|
|
|
|
Joined: Apr 2005
Posts: 380
Sage
|
OP
Sage
Joined: Apr 2005
Posts: 380 |
Hi guys. Can any one guide me or share the COAG calibration of SSE2L ESU. I am having problem of coag working on 7, 8,9 & 10 setting on front knob.. On low settings it is not working.. Any help will be appreciated. thanks in advance & regards.
|
|
|
|
|
Joined: Sep 2001
Posts: 499 Likes: 1
Sage
|
Sage
Joined: Sep 2001
Posts: 499 Likes: 1 |
I would suspect the potentiomenter in the first instance, I persume it works correctly on high settings only? not at all or some times? please clarify?
Which ESU tester are you using?
What load resistance are you using?
UMi-007
"WORK SMART NOT HARD !"
|
|
|
|
|
Joined: Dec 2009
Posts: 176 Likes: 1
Mentor
|
Mentor
Joined: Dec 2009
Posts: 176 Likes: 1 |
Hi, Laiq. I have SSE2L service manual. I could send you if you give me your e-mail. Ok? My e-mail is hugo_68@yahoo.com By
Last edited by Pavel; 17/12/09 3:26 AM.
|
|
|
|
|
Joined: Apr 2005
Posts: 380
Sage
|
OP
Sage
Joined: Apr 2005
Posts: 380 |
Dear Umish & Pavel Thanks for your reply. First i am using ELECTROSURGICAL ANALYSER MODEL443 from NUERODYNE DEMPSEY INC.. Settings are 500 Ohms load wattmeter.. The cut is giving max of 150 watts & coag is 80 watts max at 10 position. Secondly can any one share what compatible Power Transistors are u guys using inplace of their original ones.. I know we have to replace all transistors if we have to use compatible ones. Cant use mix of original & compatibles.. Hope to hear from u guys soon. Thanks again for all of your help. Regards
|
|
|
|
|
Joined: Dec 2009
Posts: 176 Likes: 1
Mentor
|
Mentor
Joined: Dec 2009
Posts: 176 Likes: 1 |
Laiq, I have same ESU tester but for ESU Valleylab we place 750 ohms/50W load in paralel with load ESU tester (500 ohms in NORMAL position). Load result in 300 ohms. By
|
|
|
|
|
Joined: Mar 2010
Posts: 6
Newbie
|
Newbie
Joined: Mar 2010
Posts: 6 |
Hi Pavel,
Could you send me a copy of the SSE2L service manual..
My email is vnoguedad@ap8vm.com
thanks
|
|
|
|
|
Joined: Oct 2010
Posts: 3
Newbie
|
Newbie
Joined: Oct 2010
Posts: 3 |
Hi Pavel,
Could you send me a copy of the SSE2L service manual..
My email is mesolution@gmail.com
thanks
|
|
|
|
|
Joined: Dec 2010
Posts: 12
Novice
|
Novice
Joined: Dec 2010
Posts: 12 |
Hi, Laiq if nothing fuse blow on power amplifier modul, you must change resistor 14(18k-27k ohm) & 15(1,6-2,2k ohm) on multivibrator board(A3)
|
|
|
|
|
Joined: Apr 2005
Posts: 380
Sage
|
OP
Sage
Joined: Apr 2005
Posts: 380 |
Thanks andy.. I have a stubborn SSE2L, giving me hard time. I will try it.
|
|
|
|
|
Joined: Jul 2010
Posts: 18
Novice
|
Novice
Joined: Jul 2010
Posts: 18 |
hi every one my name is fadel .I hv the service manual for SSE2L I can give it to any one but i need the service manual for MARTIN ME MB1 pls if any one hv pls send to me my email if fadelbiomedical@yahoo.com
|
|
|
|
2 members (Huw, daisizhou),
11,857
guests, and
146
robots. |
|
Key:
Admin,
Global Mod,
Mod
|
|
|
Forums26
Topics11,281
Posts74,584
Members10,392
| |
Most Online69,877 Jul 17th, 2026
|
|
|
|
|
|