Home Articles Downloads Forum Products Services EBME Expo Contact ☰
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 2 of 3 1 2 3
Joined: Apr 2013
Posts: 44
jbt Offline OP
Technologist
OP Offline
Technologist
Joined: Apr 2013
Posts: 44
Thanks for the advice, let´s see how it goes. I have already changed the Fuseholder for a new one, and is working for the other cycles, only if a choose the drying cycle is the problem

Joined: Apr 2013
Posts: 44
jbt Offline OP
Technologist
OP Offline
Technologist
Joined: Apr 2013
Posts: 44
Going crazy with this unit. I start the day and choose the drying option, the unit blocks itself and does nothing, so just turn it off. COme back again next day in the morning, turn the unit ON and it starts the drying cycle on itself, its like it is picking up what did not do the day before?

Maybe a defective EPROM?

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Have you had a good dig around the solid-state relay that controls the drying fan? Looking at the diagram, that is (should be) the only way the drying fan gets energised. smile

Joined: Apr 2013
Posts: 44
jbt Offline OP
Technologist
OP Offline
Technologist
Joined: Apr 2013
Posts: 44
Yes, the SSR are working perfectly, and as I say, it only doesnt work when you select a Dry Only program, but the drying pump works when the drying is part of a Sterilization cycle.

Joined: Sep 2010
Posts: 413
Likes: 2
Sage
Offline
Sage
Joined: Sep 2010
Posts: 413
Likes: 2
..any development?


--- PATIENT SAFETY FIRST ---
Joined: Apr 2013
Posts: 44
jbt Offline OP
Technologist
OP Offline
Technologist
Joined: Apr 2013
Posts: 44
Waiting on a new EPROM, hope this will solve the issue.

Joined: Sep 2010
Posts: 413
Likes: 2
Sage
Offline
Sage
Joined: Sep 2010
Posts: 413
Likes: 2
...let us Know.


--- PATIENT SAFETY FIRST ---
Joined: Apr 2013
Posts: 44
jbt Offline OP
Technologist
OP Offline
Technologist
Joined: Apr 2013
Posts: 44
Just replaced the Eprom, now the crying cycle works very well. I recalibrated the income water time which was 140s to 40s, but now 2 new problems arised, one is with the water, the unit starts the water pump for the set time, stops 1s then turn on again the water pump for 20-25s more. The other problem is that the unit reaches 134° but pressure seems to be a little down for that temp. and once the temp is reached, it doesnt go into sterilization time????
When the cycles are for 121° is working just fine. Going crazy with this unit.

Joined: Sep 2010
Posts: 413
Likes: 2
Sage
Offline
Sage
Joined: Sep 2010
Posts: 413
Likes: 2
How come Eprom is the problem? maybe there's something else or there's a reason why? smile


--- PATIENT SAFETY FIRST ---
Joined: Apr 2013
Posts: 44
jbt Offline OP
Technologist
OP Offline
Technologist
Joined: Apr 2013
Posts: 44
For the lack of pressure I found a small crack in the door gasket, we will see how it goes after we replace it.

Page 2 of 3 1 2 3

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (daisizhou), 11,665 guests, and 202 robots.
Key: Admin, Global Mod, Mod
Newest Members
AraEmin, Edouardohaag, Mark B, Royal, davidarhui
10,392 Registered Users
Forum Statistics
Forums26
Topics11,282
Posts74,585
Members10,392
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5