OK, here we go ...
Biomed 101. Come on, Granny ... get sucking on those eggs!
First of all, let’s agree terms. By
biomed I am referring to the in-house technical support of medical equipment, carried out by resident technicians employed by (or closely allied with) the hospital. As we all know, there are other ways of supporting in-service equipment, but let’s stick with the in-house biomed model for sake of argument.
As I never tire of repeating, the
raison d'être of the biomed department is PM. From the dawns of biomed time, we have always done our regular "inspections" aimed at ensuring that the kit is kept in good shape. Myself I like the terminology (that originated from ECRI):-
I/PM.
Inspection(s) and Preventive Maintenance. After all these years, if biomeds are still not wholeheartedly on board with the notion of I/PM, and indeed actively engaged in carrying out an I/PM program, then (quite frankly) they may as well shut up shop, as they’re not doing any good.
(And I
am aware that some biomed shops see things somewhat differently. Basically they just hang-out in the workshop waiting for jobs to come to them. They are wrong, the management is wrong, and the service they provide is sub-standard. To paraphrase John Sandham, they are being reactive, but not proactive! This is not the way!)
Secondly, electrical safety testing is (and always has been) part of I/PM. If anyone disagrees, then they’re also just plain wrong! And it has also always been standard practice to carry out EST following a repair. It’s all just a matter of common sense after all.
If you don’t document the work you have carried out, then as far as "the record" is concerned it was never done! That’s why biomed also needs a decent database system to log, record and control the work.
So, there you have it. Three easy steps to true enlightenment:-
1) Embrace I/PM
2) Carry out EST as part of I/PM
3) Record your work in a database system
Folks, this stuff is not rocket science!
A quick word about "standards". It doesn’t really matter much which one you follow. In real terms, they are all much of a muchness, and even if you look closely you will be unlikely to distinguish between them (in terms of the
rationale behind them, allowable leakage currents
etc.). To my mind, you may as well use the latest one, which happens to be
IEC 62353. There is no need to go to the nth degree of
esoteria. Just stick to the basic tests, because that is all that’s required at first line. And ... I would suggest always following the latest equipment management "guidelines" too. As far as I’m aware, in the UK that would be
DB 2006(05). It’s there, and it’s an official document, so why not use it? (and section 4.4 even lays out, in reasonably clear terms, the legal requirement for electrical safety testing, and goes as far as to say "these include pre-use testing of new devices in addition to subsequent maintenance tests").
But finally, take a look through John's
articles ... they are generally well-reasoned, and up to date.