Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 1 of 3 1 2 3
#64186 25/03/13 4:52 PM
Joined: Aug 2001
Posts: 17
Novice
OP Offline
Novice
Joined: Aug 2001
Posts: 17
Our 98 Defibs have been due battery change since last Aug/Sept. Is anyone else awaiting delivery of the XL battery pack?

I did the battery test last year on all devices and found an 8% reject rate. I'm now concerned this may have increased in the last six months.

Batteries were first promised in November.
A recent 'Philips 'letter dated 6th February 2013 is still unclear on delivery.

Joined: Feb 2004
Posts: 14,456
Likes: 34
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,456
Likes: 34

What's all that about, Alan? Why are you waiting? Care to explain a little further?

More "battery blues" ... ? think


If you don't inspect ... don't expect.
Joined: Sep 2006
Posts: 745
Philosopher
Offline
Philosopher
Joined: Sep 2006
Posts: 745
Hello

There was a recall on a batch of Heartstart XL batteries last year (MDA/2012/077)
Quote:
Medical Device Alert
Ref: MDA/2012/077 Issued: 08 November 2012 at 14:00
Device
Batteries for HeartStart XL defibrillator/monitor.
Battery part number M3516A.
Batch labelled ‘Made in Taiwan’ with ‘Date of Manufacture’ code ‘R-2011-12’.
Manufactured by Philips and distributed by Philips and Cardiac Services UK & Ireland.


with Philips replacing the affected batteries; is this a knock on effect?



Lee


Don't forget "we've never had it so good".
Joined: Aug 2007
Posts: 306
Master
Offline
Master
Joined: Aug 2007
Posts: 306
We are the same here, awaiting delivery of batteries since las September.

All should have been changed in November. 30ish

Must chase it again, thanks for reminding me.

Joined: Sep 2008
Posts: 31
Visionary
Offline
Visionary
Joined: Sep 2008
Posts: 31
Contacted Philips in March chasing an order for defib batteries and was told they hope to have some Q1 2013 but that was all they could offer.

Joined: Jul 2002
Posts: 2,020
Hero
Offline
Hero
Joined: Jul 2002
Posts: 2,020
Q1 2013 That would be sometime between 1st january and the 31st March. Unless they meant the financial quarter startin on 1st April.
Now if they used a general generic battery there would be several manufacturers they could buy it from.... and so could we...Oh I get it now.
Robert


My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
Joined: Aug 2008
Posts: 84
Likes: 6
Adept
Offline
Adept
Joined: Aug 2008
Posts: 84
Likes: 6
I spoke to them today they now say mid June is the earliest expected date. Just out of interest has anyone take to fitting standard 12v 2.3 ah batteries? I have just had to do this today one of our batteries was failing after 4 discharges, it was 2 years and 3 months old. We have been waiting since September for our order.

Joined: Feb 2004
Posts: 14,456
Likes: 34
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,456
Likes: 34

Again I ask:- what is the (real) problem here?

Technology being introduced before it has been sufficiently developed? think

Whatever it is ... there has to be an engineering solution. So what is it?


If you don't inspect ... don't expect.
Joined: Aug 2008
Posts: 84
Likes: 6
Adept
Offline
Adept
Joined: Aug 2008
Posts: 84
Likes: 6
The Philips manual suggests you can use non original batteries so maybe we all should. The original batteries were Panasonic, they withdrew from making this type of battery some time ago so it would appear to be a production problem with the new supplier?

Joined: Jul 2002
Posts: 2,020
Hero
Offline
Hero
Joined: Jul 2002
Posts: 2,020
What is wrong with buying the batteries from XXX* rather then Philips? They are the same battery at the end of the day.
Robert
* Insert supplier name of choice


My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
Page 1 of 3 1 2 3

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 52 guests, and 18 robots.
Key: Admin, Global Mod, Mod
Newest Members
Daniel Rothwell, tomandjogh, Pascachristi, bme2020, drleiva
9,974 Registered Users
Forum Statistics
Forums25
Topics10,858
Posts72,863
Members9,974
Most Online5,980
Jan 29th, 2020
Powered by UBB.threads™ PHP Forum Software 7.7.5