Re: Cleaning Medical Equipment
Rob1234 #71957 29/06/17 9:01 PM
Joined: Feb 2004
Posts: 14,225
Likes: 19
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,225
Likes: 19

Back to Alcowipes, then (individual sachets). smile

70% v/v Isopropyl Alcohol BP


If you don't inspect ... don't expect.
Re: Cleaning Medical Equipment
Lee S #71958 30/06/17 5:37 AM
Joined: May 2004
Posts: 77
Adept
Offline
Adept
Joined: May 2004
Posts: 77
We are definately seeing widespread degredation of plastic components with cracking of handset cases and cables on Digiterm chairs cleaned with Clinell wipes which is currently being tested at Digiterms factory.
We think one of the components in the fluid is removing the plastisiser.
Also had reports from customers of plastic components on the front of dialysis machines cracking and a screen damaged.

Re: Cleaning Medical Equipment
Chris Pearson #71962 30/06/17 10:47 AM
Joined: Feb 2004
Posts: 14,225
Likes: 19
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,225
Likes: 19

There is always going to be a downside somewhere along the line when "industrial chemicals" are used as disinfectants or cleansing agents - especially on "plastics" of various compositions.

No doubt all manufacturers these days make recommendations about how (what to use) to clean their equipment. Our old friend "hot soapy water" (and tepid, too) seems to be favourite.

Meanwhile, I have had good results in the past when experimenting with "natural" substances, usually locally (and cheaply) available; freshly squeezed lemon - and even clear spring water - come to mind! smile


If you don't inspect ... don't expect.
Re: Cleaning Medical Equipment
Chris Pearson #71966 30/06/17 3:29 PM
Joined: Feb 2017
Posts: 15
Novice
Offline
Novice
Joined: Feb 2017
Posts: 15
As new kid on the block have seen similar problems to Chris, but its not the Cinell wipes that is the problem in Dialysis clinic's I have seen the using a product called "viacon" which is very agressive and causes problems with plastic

Re: Cleaning Medical Equipment
Stargolf #71967 30/06/17 6:27 PM
Joined: Feb 2004
Posts: 14,225
Likes: 19
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,225
Likes: 19

Did you mean Virkon? think


If you don't inspect ... don't expect.
Re: Cleaning Medical Equipment
Geoff Hannis #71969 01/07/17 5:54 PM
Joined: Feb 2004
Posts: 14,225
Likes: 19
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,225
Likes: 19

Meanwhile, I forked out a quid for a pack of forty "Anti+Bac Original Multi-Surface Wipes" at Tesco's today. smile

Non-aggressive, yet more than adequate for most run-of-the-mill surface disinfecting, I would have thought.

Anything contaminated with the more esoteric bugs should be referred to Infection Control and (or) CSSD anyway, I reckon.

1) Wetwipe info.
2) CosIng database (example)


If you don't inspect ... don't expect.
Re: Cleaning Medical Equipment
Lee S #71971 02/07/17 8:25 PM
Joined: Feb 2017
Posts: 15
Novice
Offline
Novice
Joined: Feb 2017
Posts: 15
Geoff,

Thanks yes i did mean Vircon had mental block

Re: Cleaning Medical Equipment
Chris Pearson #71973 03/07/17 11:37 AM
Joined: Sep 2006
Posts: 745
Lee S Offline OP
Philosopher
OP Offline
Philosopher
Joined: Sep 2006
Posts: 745
Hello

Chris do you know what type of plastic is being affected?

Lee


Don't forget "we've never had it so good".
Re: Cleaning Medical Equipment
Lee S #71974 03/07/17 11:56 AM
Joined: Dec 2015
Posts: 86
Likes: 11
Adept
Offline
Adept
Joined: Dec 2015
Posts: 86
Likes: 11
Welch Allyn provide labels that state "clinell safe" for the users to place on their products. Clinell needs to be applied and then removed several minutes later with a damp cloth, this is recommended by Clinell. Be very cautious when using Clinell on any older plastic devices as they will degrade the strength over time.

Last edited by Dustcap; 03/07/17 11:56 AM.

30 years since the Chernobyl disaster and yet we still have no super heroes or zombies.
Re: Cleaning Medical Equipment
Lee S #71976 03/07/17 2:09 PM
Joined: Mar 2004
Posts: 31
Visionary
Offline
Visionary
Joined: Mar 2004
Posts: 31
Not sure if it's of interest, but some of my colleagues looked into disinfection including different wipe types/ingredients, link here

Page 2 of 3 1 2 3

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
2 members (Neoteny, DaveC in Oz), 135 guests, and 22 robots.
Key: Admin, Global Mod, Mod
Newest Members
Maria LN, Maher, rudied, Dave Aldrete, alvinobogne
9763 Registered Users
Forum Statistics
Forums25
Topics10,672
Posts72,077
Members9,763
Most Online5,980
Jan 29th, 2020
Powered by UBB.threads™ PHP Forum Software 7.7.4