Dear all,

As a safety measure we have been issued with bench top Isolating transformers. I have measured the voltage between the earth and live at the output of the transformer and found that it is 100 volts. My understanding of these is that the live is no longer referenced to earth and there should be 0 volts P.D.

Does anyone have any experience with these?

Best wishes,

Paul.

Joined: Sep 2006
Posts: 25
Dreamer
Offline
Dreamer
Joined: Sep 2006
Posts: 25
Originally Posted By: Paul Latham
Dear all,

<snip> I have measured the voltage between the earth and live at the output of the transformer and found that it is 100 volts. My understanding of these is that the live is no longer referenced to earth and there should be 0 volts P.D.
<snip>



Hi Paul,

So your expectation would be that there is 0 volts PD between Earth and Live? Sounds scary!

As you say, the point of an isolating transformer is that it is "isolated" from the main earth. Therefore i would think you cannot reference the output to some random "earth" point, as it has no return path.

Regards,
Robert

Joined: Aug 2001
Posts: 150
Mentor
OP Offline
Mentor
Joined: Aug 2001
Posts: 150
Hello Robert,

The rationale of these transformers seems to be that most electrical accidents take place because someone is more likely to be in contact with earth and live than neutral and live. I have measured a voltage of 100v which I accept is a step in the right direction but still think that it is a high enough voltage to cause an accident.

Best wishes

Paul.

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

The idea of the bench isolating transformer is that the hapless tech can safely (stupidly?) have one paw on an energised "live" part of a piece of equipment and the other paw (or some other extremity) at "earth" ... and still live to tell everyone how dumb he is! frown

I believe that, technically speaking (and we are, after all, technical folk), it is incorrect to talk about "Live" and "Neutral" in the context of isolating transformers. I should imagine that you'll also see circa. 100 VAC on the other terminal, too, Paul (rather than the near as dammit zero that we hope you find at the Neutral of a normal, earth referenced mains outlet).

As you may imagine, I am not wholly in favour of isolating transformers being used to protect idiots working at benches, mainly because they introduce a non-standard electrical environment (one that the equipment itself will not encounter outside of the operating theatres). They can also lull the tech into an entirely false sense of security. Far better to engage brain before attacking the job at hand. If some kind of safety device is deemed necessary, I suggest the Residual Current Circuit Breaker (sometimes known as a residual-current device).

I have a plug-in RCD outlet adaptor (that I bought cheaply from Wickes) that I carry around for use when working in dodgy (or electrically uncertain) surroundings. It trips out on both poles in less than 30 ms when it detects a imbalance between input and return (due to a fault, or heavy leakage) of 30 mA. Perhaps one day I'll find one that I can adjust to a lower tripping current, but in the meantime the one I have works for me, and I'm not dead yet.

Lastly, isolation transformers sometimes also cause problems for the unwary. Does your electrical safety tester get passed the mains voltage check when plugged in to the iso Tx, for instance? smile


If you don't inspect ... don't expect.

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 11,593 guests, and 194 robots.
Key: Admin, Global Mod, Mod
Newest Members
AraEmin, Edouardohaag, Mark B, Royal, davidarhui
10,392 Registered Users
Forum Statistics
Forums26
Topics11,282
Posts74,585
Members10,392
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5