Simple answer "Yes".
There is a perenial argument over KVO that is why it is an option to turn it on or not on modern syringe pumps and to let the hospital decide the protocol. Here, at GOSH, it is turned on except on pumps with a large red sticker, the nursing staff then decide whether they want KVO or not. Ther is no correct answer only one that is right for that situation.
I am worried that you are using IV pumps at such a low rate. it might be possible to set them but manufacturers often give a minimum recommended usable rate. The error of delivery is often given as a percentage of set rate for high rates and as an absolute for low rates. if you are using it at the minimum rate the error might be greater than the set rate. Also a slight movement of the tubing will deliver a bolus that could easily be equal to an hours dose. See the user manual for details for your pumps but I am sure IV pumps are not supposed to be used at this low a rate.
For such low rates of fluid delivery I would definitely use a syringe pump and then make sure a small syringe was used to get a decent linear speed. The rubber plunger can flex and give a pulsatile flow. For a 50ml syringe 1ml/hr is approximatele 1mm/hr linear speed, so a flex of 0.1mm (which is hardly anything) can cause a significant bolus - and below that speed...well.
As a last resort refer to the new guidance from the MDA DB2003(02) which tells you everything you need to know about infusion systems. (Free to the NHS)
Hope this helps or at least gives food for thought.