810:15 Air in line sensor saturated for 1.0 mL (infusion)
or 5 seconds (stopped)
1. Check for moisture and clean the tubing channel.
2. Perform AIL test on page 11-14 to confirm AIL is
functioning properly.
3. If failure recurs, replace pump module per “Pump
Module,” 8-180.
HTH
Rob.