Electrical Safety Testing of BEDS #70026
Joined: May 2011
Posts: 13
Australia The land that time f...
iozzie Offline OP
Novice
iozzie  Offline OP
Novice

Joined: May 2011
Posts: 13
Australia The land that time f...
Hi All

As part of our contract we are required to perform EST to patient beds. We have to test the beds with the patients in them !!

Just wanted to hear other peoples views from a H&S point of view of how a EST to a patient bed should take place.

Do you have a policy ?

Re: Electrical Safety Testing of BEDS [Re: iozzie] #70028
Joined: Feb 2004
Posts: 13,645
the path less trodden
Geoff Hannis Offline
Super Hero
Geoff Hannis  Offline
Super Hero

Joined: Feb 2004
Posts: 13,645
the path less trodden

... yes; I wouldn't do it. whistle

Other than that, what do your company's lawyers (and insurance brokers) have to say about it? think

And lastly, when you say EST ... which specific tests do you mean? Protective earth continuity only?

Re: Electrical Safety Testing of BEDS [Re: iozzie] #70053
Joined: May 2011
Posts: 13
Australia The land that time f...
iozzie Offline OP
Novice
iozzie  Offline OP
Novice

Joined: May 2011
Posts: 13
Australia The land that time f...
Hi Geoff

We are performing full EST, earth continuity, Insulation resistance, earth leakage and touch current.

Re: Electrical Safety Testing of BEDS [Re: iozzie] #70054
Joined: Feb 2007
Posts: 224
Yorkshire
MikeX Offline
Master
MikeX  Offline
Master

Joined: Feb 2007
Posts: 224
Yorkshire
Clearly there is a potential hazard to the patient if you perform these tests while the patient is still residing in it! In the case of the earth leakage test the protective earth will be broken during the test and potentially exposes the patient to a electric shock risk.

I assume you are using IEC 62353, which states:
Quote:
All tests shall be performed in such manner that no hazardous situations arise for testing personnel, patients or other individuals.

Clearly you are failing in this duty of care when you test while the patient is still connected to the medical device!

Re: Electrical Safety Testing of BEDS [Re: iozzie] #70055
Joined: Feb 2004
Posts: 13,645
the path less trodden
Geoff Hannis Offline
Super Hero
Geoff Hannis  Offline
Super Hero

Joined: Feb 2004
Posts: 13,645
the path less trodden

Primum non nocere! frown

Re: Electrical Safety Testing of BEDS [Re: iozzie] #70063
Joined: Jun 2001
Posts: 242
United Kingdom
Electric Blobby Offline
Master
Electric Blobby  Offline
Master

Joined: Jun 2001
Posts: 242
United Kingdom
We ask the patients who are able to vacate the bed say for their physio or to move around. This is agreed prior and entered to the patient and ward notes and if needed clinical staff are paid for to manage the time out of the bed. If the patient is bed bound then arrangements upon turning the patient incorporates a bed change to substitute the good from bad. In no way do we compromise patient care,safety or excessive sheer issues.
I can send a policy of extract to support this as needed.
Paul


ERRATIC MEANS STATIC SO BE ERRATIC AND NOT STATIC
WE ARE ALL IN THE NHS AND THIS IS
"ERRATIC AND STATIC"
Re: Electrical Safety Testing of BEDS [Re: Electric Blobby] #70095
Joined: Feb 2004
Posts: 13,645
the path less trodden
Geoff Hannis Offline
Super Hero
Geoff Hannis  Offline
Super Hero

Joined: Feb 2004
Posts: 13,645
the path less trodden

I'm wondering what the urgency is in this situation; something contractual, perhaps?

But what does the Risk Assessment say?

Do such beds ever fail an EST? And what happens (ie, to the returning patient) if they do?

Why not just wait until the nursing staff let you know when a bed finally becomes free, and then go along and EST (and PM) it? smile

Re: Electrical Safety Testing of BEDS [Re: Geoff Hannis] #70343
Joined: May 2011
Posts: 13
Australia The land that time f...
iozzie Offline OP
Novice
iozzie  Offline OP
Novice

Joined: May 2011
Posts: 13
Australia The land that time f...
Originally Posted By: Geoff Hannis

I'm wondering what the urgency is in this situation; something contractual, perhaps? Yes we have a contract personally I would not have touched the job as I am sure we will lose out on it!!

But what does the Risk Assessment say? There isn't one

Do such beds ever fail an EST? And what happens (ie, to the returning patient) if they do?

Why not just wait until the nursing staff let you know when a bed finally becomes free, and then go along and EST (and PM) it? smile
We have a very small team for a 600+ bed hospital and it is quite had to do with our PM's to complete and approx 350 breakdowns a month.....

Re: Electrical Safety Testing of BEDS [Re: iozzie] #70345
Joined: Feb 2004
Posts: 13,645
the path less trodden
Geoff Hannis Offline
Super Hero
Geoff Hannis  Offline
Super Hero

Joined: Feb 2004
Posts: 13,645
the path less trodden

Yes, there are only so many hours in a day (working week, whatever).

It sounds like you need to adopt that time-honoured biomeds' approach to the work of tackling tasks in order of priority.

Setting up a Risk-Based PM programme might be a good place to start (regardless of what contracts might "require").

When you assign Risk Levels to the various equipment types you look after, you may well find that carrying out routine electrical safety tests on patient beds comes way down the list (and therefore will probably never get done). I would suggest, "Repair when failed" might be a good maintenance support policy for run-of-the-mill patient beds. smile

Re: Electrical Safety Testing of BEDS [Re: iozzie] #70347
Joined: Oct 2006
Posts: 155
Rome
Malcolm Offline
Mentor
Malcolm  Offline
Mentor

Joined: Oct 2006
Posts: 155
Rome
I would reiterate what Geoff has said above, risk management is the way forward.
By the way, are the beds class I or class II??
We have a similar albeit smaller problem within two hospices in a service contract,frequently we part from our central office with a blue light when the beds become vacant :-((

Malcolm.

Re: Electrical Safety Testing of BEDS [Re: iozzie] #70348
Joined: Oct 2006
Posts: 155
Rome
Malcolm Offline
Mentor
Malcolm  Offline
Mentor

Joined: Oct 2006
Posts: 155
Rome
Geoff....do we read between the lines"We have a very small team" that they are looking for new recruits....Australia is on my bucket list,what about you....

Re: Electrical Safety Testing of BEDS [Re: Malcolm] #70349
Joined: Feb 2004
Posts: 13,645
the path less trodden
Geoff Hannis Offline
Super Hero
Geoff Hannis  Offline
Super Hero

Joined: Feb 2004
Posts: 13,645
the path less trodden

Yes, I like to think I'm pretty good at "reading between the lines". whistle

I have yet to come across any biomed team anywhere that I would describe as "over-manned". frown

That's why I keep banging on about the need (obvious, I would have thought) to prioritize the work.

Even "back in my day" (and often with myself and my gang "busting a gut", you might say), there were often times when we were way behind in what we were "supposed" to be doing - and, yes, sometimes stuff went "un-done". But, as far as I remember, nothing catastrophic resulted from our (my) management of the situation(s).

On your last point - if I were thirty (forty?) years younger:- yes, by all means. I can't see much future for any keen young guy (not me, then) here in the UK. Anytime I meet a "fertile young mind" I always advise him (and it generally is a him) accordingly. smile

Re: Electrical Safety Testing of BEDS [Re: iozzie] #70852
Joined: Mar 2014
Posts: 20
Isle of Man
H
Heather Offline
Dreamer
Heather  Offline
Dreamer
H

Joined: Mar 2014
Posts: 20
Isle of Man
We are never aloud to EST a bed with a patient in it, that's crazy! When ours are done if a patient is too ill to get out of bed, the bed is logged and tagged to be done when available!

Re: Electrical Safety Testing of BEDS [Re: Heather] #70853
Joined: Feb 2004
Posts: 13,645
the path less trodden
Geoff Hannis Offline
Super Hero
Geoff Hannis  Offline
Super Hero

Joined: Feb 2004
Posts: 13,645
the path less trodden

OK ... just to continue (prolong?) the debate here:-

1) What EST(s) do you carry out on beds?
2) Which Standard do you follow?
3) How often do you carry them out?
4) How many have "failed" (if any)?
5) Do you do these EST(s) as part of PM?
6) How are beds that "fail" EST(s) (or PM) dealt with?

Re: Electrical Safety Testing of BEDS [Re: Geoff Hannis] #73104
Joined: May 2011
Posts: 13
Australia The land that time f...
iozzie Offline OP
Novice
iozzie  Offline OP
Novice

Joined: May 2011
Posts: 13
Australia The land that time f...
Sorry I haven't been on for a while

I have won Beds are now pulled out of service for PM's by engineering and then EST'd by biomed before returning

Re: Electrical Safety Testing of BEDS [Re: Geoff Hannis] #73108
Joined: Oct 2006
Posts: 155
Rome
Malcolm Offline
Mentor
Malcolm  Offline
Mentor

Joined: Oct 2006
Posts: 155
Rome
Hi Geoff, I give support to two 50 bed Hospice's here in Rome, as the is no fixed technical presence I give support on a ad hoc basis for repairs and a programmed visit for the PM/EST, as sure as the Pope is a catholic during my PM visit there is almost always 100% bed occupancy.
In one of the locations the mains supply is 220v split phase, end result the Fluke ESA612 reports a no reference earth message, I have therefore decided ( with the consent of the person responsible for the QC) to use use the alternative method of the EN 62353 for the EST's.
I'm "obliged" to perform the EST, where possible, on beds with patients, as this method isolates the DUT from the mains supply and is current limited, this to me is the safest method available apart from not carrying out the EST until the bed is empty. As both structures have a waiting list with a turn around bed occupancy of approximately 12 hours I can't see the alternative solution (open to suggestions....)
Rare failures are always earth bond centered which result in the programmed replacement of the mains cable.

Re: Electrical Safety Testing of BEDS [Re: Malcolm] #73110
Joined: Feb 2004
Posts: 13,645
the path less trodden
Geoff Hannis Offline
Super Hero
Geoff Hannis  Offline
Super Hero

Joined: Feb 2004
Posts: 13,645
the path less trodden

It sounds like you're doing the best you can there, Malcolm. smile

As you know, the Alternative Method does not activate the DUT, and so therefore is not comparable to all 60601 tests (especially leakage currents) ... however, for occupied beds and similar items I should imagine it's more or less ideal!

And ... I'm guessing that where you are, mains polarity is not an issue (ie, can be either way around).

I have encountered the "split phase" situation myself in strange locations many moons ago; our tester didn't pass it's Self Test!

But (more importantly), how good is the building earth in the places you mention? think

Lastly, are those mains cables fixed, or detachable, I wonder?

Re: Electrical Safety Testing of BEDS [Re: Geoff Hannis] #73111
Joined: Oct 2006
Posts: 155
Rome
Malcolm Offline
Mentor
Malcolm  Offline
Mentor

Joined: Oct 2006
Posts: 155
Rome
Fortunately the building earth is very good.The beds are all Hill-Rom with fixed mains.

Re: Electrical Safety Testing of BEDS [Re: Geoff Hannis] #73114
Joined: Feb 2007
Posts: 224
Yorkshire
MikeX Offline
Master
MikeX  Offline
Master

Joined: Feb 2007
Posts: 224
Yorkshire
Originally Posted by Geoff Hannis
As you know, the Alternative Method does not activate the DUT, and so therefore is not comparable to all 60601 tests (especially leakage currents) ... however, for occupied beds and similar items I should imagine it's more or less ideal!
The Alternative method will give you the same result as a open neutral 60601-1 test, so is comparable.

Using the Alternative method offers a massive time saving over all other methods, since you don’t need to reverse the polarity of the mains and you don’t have to wait for systems to boot up and down between tests. It also is safer for both the tester and those in the vicinity of the EUT. No need for insulation tests if you suspect an insulation problem, such as liquid ingress.

This is the reason many equipment manufacturers are switching to using IEC 62353 and the Alternative method. My experience is that very little equipment is unsuitable for the Alternative method.

Re: Electrical Safety Testing of BEDS [Re: MikeX] #73115
Joined: Feb 2004
Posts: 13,645
the path less trodden
Geoff Hannis Offline
Super Hero
Geoff Hannis  Offline
Super Hero

Joined: Feb 2004
Posts: 13,645
the path less trodden

Just out of interest ... I understand (?) that the Alternative Method issues predictive leakage current results.

Does anyone know how (or even better, actually tested) these leakage currents compare with actual values measured using either the Direct Method or traditional 60601 tests?

In other words, how accurate are these predictions? smile

Re: Electrical Safety Testing of BEDS [Re: Geoff Hannis] #73118
Joined: Feb 2004
Posts: 13,645
the path less trodden
Geoff Hannis Offline
Super Hero
Geoff Hannis  Offline
Super Hero

Joined: Feb 2004
Posts: 13,645
the path less trodden

Originally Posted by Geoff Hannis

In other words, how accurate are these predictions?


See here. smile

Re: Electrical Safety Testing of BEDS [Re: iozzie] #73197
Joined: Sep 2010
Posts: 383
Philippines
ariel dulay Offline
Sage
ariel dulay  Offline
Sage

Joined: Sep 2010
Posts: 383
Philippines
..what the use of test instrument by the way?


--- PATIENT SAFETY FIRST ---
Re: Electrical Safety Testing of BEDS [Re: Malcolm] #73198
Joined: Feb 2009
Posts: 1,811
Jeddah, Saudi Arabia
Neil Porter Offline
Hero
Neil Porter  Offline
Hero

Joined: Feb 2009
Posts: 1,811
Jeddah, Saudi Arabia
Healthcare institutes must be single phase, dual phase was, and still is the lazy way to upgrade from 110V to 220V


If you think hiring professionals is expensive, try hiring amateurs!
Re: Electrical Safety Testing of BEDS [Re: Neil Porter] #73199
Joined: Feb 2004
Posts: 13,645
the path less trodden
Geoff Hannis Offline
Super Hero
Geoff Hannis  Offline
Super Hero

Joined: Feb 2004
Posts: 13,645
the path less trodden

Eh? What about 3-phase (for x-ray systems, etc.)? think

I think what you meant to say there, Neil, was that (in some locations) both 110 and 220 VAC may be taken from a supply transformer in star configuration!

But (and perhaps more importantly in such cases) where is the earth connected?

I have always thought that "two-phase power supplies" was an "American idea". I have only ever encountered them in buildings built to "domestic standards" in places like Saudi Arabia (village clinics, and the like). Only once did I come across a piece of "medical equipment" that actually required a 220 VAC two-phase supply in a proper hospital, and that was an ancient baby milk-bottles sterilizer (I think it was) made in the USA.

Re: Electrical Safety Testing of BEDS [Re: ariel dulay] #73200
Joined: Feb 2004
Posts: 13,645
the path less trodden
Geoff Hannis Offline
Super Hero
Geoff Hannis  Offline
Super Hero

Joined: Feb 2004
Posts: 13,645
the path less trodden

Originally Posted by ariel dulay

..what the use of test instrument by the way?


Do you mean the 62353 Alternative Method?

For more, see this earlier thread.

You will need an Safety Tester that can accommodate this (has this "feature").

Re: Electrical Safety Testing of BEDS [Re: iozzie] #73202
Joined: Nov 2007
Posts: 115
HPUiTM, Malaysia
M
mammad Offline
Savant
mammad  Offline
Savant
M

Joined: Nov 2007
Posts: 115
HPUiTM, Malaysia
Going back with the initial queries. In our policy in Malaysia, when we performing PPM inclusive with EST for any medical equipment regardless class 1 or class 2, it shall be isolated from patient. This shall minimize any potential risk exposed to the patient which will lead into severe damage.

Re: Electrical Safety Testing of BEDS [Re: mammad] #73203
Joined: Feb 2004
Posts: 13,645
the path less trodden
Geoff Hannis Offline
Super Hero
Geoff Hannis  Offline
Super Hero

Joined: Feb 2004
Posts: 13,645
the path less trodden

OK, that's the policy. But how do you implement the policy in cases where beds are hardly ever (if at all) vacated?

Are you following 62353 ... and especially, the Alternative Method? think

Re: Electrical Safety Testing of BEDS [Re: Geoff Hannis] #73207
Joined: Nov 2007
Posts: 115
HPUiTM, Malaysia
M
mammad Offline
Savant
mammad  Offline
Savant
M

Joined: Nov 2007
Posts: 115
HPUiTM, Malaysia
We never experienced that.

Normally bed, the frequency for the PPM is once a year and bed hardly vacated is normally in ICU.

We still emphasize in 60601 and not yet 62353.

Re: Electrical Safety Testing of BEDS [Re: mammad] #73208
Joined: Feb 2009
Posts: 1,811
Jeddah, Saudi Arabia
Neil Porter Offline
Hero
Neil Porter  Offline
Hero

Joined: Feb 2009
Posts: 1,811
Jeddah, Saudi Arabia
For ICU, ask the head nurse to inform BME when the room becomes vacant, death, theatre procedure or any other service outside of the ICU and PPM all the equipment for that room.


If you think hiring professionals is expensive, try hiring amateurs!
Re: Electrical Safety Testing of BEDS [Re: Neil Porter] #73209
Joined: Feb 2004
Posts: 13,645
the path less trodden
Geoff Hannis Offline
Super Hero
Geoff Hannis  Offline
Super Hero

Joined: Feb 2004
Posts: 13,645
the path less trodden

1) Patient goes to Theatre
2) Nurse informs BME
3) BME rushes to ICU
4) Bed fails EST (PPM)
5) Bed has a fault
6) Bed needs extra work
7) Bed needs parts
8) Patient returns
9) What happens next?

Re: Electrical Safety Testing of BEDS [Re: Geoff Hannis] #73211
Joined: Feb 2009
Posts: 1,811
Jeddah, Saudi Arabia
Neil Porter Offline
Hero
Neil Porter  Offline
Hero

Joined: Feb 2009
Posts: 1,811
Jeddah, Saudi Arabia
Either the department has a spare bed that they can use, the patient sleeps on the transport stretcher until the bed is repaired or the patient may sleep on the mattress on the floor.

Better to find the fault whilst the patient is out of the bed, rather than the fault causing harm to the patient.


If you think hiring professionals is expensive, try hiring amateurs!
Re: Electrical Safety Testing of BEDS [Re: Neil Porter] #73214
Joined: Feb 2004
Posts: 13,645
the path less trodden
Geoff Hannis Offline
Super Hero
Geoff Hannis  Offline
Super Hero

Joined: Feb 2004
Posts: 13,645
the path less trodden

Not a bad answer, Neil. smile

... or the nursing staff set up a bay on one of the regular wards (hopefully not too far away), and appoint staff from the ICU to keep an eye on the patient there. The theatres Recovery Room may be another option, depending on how long repair of the ICU bed is likely to take.

There may well be other options, depending upon the actual condition of the patient and what kit is needed (and available); for instance, it may be possible to drag an ordinary hospital bed into the ICU temporarily.

Needs must!

Re: Electrical Safety Testing of BEDS [Re: Geoff Hannis] #73217
Joined: Nov 2007
Posts: 115
HPUiTM, Malaysia
M
mammad Offline
Savant
mammad  Offline
Savant
M

Joined: Nov 2007
Posts: 115
HPUiTM, Malaysia
Originally Posted by Geoff Hannis

Not a bad answer, Neil. smile

... or the nursing staff set up a bay on one of the regular wards (hopefully not too far away), and appoint staff from the ICU to keep an eye on the patient there. The theatres Recovery Room may be another option, depending on how long repair of the ICU bed is likely to take.

There may well be other options, depending upon the actual condition of the patient and what kit is needed (and available); for instance, it may be possible to drag an ordinary hospital bed into the ICU temporarily.

Needs must!


a good approached Geoff. Will need to revise my policy then.