Joined: Sep 2006
Posts: 10
Novice
|
OP
Novice
Joined: Sep 2006
Posts: 10 |
We have had the McKinley BodyGuard pumps for about a number of years now and recently bought the T34 (24Hr ambulatory pump) ... we have only just discovered that the pumps contain an internal lithium battery which controls the time and date and needs routinely checking and replacing!
BE WARNED - THIS IS NOT IN THEIR MAINTENENCE MANUAL!!
Is anyone else finding that the technical information that they provide is totally and utterly rubbish?
|
|
|
|
|
Joined: Nov 2008
Posts: 40
Technologist
|
Technologist
Joined: Nov 2008
Posts: 40 |
Thanks for the warning. Given some of the competetion, I still hold the McKinley pumps in very high regard! I checked the service manual, and you are quite correct. However I would not expect date and time to be critical to the pumps operation, and hence I doubt we shall start checking them on a regular basis.
|
|
|
|
|
Joined: Aug 2009
Posts: 66
Scholar
|
Scholar
Joined: Aug 2009
Posts: 66 |
Hi What version of the manual do you have - I have Jan 2007 and fine the lithium battery mention on page 17.
I can send you the document if you need.
Phil
|
|
|
|
|
Joined: Feb 2004
Posts: 14,815 Likes: 72
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,815 Likes: 72 |
I believe that the OEM of these pumps is CME in Israel. So there's another source for information, parts, training ... and, indeed, the pumps themselves right there! 
If you don't inspect ... don't expect.
|
|
|
|
|
Joined: Sep 2006
Posts: 10
Novice
|
OP
Novice
Joined: Sep 2006
Posts: 10 |
If the device is involved in a clinical incident then the date and time is critical. its not just an offset with the pumps its drifting
|
|
|
|
|
Joined: Feb 2004
Posts: 14,815 Likes: 72
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,815 Likes: 72 |
Two thoughts:- 1) Has anyone taken this up with CME McKinley in the UK? 2) Does anyone use the BodyComm device to download data from the T34 (it looks like a useful piece of kit)?
If you don't inspect ... don't expect.
|
|
|
|
|
Joined: Feb 2004
Posts: 14,815 Likes: 72
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,815 Likes: 72 |
Can anyone supply a .pdf service manual for the T34? 
If you don't inspect ... don't expect.
|
|
|
|
|
Joined: Dec 2004
Posts: 578 Likes: 1
Philosopher
|
Philosopher
Joined: Dec 2004
Posts: 578 Likes: 1 |
The service manual can be obtained from McKinley on their training course.
As for the BodyComm device Geoff, yes lovely little device, allows you to quickly make modifications to the config via the pc without having to mess with entering the codes via the syringe driver.
The only thing I've found on these is that compared to the Graseby MS16A the parts are more expensive, but then again what goes wrong on a MS16A is cheap and cheerful.
|
|
|
|
|
Joined: Feb 2004
Posts: 14,815 Likes: 72
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,815 Likes: 72 |
The service manual can be obtained from McKinley on their training course. Apparently so ... but £ 1,800 is a bit beyond my budget, I'm afraid. 
If you don't inspect ... don't expect.
|
|
|
|
|
Joined: Sep 2010
Posts: 20
Dreamer
|
Dreamer
Joined: Sep 2010
Posts: 20 |
McKinley are a waste of space. Their technical department has got well down hill in the last 3 or so years!
I think its the case of a small company getting bigger and not giving a hoot.
|
|
|
|