We are having problems with earth bonding tests on diagnostic microscope which have various aluminium swivel joints.

Unless I can overcome this problem, I will condemn these scopes.

Have any of you used this kind of grease on close fit joints to try and overcome the problem of oxidisation on aluminium?

Cheers

Mark

Joined: Aug 2001
Posts: 150
Mentor
Offline
Mentor
Joined: Aug 2001
Posts: 150
Mark,

I have never come across conductive grease being used for safety/grounding purposes. I am not sure I would trust it for maintaining an earth path.

How feasible would it be to run an earth braid from the base of the microscope (which I assume is grounded)to the rest of the Microscope?

Best wishes,

Paul

Joined: Jun 2001
Posts: 464
Mark.N Offline OP
Sage
OP Offline
Sage
Joined: Jun 2001
Posts: 464
Paul,

The grease idea was a 'stab in the dark'.

I presume that the aluminium has been anodised so that it's hard-wearing. Does anodising cause higher electrical resistance?

I've thought about adding additional cross-bonding, but this would look a mess(too much isolated exposed metal). Plus, would that not be a modification?

I think that I'll be happier if they're condemned.

Cheers

Mark



Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72
If you are insisting on condemning that microscope, Mark (and earth bonding is the answer), don't forget the well-known charity group based just twelve miles from where you are (Adrian knows the lady concerned) ... and I happen to be there right at this moment! smile


If you don't inspect ... don't expect.
Joined: Jun 2001
Posts: 464
Mark.N Offline OP
Sage
OP Offline
Sage
Joined: Jun 2001
Posts: 464
Geoff,

Let's just say that I'm a bit involved with this one for the time being.

This particular problem is not an isolated one, so I would urge anyone else who has one to check it's earth bonding from head to toe.

Cheers

Mark


Joined: Jul 2000
Posts: 1,968
Likes: 32
Hero
Offline
Hero
Joined: Jul 2000
Posts: 1,968
Likes: 32
Mark

I would tend to go for a bonding strap as Paul suggested.


Be Proactive and reactive.
Joined: Jun 2003
Posts: 338
Master
Offline
Master
Joined: Jun 2003
Posts: 338
If you put any form of grease on those swivel joints, I suspect you would then be able to lock / orientate them in the desired position.

Bonding is the only option as sttated above.


It is better to be reactive than radioactive...
Joined: Jun 2001
Posts: 464
Mark.N Offline OP
Sage
OP Offline
Sage
Joined: Jun 2001
Posts: 464
If the aluminium components have been anodised, would they still be classed as 'exposed metal'?

Cheers

Mark

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72
Semantics, Mate. If the protective earth reading is too high, simply add earthing "straps" (green and yellow "pigtails") until the value is brought back within limits. What's the problem? Kit like this has been around since the Dawn of Biomed (without, as far as I recall, anyone having sparks jumping out of their ears)! smile


If you don't inspect ... don't expect.
Joined: Sep 2006
Posts: 745
Philosopher
Offline
Philosopher
Joined: Sep 2006
Posts: 745
In the good old days I can remember seeing earthing braid on slit lamps for the same problem.

I've also tested units where the only exposed metal is anodised and you can only get a contact by "stabbing" through that coating.

Lee


Don't forget "we've never had it so good".
Page 1 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (daisizhou), 12,013 guests, and 148 robots.
Key: Admin, Global Mod, Mod
Newest Members
AraEmin, Edouardohaag, Mark B, Royal, davidarhui
10,392 Registered Users
Forum Statistics
Forums26
Topics11,281
Posts74,584
Members10,392
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5