Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 1 of 2 1 2
Joined: Feb 2006
Posts: 4
ELIS Offline OP
Newbie
OP Offline
Newbie
Joined: Feb 2006
Posts: 4
Hi Medical Engineers
I have problem. We used in operating rooms the device with pulsed consumption. There was tested the leakage current and it was OK. But somebody connceted the device to the unearthed system and the insulation monitoring device showed the insulation fault so the resistence was not stabill the resistance was moved it seems with relationship of pulsed consumption of the device.
Has the device any defect? Can we claim the device on the manufacture so the device is damaged? The defect will not appeared if the device will not be connect into unearthed system.
Thanks for your reply and advices .
ELIS

Joined: Aug 2004
Posts: 88
Adept
Offline
Adept
Joined: Aug 2004
Posts: 88
What machine/device is it? How does it monitor the earth system?
Your questions are difficult to answer unless you provide more details.

Joined: Feb 2004
Posts: 14,767
Likes: 70
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,767
Likes: 70
OK ELIS ... so you're using a certain piece of equipment in an operating theatre which has an isolated power system, right? And, when that equipment is running in a certain mode (pulsing), it triggers an alarm (over-leakage-current) at the LIM (line isolation monitor)?

OK, it's leaking a bit of current at that stage (probably due to RF leakage if it's an electro-surgical unit, or a motor being in circuit, perhaps, if it's a pump of some kind). But it may not be faulty. What other equipment is plugged into the IPS at the same time? Don't forget that the LIM shows the sum of the leakage currents of all equipment plugged in. Try disconnecting all other items and see what happens. Also, is the LIM correctly set?

Naturally, the item will be assumed to be working normally (and I suspect that it is) when plugged into a "normal" (earthed) outlet ... because then there is no LIM to report the leakage. It will still be there, though. Try testing it with your Electrical Safety Tester. If it passes the EST, then basically it's OK.

But Bong is right. We would be interested to hear what this equipment is. My guess is an electro-surgical unit. Am I right? smile

Joined: Feb 2008
Posts: 129
Expert
Offline
Expert
Joined: Feb 2008
Posts: 129
..or perhaps it is something with a heating element using some type of proportional control?

ELIS, please tell us what the equipment is, we would all like to know! shades

Joined: Feb 2006
Posts: 4
ELIS Offline OP
Newbie
OP Offline
Newbie
Joined: Feb 2006
Posts: 4
The equipment is the laser for lithotripsie. The insulation monitoring device shows the lower insulation resistance when only the laser was connect to the unearthed system.
ELIS

Last edited by ELIS; 04/11/08 9:52 AM.
Joined: Feb 2004
Posts: 14,767
Likes: 70
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,767
Likes: 70

OK, ELIS. More questions:-

Is this a new piece of equipment, recently commissioned ... or something old, and rarely used?

Is it excessively humid in the theatre?

Have you checked out the kit with your Electrical Safety Tester?

If there is a fault, take a close look at leads and cables (especially those that may have been "run over" by wheeled equipment) ... then take a look at the power supply unit (big fat capacitors?). smile

Joined: Feb 2006
Posts: 4
ELIS Offline OP
Newbie
OP Offline
Newbie
Joined: Feb 2006
Posts: 4
Thank you Goeff you wants to help me
The laser is new approx. two month. I think in the operating room is not any excessively humid.
The laser was check by medical electrical safety tester and the leakage currents are in standard value.
The insulation monitoring device monitor the resistance between the active conductors and earth. Regarding to our low, the critical level is 47 kohms. The problem is so when to the insulation system is connect only the laser, than the gauge and his needle is not on the value "infinite", but the needle is moved between value "infinite" and critical value ( the laser is connect only and switched on without activation).
ELIS

Last edited by ELIS; 04/11/08 8:48 PM.
Joined: Feb 2004
Posts: 14,767
Likes: 70
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,767
Likes: 70

H'mmm. On the face of it, it does indeed sound like excessive leakage current there somewhere. It could indicate a defect of some kind. Or perhaps the thing just draws a heavy current anyway, upsetting (or confusing) the LIM. Odd, though, that you get comfortable readings with your electrical safety tester. You may find that you need to provide a special (that is, standard, "earthed") power outlet in the theatre just for this unit (as sometimes needs to be done for other items, like mobile C-arm x-ray units, for example). But as the unit is still "new", perhaps you should first of all seek the opinion of the supplier! smile

Joined: Feb 2008
Posts: 129
Expert
Offline
Expert
Joined: Feb 2008
Posts: 129
I agree with Geoff that you should talk to the supplier of the laser.

It is strange that the LIM and the safety tester seem to disagree...or do they?! What is your supply voltage? What leakage current readings did you see?

Joined: Feb 2004
Posts: 14,767
Likes: 70
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,767
Likes: 70

Yes, that's an interesting topic in itself:- LIM versus electrical safety tester (EST)! Put simply, doesn't the LIM essentially look at the imbalance of currents drawn at L1 and L2 (as we can't really say line and neutral when talking about an isolated power system, can we)? It then regards the difference as leakage. But, maybe it's not current leakage within the equipment itself ... perhaps there's a parallel circuit somewhere!

And, of course, the EST measures the true leakage current(s) value(s) by interrupting the protective earth conductor (and all the rest).

It would be "educational" to get to the bottom of this one. My next question to ELIS would have to be:- how much supply current does this equipment draw (consume)? Is it something hefty? smile

Page 1 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
2 members (mini nair, mub777), 643 guests, and 17 robots.
Key: Admin, Global Mod, Mod
Newest Members
Richard 1x1, Stephen UHM, AndyC85, Jose Luis, jaringa
10,258 Registered Users
Forum Statistics
Forums25
Topics11,164
Posts74,190
Members10,258
Most Online5,980
Jan 29th, 2020
Powered by UBB.threads™ PHP Forum Software 7.7.5