INDICATION 202
The transfer time of the sample from the mixing cup towards the WBC measuring chamber is too long. This can be caused by :
a. Too low capacity of the waste pump.
b. A too large sample volume in the prediluted mode.
c. Shorting between the electrodes 3 and 4 in the mixing cup
d. Blockage in the center electrode (3). Rinse with a needle.
Press <CE> and perform a PRIME from the system flow menu to restore the system.