or signal generator, an amplifier, a device called an attenuator, which controls and specifies the intensity of tones produced, and an earphone or loudspeaker. The intensity range is usually 100 decibels in steps of 5 decibels. The "zero dB" level represents normal hearing for young adults under favourable, noise-free laboratory conditions. It was established in 1964 as an international standard.

frequency graph

Operating an audiometer

In pure-tone audiometry each ear is tested separately, while the other is shielded against sound. The person being tested wears an earphone or sits in front of a loudspeaker in a quiet test chamber, with instructions to give a hand signal whenever a brief tone is sounded. The audiologist proceeds to determine the lowest intensity for each frequency at which the person reports being just able to hear the tone 50 percent of the time. For example, one who hears the tone of 4,000 hertz only half the time at the 40-decibel setting has a 40-decibel hearing level for that frequency--i.e., a threshold 40 decibels above the normal threshold. A graph showing the hearing level for each ear by octaves and half octaves across the frequency range of 125 to 8,000 hertz is called an audiogram. The shape of the audiogram for an individual who is hard-of-hearing can provide the otologist or audiologist with important information for determining the nature and cause of the hearing defect. (The audiologist is primarily concerned with measuring the degree of hearing impairment; the otologist diagnoses and treats defects and diseases of the ear by medical or surgical means.)

A calibrated bone-conduction vibrator usually is furnished with the audiometer so that hearing by bone conduction also can be measured. When an individual has otosclerosis or another conductive defect of the middle ear, there may be a sizable difference between the air-conduction and bone-conduction audiograms, the so-called air-bone gap. This difference is a measure of the loss in transmission across the middle ear and indicates the maximum improvement that may be obtained through successful corrective surgery. When the defect is confined to the organ of Corti, the bone-conduction audiogram shows the same degree of loss as the air-conduction audiogram. In such cases of sensor neural impairment, surgery is seldom capable of improving hearing, but a hearing aid may be helpful.

Recruitment

Although faint sounds may not be heard at all by the ear with a sensor neural impairment, more intense sounds may be as loud as they are to a healthy ear. This rapid increase in loudness above the threshold level is called recruitment. When the opposite ear has normal hearing, recruitment can be measured by the alternate binaural loudness balance test. The subject is asked to set the controls so that the loudness of the tone heard in the defective ear matches that of the tone heard in the normal ear. By repeating the comparison at several intensity levels, the presence or absence of recruitment can be demonstrated.  When recruitment is excessive, the range of useful hearing between the threshold and the level at which loudness becomes uncomfortable or intolerable may be narrow, so that the amplification provided by a hearing aid is of limited value to the subject.

Discrimination score

Although hearing thresholds for pure tones give some indication of the person's ability to hear speech, direct measurement of this ability is also important.  Two types of tests are used most often. In one test the speech reception threshold is measured by presenting words of spondee pattern--i.e., words containing two syllables of equal emphasis, as in "railway" or "football"--at various intensity levels until the level is found at which the person can just hear and repeat half the words correctly. This level usually corresponds closely to the average of the person's thresholds for frequencies of 500, 1,000, and 2,000 hertz. A more important measure of socially useful hearing is the discrimination score. For this test a list of selected monosyllabic words is presented at a comfortable intensity level, and the subject is scored in terms of the percentage of words heard correctly. This test is helpful in evaluating certain forms of hearing impairment in which the sounds may be audible but words remain unintelligible. Such tests usually are carried out in a quiet, sound-treated room that excludes extraneous noise. These tests may give an overly optimistic impression of the ability of the individual with a sensor neural impairment to understand speech in ordinary noisy surroundings. For this reason speech tests are best carried out against a standardized noise background as well as in the quiet. A person with a conductive defect may be less disturbed by the noisy environment than a healthy subject. More elaborate tests, which often involve speech or sound localization, are available for evaluating hearing when central defects of the auditory system are suspected as a result of aging, disease, or injury. Their interpretation may be difficult, however, and the diagnostic information they furnish may be unclear.

Electrocochleogram (ECoG)

When the hearing of infants or others (who are unable to cooperate in standard audiometric tests) must be measured, their thresholds for pure tones can be established by electrophysiological means. One type of test is the electrocochleogram (ECoG). Electric potentials representing impulses in the cochlear nerve are recorded from the outer surface of the cochlea by means of a fine, insulated needle electrode inserted through the tympanic membrane to make contact with the promontory of the basal turn. This test provides a direct sampling of cochlear function.

Brain-stem-evoked response

A non-invasive, painless, and more frequently used test is brain-stem-evoked response audiometry (BERA). In this test electrodes are pasted to the skin (one placed behind the ear) and are used to record the neural responses to brief tones. The minute potentials evoked by a train of brief sound stimuli are suitably amplified and averaged by a small computer to cancel out background activity, such as potentials from muscles or the cerebral cortex. The typical recording shows a series of five or six waves that represent the responses of successive neural centres of the auditory pathway of the brain stem and provide information about the strength and timing of their activity.

Impedance audiometry

A simple and objective means of testing hearing at the level of the cochlea and brain stem is supplied by impedance audiometry. Two small tubes are sealed into the external canal. Through one tube sound from a small loudspeaker is injected into the canal. The portion that is reflected from the tympanic membrane is picked up by the other tube and led to a microphone, amplifier, and recorder. When a sudden, moderately intense sound is applied to the opposite ear, the stapedius muscle contracts, the impedance is increased, and the recorder shows a slight excursion as more sound is reflected. This test can provide information not only about the condition of the cochlea and the auditory pathways of the medulla but also about the facial nerve that innervates the stapedius muscle. However, it does not give an actual measurement of the acoustic impedance of the ear, representing the state of the ossicular chain and the mobility of the tympanic membrane.

This information can be obtained by means of the acoustic bridge--a device that enables the examiner to listen simultaneously to a sound reflected from the tympanic membrane of the subject and a sound of equal intensity reflected in an artificial cavity, with the volume being adjusted to equal that of the external canal of the ear being tested. When the two sounds are matched by varying the acoustic impedance of the cavity, the impedance of the ear is equal to that of the cavity, which can be read directly from the scale of the instrument.  Conductive defects of the middle ear, including disarticulation (separation) of the ossicular chain and immobility of the malleus or stapes, can be recognized by the characteristic changes they cause in the impedance of the ear as revealed by tympanometry. This test procedure consists in raising and lowering the air pressure in the middle ear to alter the stiffness in the tympanic membrane while measuring the changes in its compliance in terms of the amount of sound reflected from it.

Profound sensor neural deafness can occur as a result of viral and other infection, including mumps, measles, and meningitis. Rubella (German measles) in the mother during pregnancy can cause severe damage to the organ of Corti, resulting in profound hearing impairment in the child. Cochlear abnormalities may be present also as a result of genetic defects. The electrophysiological hearing tests described above--BERA and ECoG--make it possible to detect such loss in infants. In all such cases of deafness in young children it is essential that the condition be recognized as early as possible so that appropriate counselling in matters of care and education may be obtained. As an important aid in learning to speak, surgical implantation of an electronic cochlear prosthesis should be considered before the child reaches school age. These devices, while controversial within the deaf community, are proving effective in restoring a significant degree of hearing to many young children with congenital deafness

ABR - Auditory brainstem response is an electrical signal evoked from the brainstem by the presentation of a sound such as a click. Auditory brainstem response audiometry is a screening test to monitor for hearing loss or deafness, regularly used with newborn infants. It is a method used to assess the functions of the ears, cranial nerves, and various brain functions of the lower part of the auditory system, prior to the child developing to the point of describing a possible hearing problem.

 

ear

 

ABR audiometry is a safe and painless test of auditory pathway and brainstem function in response to auditory or [click] stimuli. The procedure was first described by Jewett and Williston in 1971. ABR audiometry is the most common application of auditory evoked responses. Test administration and interpretation are typically performed by audiologists. ABR technology has been used in testing newborns for many years. Approximately 1 in every 1000 children is born deaf. Many more are born with less severe degrees of hearing impairment, whilst others may acquire hearing loss during early childhood.

Historically, only infants who met one or more criteria on the high-risk register were tested. Universal hearing screening has been recommended because about 50% of the infants later identified with hearing loss are not tested when neonatal hearing screening is restricted to high-risk groups. Many hospitals are now implementing universal newborn hearing screening programs. These programs are possible because of the combination of technological advances in ABR and otoacoustic emissions [OAE] testing methods and equipment availability, which enables accurate and cost-effective evaluation of hearing in newborns. Moreover, the auditory brainstem response can also be measured in response to a complex speech signal. A growing body of literature has evaluated speech- evoked brainstem response differences between normal children and children with learning problems indicates that brainstem measures relating to the encoding of linguistic information can serve as a biological marker for auditory function in children with language-based learning problems, such as dyslexia. A consistent finding is that about one third of children with language-based learning problems exhibit a unique pattern of auditory neural activity that easily distinguishes them from the larger population of children with learning problems.

 

Sources:

http://images.google.co.uk/

http://www.audiometrictesting.co.uk

http://en.wikipedia.org/wiki/Auditory_brainstem_response

http://www.audiometry.co.uk/

www.boystownhospital.org

www.casellausa.com

Compiled and Edited by John Sandham IEng MIET MIHEEM Updated - April '07