hi dear friends
i have a problem in Bennett 760 ventilator that it restart during in normal working. the ac mail supply and internal battery situation is ok and EST procedure is passed.
does anybody help me to solve this malfunction?
thanks

Recruitment / Jobs
Advertise your jobs with EBME
Joined: Jul 2002
Posts: 2,020
Hero
Offline
Hero
Joined: Jul 2002
Posts: 2,020
I have managed to totally reset one of these vents by using a walkie-talkie near it (in a controlled test situation I might add). They seem particularly prone to RF interference. Was anyone using a radio transmitter or telephone near it at the time?
Robert
PS It went in to fault mode and had to have a total reset before it would operate again.


My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
Joined: Dec 2009
Posts: 72
Scholar
OP Offline
Scholar
Joined: Dec 2009
Posts: 72
dear Rojo
it's a good point you have mentioned. but there is no RF source near it.

Joined: Feb 2004
Posts: 14,658
Likes: 61
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,658
Likes: 61

How many times (and how often, and at what time of day) has this happened, I wonder? think

Although you have suggested that the mains power supply is OK, that's where I would suspect the problem might lay. It would be nice to put a UPS in circuit and see (listen) how many times it bleeps!


If you don't inspect ... don't expect.
Joined: Dec 2009
Posts: 72
Scholar
OP Offline
Scholar
Joined: Dec 2009
Posts: 72
َActually it's connected to ups and i checked there was no interrupt in power line.
abnormal restart almost happens every 3 min approximately.

Joined: Feb 2004
Posts: 14,658
Likes: 61
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,658
Likes: 61

How about the output from the UPS? Are you sure it's OK?

Have you actually been alongside the 760 when it re-boots? Any clues from indicator lights etc.?

No doubt you've already checked inside for "bad connections" ... and re-seated boards, and so forth.

Is there any NV-RAM in there? Battery-backed? What's the state of the battery? think

Joined: Dec 2009
Posts: 72
Scholar
OP Offline
Scholar
Joined: Dec 2009
Posts: 72
the problem solved!
the gear of motor that drive the piston shaft push it very hard, so sometimes the piston does not work probably.

Joined: Feb 2004
Posts: 14,658
Likes: 61
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,658
Likes: 61

Interesting. smile

So what was happening ... motor over-load causing a breaker to trip? Or an error signal causing a reset, or partial close-down? Or what?

Joined: Jul 2002
Posts: 2,020
Hero
Offline
Hero
Joined: Jul 2002
Posts: 2,020
Quote:
the piston shaft push it very hard

Lack of maintenance?...oil for example?
Robert


My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
Joined: Dec 2009
Posts: 72
Scholar
OP Offline
Scholar
Joined: Dec 2009
Posts: 72
no, i think it happens because of vibration and shaking in bad transportation.

Page 1 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 314 guests, and 15 robots.
Key: Admin, Global Mod, Mod
Newest Members
rajvenugopal, Arzo Momand, steve_shomz, trat, Mitul Patel
10,178 Registered Users
Forum Statistics
Forums25
Topics11,061
Posts73,725
Members10,178
Most Online5,980
Jan 29th, 2020
Powered by UBB.threads™ PHP Forum Software 7.7.5