What can you tell me about thrombolysis pumps?
One of our nurses has asked the following:
"there are 2 types of thrombolysis, Streptokinase and TPA each with a different pump to administer?????. With TPA, I am right in thinking that with the TPA pump it takes longer to draw up and start, but once started can be left, or alternatively can be drawn up in 2 lots so treatment can be started quicker, but needs someone there for the 2nd lot to be administered? Please could you clarify/give your views".

Any thoughts on this matter??? confused


Be Proactive and reactive.
#20419 18/06/02 9:44 AM
Joined: Jul 2000
Posts: 1,916
Likes: 23
Hero
OP Offline
Hero
Joined: Jul 2000
Posts: 1,916
Likes: 23
TPA and streptokinase are each much more effective if begun quickly after
symptoms appear. For example, streptokinase infused/injected within an hour
of a heart attack was shown in one study to reduce deaths 47 percent -- a
reduction that declined to 26 percent if administered in the first three
hours and further declined to 17 percent if administered in three to six
hours.

The constraints here are:
Some staff have concerns about their ability to interpret ECGs with
sufficient confidence to initiate therapy and may need training to increase
their confidence to levels where they will be happy.

Experience with thrombolytic agents has raised concerns amongst some
clinicians about reactions - haemorrhagic, anaphylactic or arrhythmic.
However with the development of newer agents, it has become clear that most
of these concerns are unfounded. Training and education will need to address
this issue if wider use is to be achieved.
For agents that need complicated infusion or injection schedules, infusion
pumps are required. A number of scenarios may exist which is dependant on
the patients condition. A syringe driver may be required if fast acting
drugs are being used at low rates [i.e. less than 5ml/hr] the reason for this is that
syringe drivers are more accurate and deliver fast acting drugs smoothy due
to their linear operation. Some infusion pumps are very accurate over an hour, but over a few minutes at low rates the fluctuation in delivery (due to the pumping mechanism) can change by as much as 80%. At higher infusion rates with larger volumes it
may be neccessary to use infusion pumps. Dependant upon the skill of the
clinical user, the infusion pump may also be used in taper mode to deliver
at different rates automatically.

Thrombosis can also be treated with the use of DVT compression systems and
anti embolism stockings


Be Proactive and reactive.

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
4 members (stevo72, KeithO, kristy07, Ivan Lai), 75 guests, and 31 robots.
Key: Admin, Global Mod, Mod
Newest Members
Carys, Ivan Lai, Chris.S, NimBuS06, Januar
9,997 Registered Users
Forum Statistics
Forums25
Topics10,881
Posts72,935
Members9,997
Most Online5,980
Jan 29th, 2020
Powered by UBB.threads™ PHP Forum Software 7.7.5